Hi all
I often use my RTB for line power measurements, this trace bounce situation can happen with other signals, but this is when I often notice it.
Channel 1 has differential probe connected to measure 120 VAC line @ 60 hz. Channel 4 has the current probe connection. I usually view in 5ms time division for 60 hz, triggered from channel one or line voltage.
At the 5ms time base, the blue trace of channel 4 bounces slightly above and below zero volts at a very noticeable rate. If I change time base setting above or below 5ms, bounce stops. If I change sample rate from auto to 10 MSa or below seems to stop.
If I turn on channel 3, bouncing stops, and turning off channel 4 and connecting current probe to channel 3 bounce stops.
Full power cycle, hard switch turn off, seems to stop this from happening until I change the time base and back to the same condition. 5ms is the only time division where this seems to appear.
Right now I'm using Firmware 02.202, but this also happens on the newer firmware. I have of course done numerous self alignment tests with no error.
Has anyone else experienced this, any ideas? This wasn't an issue in the past.
Thanks
Russell