Still can't figure out howto upgrade my DHO914S into DHO924S....
Had a DHO800 upgraded to DHO924 but can't know for this extended version. Any help appreciated.
You need a vendor.bin for DHO924S.
(you can make your own using the script that's floating around)
specificity is much more helpful, here i pin locked among the most important info, not sure if its still applicable (any hack related discussion should be there fwiw)
https://www.eevblog.com/forum/testgear/hacking-the-rigol-dho800900-scope/msg5254071/#msg5254071 or the another...
https://www.eevblog.com/forum/testgear/hacking-the-rigol-dho800900-scope/msg5148330/#msg5148330

Nature: Evolution and the Illusion of Randomness (Stephen L. Talbott): Its now indisputable that... organisms “expertise” contextualizes its genome, and its nonsense to say that these powers are under the control of the genome being contextualized - Barbara McClintock
 

Offline Houseman

  • Regular Contributor
  • *
  • Posts: 176
  • Country: it
Re: Rigol's New DHO800 Oscilloscope unbox & teardown
« Reply #1951 on: February 10, 2024, 07:16:37 am »
Still can't figure out howto upgrade my DHO914S into DHO924S....
Had a DHO800 upgraded to DHO924 but can't know for this extended version. Any help appreciated.
You need a vendor.bin for DHO924S.
(you can make your own using the script that's floating around)
specificity is much more helpful, here i pin locked among the most important info, not sure if its still applicable (any hack related discussion should be there fwiw)
https://www.eevblog.com/forum/testgear/hacking-the-rigol-dho800900-scope/msg5254071/#msg5254071 or the another...
https://www.eevblog.com/forum/testgear/hacking-the-rigol-dho800900-scope/msg5148330/#msg5148330
Thank You so much. I had the ability to successfully upgrade my old DHO804 into 924 using ready-to-use fungus method explained here at reply #111 but when I got the 914S in my hands I was afraid that upgrading it to upper bandwidth with same method I would loose the native S feature..
By the way during upgrade and revert to original 804 I had to manually hard reset the oscope by plugging out the powercord (even with the alert on his thread) since the adb reboot option didn't work for me in either phases.. The oscope was freezed durning reboot and even after 15minutes it was still hung in the middle of nowhere.. And that was some really panic moments, expecially the first time.
And also I have lost the thread discussion update since then ( from reply 111 to 1950) wow it's a growth..
Thanks for the help and the link
 

Offline Fungus

  • Super Contributor
  • ***
  • Posts: 16677
  • Country: 00
Re: Rigol's New DHO800 Oscilloscope unbox & teardown
« Reply #1952 on: February 10, 2024, 07:22:31 am »
By the way during upgrade and revert to original 804 I had to manually hard reset the oscope by plugging out the powercord (even with the alert on his thread) since the adb reboot option didn't work for me in either phases..

Yeah, I don't think reboot syncs the file system. I've had a couple of fails with that.

You can do do "adb sync" before reboot but these days I do this:

Code: [Select]
adb shell kill -9 $(pidof com.rigol.scope)
That kills the 'scope app.

It comes back automatically after a few seconds as the new model, it's much faster than rebooting.

(nb. You need to be root, do "adb root" if you aren't already)
« Last Edit: February 10, 2024, 07:24:51 am by Fungus »
 
The following users thanked this post: Houseman

Online shapirus

  • Super Contributor
  • ***
  • Posts: 1360
  • Country: ua
Re: Rigol's New DHO800 Oscilloscope unbox & teardown
« Reply #1953 on: February 10, 2024, 04:38:00 pm »
Has someone found a bag/case for the scope that would be just right size-wise and ideally having internal compartments for the PSU and probes?

I know that there is some rigol-branded bag made specifically for this scope, but I'm looking for what's available specifically on Aliexpress -- ordering elsewhere is currently non-viable logistics-wise for me.

Here's two that I've found so far that seem to be the best:

https://www.aliexpress.com/item/33004483750.html -- bigger, has a separated compartment inside, but a little too big, could be smaller:





https://www.aliexpress.com/item/1005006312639453.html -- smaller, cheaper, no separate compartment.




If someone has found something better, please share.
 

