your English is much better than my Spanisch ...
OK, the 15V are ok, the regulators are set high by a silicon diode each, so it should be about +15.6V and -15.6V . (The voltages in the schematic are for information only, not a must be!) The 20V after the rectifier i.e. are about 24V depending on the mains voltage.
Your reference is ok, I have seen references from 6.91V to 7.01V, your voltage matches.
The +14V and -14V have to be the exact double of your reference, this matches, too.
But there are some other parts wich are influencing the A/D converter, i.e. the resistors in the TF-245 network.
My K2000 has an issue with this TF-245, one resistor has changed its value. So there is a 101.2 error.
Good luck
EMUD