Purchased used Fluke 289 - need some advice
« on: March 20, 2019, 10:03:03 pm »
Hi all -

I picked up a used Fluke 289 recently and in inspecting it feel that something is wrong with it.

Having no leads plugged into it, when I set it into mVAC mode, the meter does not even out to 0. It hovers around 5.2 mVAC, putting it into regular VAC mode (again the leads are not connected), it settles to about 0.0315 VAC.

Both DC and mDC settle to around 0.

For anyone with experience with this, is this a faulty meter?

Thank you.
J
 

Offline 0culus

  • Super Contributor
  • ***
  • Posts: 3032
  • Country: us
  • Electronics, RF, and TEA Hobbyist
Re: Purchased used Fluke 289 - need some advice
« Reply #1 on: March 20, 2019, 10:11:28 pm »
If you short the leads together, it should drop to zero. Basically, the leads will act like antennas and pick up local EMI.
 

Offline m3jalaTopic starter

  • Newbie
  • Posts: 8
  • Country: us
Re: Purchased used Fluke 289 - need some advice
« Reply #2 on: March 20, 2019, 10:21:38 pm »
Thank you - I appreciate the response. I'm not seeing it go to 0, see these photos. This isn't normal right?
 

Offline 0culus

  • Super Contributor
  • ***
  • Posts: 3032
  • Country: us
  • Electronics, RF, and TEA Hobbyist
Re: Purchased used Fluke 289 - need some advice
« Reply #3 on: March 20, 2019, 10:31:19 pm »
I think it's fine. My 87-V does about the same thing when you short the leads out like that. Goes down to 0.2 mVac and stays there.
 

Offline helius

  • Super Contributor
  • ***
  • Posts: 3642
  • Country: us
Re: Purchased used Fluke 289 - need some advice
« Reply #4 on: March 20, 2019, 10:32:54 pm »
There are low-frequency electric fields in most places, unless inside a Faraday cage or deep in the woods. The loop enclosed by your leads is enough to pick them up in AC mode (no surprise that they do not register in DC mode). Try putting a shorting bar in the jacks instead of the leads, the much smaller loop area will register as a much lower AC voltage.

If you turn on frequency display, you'll probably see a big fat "60.000 Hz" and understand where that reading is coming from.
« Last Edit: March 20, 2019, 10:34:34 pm by helius »
 
The following users thanked this post: Trader

Offline commongrounder

  • Frequent Contributor
  • **
  • Posts: 372
  • Country: us
Re: Purchased used Fluke 289 - need some advice
« Reply #5 on: March 20, 2019, 10:33:39 pm »
From the 289 manual:

True-rms:
AC mV, AC V, AC μA, AC mA, and AC A specifications are ac-coupled, true rms, and are specified from 2 % of range to 100 % of range, except 10 A range is specified from 10 % to 100 % of range.

AC Floor:
When the input leads are shorted together in the ac functions, the Meter may display a residual reading up to 200 counts. A 200 count residual reading will cause only a 20 count change for readings at 2 % of range. Using REL to offset this reading may produce a much larger constant error in later measurements.
 
The following users thanked this post: Marco1971

Online rsjsouza

  • Super Contributor
  • ***
  • Posts: 5986
  • Country: us
  • Eternally curious
    • Vbe - vídeo blog eletrônico
Re: Purchased used Fluke 289 - need some advice
« Reply #6 on: March 20, 2019, 10:34:26 pm »
Are the probes shorted in your third photo?

If so, keep in mind the meter is measuring 300µV, which is well within the noise margin of lots of True RMS converters.

In this case, can you try to clean the probes thoroughly and rotate its banana jacks? Given the magnitude of the error, I wouldn't be surprised if grease is the culprit here.
« Last Edit: March 21, 2019, 01:28:26 pm by rsjsouza »
Vbe - vídeo blog eletrônico http://videos.vbeletronico.com

Oh, the "whys" of the datasheets... The information is there not to be an axiomatic truth, but instead each speck of data must be slowly inhaled while carefully performing a deep search inside oneself to find the true metaphysical sense...
 

Offline 0culus

  • Super Contributor
  • ***
  • Posts: 3032
  • Country: us
  • Electronics, RF, and TEA Hobbyist
Re: Purchased used Fluke 289 - need some advice
« Reply #7 on: March 20, 2019, 10:38:51 pm »
Yep, it's almost certainly 60 hz hum you're seeing. If you have an oscilloscope you will see it there as well, from a non grounded probe.
 

