It has only been a week since I received the R&S CMU200 tester, after replacing the HDD and installing a PCMCIA to SD card interface tonight, I thought I am good to go. Unfortunately, I think I just discovered a serious problem in my analyzer: while sweeping 10 MHz - 40 MHz using the spectrum analyzer, it reports an error message: Overload at connector RF 1! The error is reported every time it completes a sweep
. But the problem is, nothing is connected to RF 1
. See the screenshot attached.
It can still detect the signal within the range, but every time it sweeps, it reports the error message. It only happens when I'm sweeping 10 MHz - 40 MHz, if I shift the sweep range a little bit, I can no longer reliably reproduce the problem. The problem doesn't seem to occur at elsewhere, it can perfectly sweep around 100 MHz a 1 GHz without reporting an overload. Also, the error only occurs at a fine resolution bandwidth, I can only reliably reproduce the problem with a RBW around 5 kHz or less. The only other abnormality is a spur at 38.8 MHz.
The spectrum analyzer is in manual mode, with a maximum power level set to 30 dBm, and a reference level of -20 dBm, there's no way it can overload the input.
Even worse, in the R&S operating and service manuals, there's no any explanation of this error message. I'm going to ask the vendor if he knows anything about it, but I don't think the answer will be positive.
What on Earth could create such a strange issue?! I suspect the problem is only affecting one signal path, that would explain the symptoms. Unfortunately, the official block diagram doesn't show any detail on the operation of the instrument besides three IFs.