We finally decided to get one STO1104 to try. And we like it enough to get more.
1. Battery and Charger
When the scope is powered off and supplied AC-DC power adaptor connected, the green power button indicates if the internal battery is being charged.
The green power button blinks to indicate the battery is still being charged. Once the battery is fully charged, the power button will stay on solid green.
When the scope is first turned off with power adaptor attached, the screen shows a large battery icon momentarily. It indicates if the internal battery is charged. (It always starts from zero to 100% before restarting again at the current battery charged precentage.)
When the scope is powered on, check the small battery icon at the lower right corner for battery status. The flash sign seems to appear if the power adaptor is connected.
2. Grounding
The scope itself is grounded via the supplied AC-DC power adaptor.
The resistance between the ground pin on the scope and the earth pin of the supplied power adaptor is about 1K ohm.
If this is not desired, the scope can be operated from its internal battery.
Take extra caution when operating the scope with floating ground.
3. HDMI Connection
The HDMI output is at 800x600 resolution and cannot be changed. Thus it will be the connected device that will perform the scaling to display the scope image on the connected device.
Another thing to watch out for is scope grounding when HDMI cable is connected.
4. Screen Capture
As for firmware 1.0.56, the background color of screen capture is black and it cannot be changed.
Micsig customer support suggested that they will allow a choice of white background in the next firmware release.
5. Serial Decoding
When the scope is put into text-only display serial bus decoding mode, the "AUTO" button does not reset this mode. The only way to exit this text-only display mode is to press the icon from the bottom pull-up on-screen menu again.
We were thinking the "AUTO" button will automatically get the scope back to graphic display but we were wrong.
6. Horizontal and Vertical Cursors
When these are turned on, the corresponding measurement numbers will always appear at the upper right quardant of the screen, which sometimes happen to overlap with the displayed waveforms. Unfortunately there seemed to be no means to move these displayed numbers, i.e. we need to move the overlapping waveforms away.
7. Fan Noise
Micsig has addressed the noisy cooling fan issue from their earlier products by using thermal control on fan speed. However some people still find the fan noise a bit loud, especially for home environments.
8. Vertical Waveform Position
We can use the knob to adjust vertical position of the waveforms. However, this scope has no numerical display of the position unlike a TEK scope where a number corresponding to the vertical postion in divisions is shown during the adjustment.