Could someone please help check if I am interpreting the attached datasheet correctly?
- Resolution varies from 0.1A to 1A over the measurement range 0.5A to 2000A (not sure if I've understood this...)
- The accuracy on the 600A setting is +/- 1%
Yes, it looks like the current has a resolution of 0.1A, probably up to 999.9A, and then 1000A above will be a 1A resolution.
The accuracy on the 600A range is +/-1% +12 digits, on the 2000A range it is +/-1% +8 digits.
Therefore if a circuit has a actual current of 1.5A, I should expect the meter would measure a value anywhere between 1.485A to 1.515A, but since the resolution is 0.1A it would simply display 1.5A?
If a circuit has a actual current of 15A, I should expect the meter would measure a value anywhere between 14.85A to 15.15A, but since the resolution is 0.1A it would display a value anywhere between 14.9A to 15.2A?
For an actual current of 1.5A, the meter would display anywhere between 0.5A and 2.7A, at 15A the reading would be between 13.7A and 16.3A.
The downside to using current clamps, is that the accuracy is affected by the placement of the conductor within the jaws, this is excaserbated by the low current being measured using a high relatively high 600A range. In practice, with careful use, you are likely to achieve better accuracy than stated, but the manufacturer is likely to specify worst case scenario.
The meter does have some interesting functionality, but if you want better accuracy at the lower amps, you would probably be better off looking for a meter with a lower range than 600A.
Kind regards