Offline Houseman

  • Regular Contributor
  • *
  • Posts: 176
  • Country: it
Re: Rigol's New DHO800 Oscilloscope unbox & teardown
« Reply #1954 on: February 11, 2024, 01:42:26 pm »
Has someone found a bag/case for the scope that would be just right size-wise and ideally having internal compartments for the PSU and probes?

I know that there is some rigol-branded bag made specifically for this scope, but I'm looking for what's available specifically on Aliexpress -- ordering elsewhere is currently non-viable logistics-wise for me.

Here's two that I've found so far that seem to be the best:

https://www.aliexpress.com/item/33004483750.html -- bigger, has a separated compartment inside, but a little too big, could be smaller:

(Attachment Link)



https://www.aliexpress.com/item/1005006312639453.html -- smaller, cheaper, no separate compartment.

(Attachment Link)


If someone has found something better, please share.
me too even if there is a dedicated thread for the bag here:
https://www.eevblog.com/forum/testgear/rigol-dho800900-case-options/msg5235294/#msg5235294.
Some user suggests a projector case.
I guess Aliexpress can fully comply you needs.
Regards.
« Last Edit: February 11, 2024, 01:44:47 pm by Houseman »
 

Online shapirus

  • Super Contributor
  • ***
  • Posts: 1360
  • Country: ua
Re: Rigol's New DHO800 Oscilloscope unbox & teardown
« Reply #1955 on: February 11, 2024, 05:30:48 pm »
Yeah I didn't realize there was a separate thread, thanks for the hint.
Not much info there, though.
 

Offline Randy222

  • Frequent Contributor
  • **
  • Posts: 643
  • Country: ca
Re: Rigol's New DHO800 Oscilloscope unbox & teardown
« Reply #1956 on: February 12, 2024, 09:28:37 pm »
Needing some teardown info.

Wondering where the fan wires actually terminate? Do they land on just Vcc-Gnd, or do they land on some pins of a chip?
I ask because the rigol start script makes commented reference to fan PWM KLM, but the ko is not in the GEL FW set.

Code: [Select]
# 加载 PWM Fan 驱动
# insmod /rigol/driver/pwm_fan.ko
# chmod 777 /dev/pwm_fan


I confirm, in the 1000/4000 GEL they enable PWM FAN and it includes the KLM pwm_fan.ko binary.

I suspect 800/900 does not have pwn chip used for pwm, but I don't know 100%.

I will take the 1000/4000 KLM ko and put it on my 804, adjust start script, and see what happens.

Some strings from the pwm_fan.ko
Code: [Select]
pwm-fan
license=GPL
description=PWM FAN driver
alias=platform:pwm-fan
author=Kamil Debski <k.debski@samsung.com>
« Last Edit: February 12, 2024, 09:37:55 pm by Randy222 »
 

Online shapirus

  • Super Contributor
  • ***
  • Posts: 1360
  • Country: ua
Re: Rigol's New DHO800 Oscilloscope unbox & teardown
« Reply #1957 on: February 17, 2024, 07:02:27 pm »
-3db bandwidth measurement:



(spoiler: better than advertised; measures over 300 MHz)
 

Offline norbert.kiszka

  • Regular Contributor
  • *
  • Posts: 227
  • Country: pl
Re: Rigol's New DHO800 Oscilloscope unbox & teardown
« Reply #1958 on: February 17, 2024, 10:18:23 pm »
(spoiler: better than advertised; measures over 300 MHz)

Bit surprise for me, because I was expecting way more. It has same input stages as in DHO4000. Maybe it was limited by capacitor (different value than DHO4000) or by ADC.

Or... maybe its a generator/coax fault of author of that test.

Online shapirus

  • Super Contributor
  • ***
  • Posts: 1360
  • Country: ua
Re: Rigol's New DHO800 Oscilloscope unbox & teardown
« Reply #1959 on: February 17, 2024, 10:24:28 pm »
Bit surprise for me, because I was expecting way more. It has same input stages as in DHO4000. Maybe it was limited by capacitor (different value than DHO4000) or by ADC.

Or... maybe its a generator/coax fault of author of that test.
Rise time measurements also indicate max bandwidth on the order of ~300 MHz.
 

