Hello Martin,
Thank you for pointing out the adjustability of the slew rate on the UDP6932B. The device has three settings: Normal, CV-Slew Rate, max. 124 V/s, and CC-Slew Rate, max. 21 A/s.
The measurements in Post #97 were taken with the CC-Slew Rate setting at 21 A/s.
Today, I repeated the measurements with the Normal and CV-SlewRate 124 V/s settings.
While the result for Normal is the same as for CC-SlewRate 21 A/s, there is a significant overshoot with the CV-SlewRate 124 V/s setting.
Voltage and current rise to approx. 4 Volts and approx. 84 mA within approx. 50 ms and then fall to the set limit value within approx. 450 ms.
Best Regards
Bernd