Results from my SDM4065A-SC, 10PLC, auto-zero on, input short-circuited.
All manual ranges, once without, once with rel function.
I don't see any problems with the meter.
Thank you for the detailed measurements. You say you dont see any problems, but your meter is also out of tolerance when not using Autozero. It confirms the measurements i made earlier.
Obviously you can Rel out the zero point, but that is bad practice for current measurements, since you are now just shifting the entire calibration (so basically everything else is measuring wrong now). Rel should only be used when you want to make a true relative measurement, not as a band-aid for a bad calibration.
And sure, Rel is typically used to subtract unwanted probe / lead capacitance or resistance when measuring Capacitance or Resistance, so effectively zeroing out those measurement modes. But in those cases, the calibration isn't affected, you are just making a true relative measurement.
Autozero is not something i like to use, because it has it own drawbacks as mentioned earlier. Even as Kleinstein said, Autozero is a tradeoff; less drift vs more noise and lower speed. So for accurate measurements i'd rather rely on having Autozero switched off.
Now what is interesting from the Defpom video i posted earlier is that the SDM3065X has a separate calibration for each mode and range with and without Autozero. So there is no reason why the zero point calibration would be out of tolerance when not using Autozero. Take a look at the 4:26 mark in the video:
https://youtu.be/8gZZg605ioI?si=3r1mO64nqXR1L-BL&t=267Interesting right? Assuming the same applies to the 4065A...