Offline norbert.kiszka

  • Regular Contributor
  • *
  • Posts: 227
  • Country: pl
Re: Rigol's New DHO800 Oscilloscope unbox & teardown
« Reply #1960 on: February 17, 2024, 10:38:05 pm »
Rise time measurements also indicate max bandwidth on the order of ~300 MHz.

DHO800 and DHO900 series has 1.25 GSa/s sample rate. That means one sample per 0.8 ns. 300 MHz is equal to rise time of 1.1(6) ns (about one and half sample time). So its practically impossible to make precise measure of rise time in this scope.
« Last Edit: February 17, 2024, 10:39:56 pm by norbert.kiszka »
 

Offline Fungus

  • Super Contributor
  • ***
  • Posts: 16677
  • Country: 00
Re: Rigol's New DHO800 Oscilloscope unbox & teardown
« Reply #1961 on: February 17, 2024, 10:44:10 pm »
I measured ~280Mhz on my DHO804 with a DHO924 vendor.bin.
 

Online shapirus

  • Super Contributor
  • ***
  • Posts: 1360
  • Country: ua
Re: Rigol's New DHO800 Oscilloscope unbox & teardown
« Reply #1962 on: February 17, 2024, 10:47:01 pm »
I measured ~280Mhz on my DHO804 with a DHO924 vendor.bin.
Which method? -3db amplitude?
 

Offline Fungus

  • Super Contributor
  • ***
  • Posts: 16677
  • Country: 00
Re: Rigol's New DHO800 Oscilloscope unbox & teardown
« Reply #1963 on: February 17, 2024, 11:16:08 pm »
I measured ~280Mhz on my DHO804 with a DHO924 vendor.bin.
Which method? -3db amplitude?

Pulse.

But other forum members have done -3dB and got the same results.

 

Online Aleksandr

  • Contributor
  • Posts: 46
  • Country: ru
Re: Rigol's New DHO800 Oscilloscope unbox & teardown
« Reply #1964 on: February 18, 2024, 04:32:24 am »
Here is a link to increasing the frequency to 400 MHz; replacing inductors expands the band. https://4pda.to/forum/index.php?showtopic=1080959#entry128030525 Post №21-30.
« Last Edit: February 18, 2024, 04:37:44 am by Aleksandr »
 
The following users thanked this post: norbert.kiszka

Offline norbert.kiszka

  • Regular Contributor
  • *
  • Posts: 227
  • Country: pl
Re: Rigol's New DHO800 Oscilloscope unbox & teardown
« Reply #1965 on: February 18, 2024, 04:58:17 am »
Here is a link to increasing the frequency to 400 MHz; replacing inductors expands the band. https://4pda.to/forum/index.php?showtopic=1080959#entry128030525 Post №21-30.

Google translate doesnt helps well. I see on images LC filters, so Im guessing I should change inductors to 0 ohm resistor and maybe remove capacitors.

Online Aleksandr

  • Contributor
  • Posts: 46
  • Country: ru
Re: Rigol's New DHO800 Oscilloscope unbox & teardown
« Reply #1966 on: February 18, 2024, 05:30:10 am »
Here is a link to increasing the frequency to 400 MHz; replacing inductors expands the band. https://4pda.to/forum/index.php?showtopic=1080959#entry128030525 Post №21-30.

Google translate doesnt helps well. I see on images LC filters, so Im guessing I should change inductors to 0 ohm resistor and maybe remove capacitors.

It is enough to replace the blue inductors with inductors with a nominal value of 12-15 nH.
« Last Edit: February 18, 2024, 05:32:32 am by Aleksandr »
 

Offline norbert.kiszka

  • Regular Contributor
  • *
  • Posts: 227
  • Country: pl
Re: Rigol's New DHO800 Oscilloscope unbox & teardown
« Reply #1967 on: February 18, 2024, 05:37:09 am »
It is enough to replace the blue inductors with inductors with a nominal value of 12-15 nH.

I know sample rate will be the same but higher bandwidth (frequency) will make possible to see properly square waves at higher frequency (1/10 - 1/5 of bandwidth).

