Nice, she's looking happier! - The circuitry in these meters are full of JFETs acting as switches... they can go leaky and cause issues. I had a problem with a Filter module, where the switching between different filter sections was not working well - causing the Average reading to be significantly lower than the plain reading (because Average mode automatically turns on the "heavy" filter section). I can easily imagine a leaky FET in a module loading the analog bus and causing false readings, if the module cannot properly disconnect from the bus.
I made some progress with my Unit 4 with the noisy ohms: I looked at the sense current with another meter and an oscilloscope, comparing the noisy unit to a good one. The current turned out to be equally noisy on both the "near perfect" unit and the broken one - so there is something else wrong than the current generator circuit. More work for another day!