Offline mr.fabe

  • Regular Contributor
  • *
  • Posts: 171
  • Country: us
Re: Purchased used Fluke 289 - need some advice
« Reply #8 on: March 20, 2019, 10:41:25 pm »
I just checked my 289 and it exhibits similar readings.  My unit was recently repaired with a new board swap and calibration.

You may want to check your FW version, check the super cap and also do a low ohms test.  Put your unit in Low Ohms mode, short your leads and if the display starts to fade, your unit is defective.
« Last Edit: March 20, 2019, 10:47:13 pm by mr.fabe »
 

Offline m3jalaTopic starter

  • Newbie
  • Posts: 8
  • Country: us
Re: Purchased used Fluke 289 - need some advice
« Reply #9 on: March 20, 2019, 11:23:53 pm »
Thanks all, I appreciate all your responses.

Mr. Fabe - I'm on an old FW version (1.0) with calibration from '07. Already checked the supercap, appears fine. Just ran the low ohms mode you mentioned, no fading.

When nothing is plugged in, I read 170 hz and 4.375 mVAC.

When I put a shorting bar (a paperclip) into the jacks, I read 0hz and 0.281 mVAC.

I'm in a townhouse area with a ton of power lines, computers, wifi and neighbors, so sounds about right for the background noise assessment?

Thanks much all.
 

Offline 0culus

  • Super Contributor
  • ***
  • Posts: 3032
  • Country: us
  • Electronics, RF, and TEA Hobbyist
Re: Purchased used Fluke 289 - need some advice
« Reply #10 on: March 20, 2019, 11:25:33 pm »
Yep that sounds about right. I only see 60 Hz where I am, but my neighbors are not packed in either!
 

Offline m3jalaTopic starter

  • Newbie
  • Posts: 8
  • Country: us
Re: Purchased used Fluke 289 - need some advice
« Reply #11 on: March 20, 2019, 11:29:53 pm »
Perfect, off to kick off some projects with my new to me 289 in that case! Thank you.
 

Offline 0culus

  • Super Contributor
  • ***
  • Posts: 3032
  • Country: us
  • Electronics, RF, and TEA Hobbyist
Re: Purchased used Fluke 289 - need some advice
« Reply #12 on: March 20, 2019, 11:41:54 pm »
No problem! BTW, welcome to the forum! Visit us in Test Equipment Anonymous if you're into test gear.  :-DMM  :-DD
 

Offline m3jalaTopic starter

  • Newbie
  • Posts: 8
  • Country: us
Re: Purchased used Fluke 289 - need some advice
« Reply #13 on: March 21, 2019, 02:14:10 am »
I'm definitely easing into it. I have to say, can already tell this is a great board with all the awesome responses.

I come here on a couple car projects of mine that require electrical work. Figured I'd teach myself the trade while at it :box:

Take care and see you around.
 
The following users thanked this post: 0culus

Offline mr.fabe

  • Regular Contributor
  • *
  • Posts: 171
  • Country: us
Re: Purchased used Fluke 289 - need some advice
« Reply #14 on: March 21, 2019, 04:34:54 am »
You may want to consider upgrading your firmware or use electrical tape on your optical sensor on top of your unit.  The current firmware available for the 289 is v1.16.

From the Fluke website...

Version 1.16 corrects an issue where recording sessions could be interrupted or terminated when using the record function in combination with ambient conditions of bright sunlight or other sources of bright light, when used without the IR cable attached. Version 1.16 does not apply for units labeled 287/TPWS or 287/NUC

If you feel the need to update your firmware, you should contact Fluke technical support to make sure calibration is not required after updating your unit.  Version 1.0 is pretty old...
 

Offline m3jalaTopic starter

  • Newbie
  • Posts: 8
  • Country: us
Re: Purchased used Fluke 289 - need some advice
« Reply #15 on: March 21, 2019, 01:21:27 pm »
Yes, thank you I saw that.
 

Offline agdr

  • Supporter
  • ****
  • Posts: 105
  • Country: us
    • agdr Audio
Re: Purchased used Fluke 289 - need some advice
« Reply #16 on: March 23, 2019, 10:39:07 am »

Having no leads plugged into it, when I set it into mVAC mode, the meter does not even out to 0. It hovers around 5.2 mVAC, putting it into regular VAC mode (again the leads are not connected), it settles to about 0.0315 VAC.

Both DC and mDC settle to around 0.


I also just received two 289's back from Fluke for repair (supercaps) and calibration a couple of weeks ago.  They do exactly the same.  Looks like Fluke did board swaps with recent (mfg date 2 years ago) boards in both.
« Last Edit: March 23, 2019, 10:44:30 am by agdr »
 

Offline Exhale

  • Newbie
  • Posts: 6
  • Country: ru
Re: Purchased used Fluke 289 - need some advice
« Reply #17 on: June 08, 2019, 03:03:54 pm »
Hello.