Online ebastler

  • Super Contributor
  • ***
  • Posts: 6510
  • Country: de
Re: Rigol's New DHO800 Oscilloscope unbox & teardown
« Reply #1968 on: February 18, 2024, 06:55:39 am »
I know sample rate will be the same but higher bandwidth (frequency) will make possible to see properly square waves at higher frequency (1/10 - 1/5 of bandwidth).

But only in single-channel mode. As soon as you use a second (or even a third) channel, aliasing will result because the sampling rate no longer meets the Nyquist requirement.

It will also not be trivial to properly feed the signal into the scope, given that the input section offers no internal 50 Ohm termination. So to "properly see" that fast-edged square wave, and e.g. know whether any ringing you see (or don't see) is real or due to your probing technique, will be a challenge.
 

Online Mechatrommer

  • Super Contributor
  • ***
  • Posts: 11652
  • Country: my
  • reassessing directives...
Re: Rigol's New DHO800 Oscilloscope unbox & teardown
« Reply #1969 on: February 18, 2024, 08:33:04 am »
no if you have 50 ohm bandwidth limiter at the dso input ;D
Nature: Evolution and the Illusion of Randomness (Stephen L. Talbott): Its now indisputable that... organisms “expertise” contextualizes its genome, and its nonsense to say that these powers are under the control of the genome being contextualized - Barbara McClintock
 

Online shapirus

  • Super Contributor
  • ***
  • Posts: 1360
  • Country: ua
Re: Rigol's New DHO800 Oscilloscope unbox & teardown
« Reply #1970 on: February 18, 2024, 09:03:02 am »
It will also not be trivial to properly feed the signal into the scope, given that the input section offers no internal 50 Ohm termination. So to "properly see" that fast-edged square wave, and e.g. know whether any ringing you see (or don't see) is real or due to your probing technique, will be a challenge.
It is actually the case as it is, at the existing max ~300 MHz BW: fast edges produce significant overshoot and ringing when using a probe, potentially making the scope calculate the actual rise time incorrectly, even when the spring ground connection is used, and they disappear to leave a nice clean edge when even a simple termination via a T-connector is done (an inline feed-through terminator may be even better -- I don't yet have one to check).
 

Offline zelea2

  • Regular Contributor
  • *
  • Posts: 61
  • Country: gb
Re: Rigol's New DHO800 Oscilloscope unbox & teardown
« Reply #1971 on: February 18, 2024, 02:57:41 pm »
Has someone found a bag/case for the scope that would be just right size-wise and ideally having internal compartments for the PSU and probes?
This is what I've got for my scope for 17$ delivered to UK:
https://www.aliexpress.com/item/1005005213947242.html (select large size)
Size-wise I'd say it is a perfect fit, it has space for the power supply, mains cable + 4 probes.
The case is reasonably hard but the screen will not survive if you bang it against a hard corner or if you place a heavy object on it.
I will probably add inside a layer of foam just to cover the LCD.
 
The following users thanked this post: myf, shapirus

Offline MalikovVV

  • Newbie
  • Posts: 4
  • Country: by
Re: Rigol's New DHO800 Oscilloscope unbox & teardown
« Reply #1972 on: February 19, 2024, 03:31:13 pm »
Wanbo X1 Max
 
The following users thanked this post: shapirus

Offline profrook

  • Newbie
  • Posts: 3
  • Country: pl
Re: Rigol's New DHO800 Oscilloscope unbox & teardown
« Reply #1973 on: February 20, 2024, 04:01:49 pm »
Just quick question, anyone with root access tried to change DPI in Android on these devices?
 

Offline Randy222

  • Frequent Contributor
  • **
  • Posts: 643
  • Country: ca
Re: Rigol's New DHO800 Oscilloscope unbox & teardown
« Reply #1974 on: February 20, 2024, 05:17:20 pm »
Just quick question, anyone with root access tried to change DPI in Android on these devices?
DPI of the DHO touchscreen? I think that was attempted in the hacking thread, IIRC no real gains.
I believe the driver for the touchscreen uses "max dpi" of the device (screen).
 


Share me

Digg  Facebook  SlashDot  Delicious  Technorati  Twitter  Google  Yahoo
Smf