I'm wondering how accurate is capacitance measurement function on a Fluke 289 in reality ?
Manual says - range 1.000 nF +-1% 5 counts.
So it means +-15 pf ?
Can it measure accurate at least from 4pf ?

How it performs compared to DE-5000 dedicated LCR ?

I'm deciding between a used 289 like all purpose DMM.
Or a combination of HP 34401a + DE-5000 (also both used)
I have a place where to get them calibrated.

Thanks.
 

Online rsjsouza

  • Super Contributor
  • ***
  • Posts: 5986
  • Country: us
  • Eternally curious
    • Vbe - vídeo blog eletrônico
Re: Purchased used Fluke 289 - need some advice
« Reply #18 on: June 08, 2019, 06:22:42 pm »
Hello.

I'm wondering how accurate is capacitance measurement function on a Fluke 289 in reality ?
Manual says - range 1.000 nF +-1% 5 counts.
So it means +-15 pf ?
Can it measure accurate at least from 4pf ?

How it performs compared to DE-5000 dedicated LCR ?

I'm deciding between a used 289 like all purpose DMM.
Or a combination of HP 34401a + DE-5000 (also both used)
I have a place where to get them calibrated.

Thanks.
Capacitance meters on DMMs are good for a quick check, but a DE-5000 is a much better choice, especially for this range of low capacitances. If you see yourself needing that range quite frequently, the 289 is really not an option (nor any DMM I know, really).

On the other hand, the bench meter is not portable. It is a very nice choice if you do not need such portability.
Vbe - vídeo blog eletrônico http://videos.vbeletronico.com

Oh, the "whys" of the datasheets... The information is there not to be an axiomatic truth, but instead each speck of data must be slowly inhaled while carefully performing a deep search inside oneself to find the true metaphysical sense...
 

Offline Exhale

  • Newbie
  • Posts: 6
  • Country: ru
Re: Purchased used Fluke 289 - need some advice
« Reply #19 on: June 08, 2019, 09:05:25 pm »
Capacitance meters on DMMs are good for a quick check, but a DE-5000 is a much better choice, especially for this range of low capacitances. If you see yourself needing that range quite frequently, the 289 is really not an option (nor any DMM I know, really).

On the other hand, the bench meter is not portable. It is a very nice choice if you do not need such portability.

I think I'll go with an option two.
I don't need a handheld.
Hate changing batteries or charging accumulators.
 

Offline Exhale

  • Newbie
  • Posts: 6
  • Country: ru
Re: Purchased used Fluke 289 - need some advice
« Reply #20 on: June 09, 2019, 09:39:27 am »
Actually I made a mistake regarding accuracy of Fluke's cap meter.
Manual states that:

"Accuracy specifications are given as: ±( [ % of reading ] + [ number of least significant digits ] )."

That means in example if we measure 10pF, +- 5 counts.

Now it looks exactly the same as DE-5000 table : 1.2%+5
All accuracy is specified as ±[(% of reading) + (value of least
significant digit)]

DE-5000

FLUKE


So, in theory, no difference ?
« Last Edit: June 09, 2019, 09:50:32 am by Exhale »
 

Online rsjsouza

  • Super Contributor
  • ***
  • Posts: 5986
  • Country: us
  • Eternally curious
    • Vbe - vídeo blog eletrônico
Re: Purchased used Fluke 289 - need some advice
« Reply #21 on: June 09, 2019, 10:39:46 am »
Yes, there is a difference. You are talking about a resolution of 0.01pF for the DER EE DE-5000 versus 1pF for the 289 for low capacitances. The number of counts will influence the precision at the lower limit of the range, which is unfavourable for the 289.

https://www.keysight.com/main/editorial.jspx?cc=US&lc=eng&ckey=1000000572:epsg:faq&nid=-11143.0.00&id=1000000572:epsg:faq
Vbe - vídeo blog eletrônico http://videos.vbeletronico.com

Oh, the "whys" of the datasheets... The information is there not to be an axiomatic truth, but instead each speck of data must be slowly inhaled while carefully performing a deep search inside oneself to find the true metaphysical sense...
 

Offline Exhale

  • Newbie
  • Posts: 6
  • Country: ru
Re: Purchased used Fluke 289 - need some advice
« Reply #22 on: June 09, 2019, 11:01:03 am »
That changes everything.
Thanks for your reply.
I missed that "resolution" column.
0.01pF vs 0.001nF is definately a difference.
;)
« Last Edit: June 09, 2019, 11:04:49 am by Exhale »
 