I recently bought an expensive toy - a Siglent SSA2021X spectrum analyzer. I don't want to harm this not only because of the high purchase price. Hence my question, for simple tests and measurements on a Breadboard and with a signal generator in the 5 to 12V DC range, is a 20 dB attenuator sufficient as protection?
SSA2021X RF input
50V DC max - I have as additional protection a DC blocker
+30dBm max - which is one Watt (if I remember right)
With -20 dB from my external Siglent attenuator (20dB@4GHz) the SSA2021X should not be damaged till +50dBm ? Or is this false?
With my Oszilloskope the matter is much more relaxed, with 1 MΩ Input (instead 50Ω on the SSA), and the 1:10 probe you are relatively safe here with voltages not exceeding 24 V. I have no practical experience with specific RF performance. I don't want to analyze transmitters either, but rather interference signals in circuits, RF noise in my home office (switching power supplies, lamps).
« Last Edit: January 13, 2024, 02:40:34 am by fnaumann »
Offline Kim Christensen
Re: SSA2021X - 20 dB attenuator sufficient?
« Reply #1 on: January 13, 2024, 03:28:40 am »
Partially true. +50dbm is 100 watts, so your 20db attenuator would also have to be rated for 100W. If it's not, and it burns out when you put 100W into it, you might get more than +30dbm into your SA.
12VDC into 50 ohms is 2.9 watts, so that would damage the typical 1W attenuator, but your SA is rated for 50Vdc with the DC blocker so it should be OK in that scenario.
The following users thanked this post: pdenisowski, fnaumann
« Reply #2 on: January 13, 2024, 08:33:06 pm »
The situation is worse than expected.
"The maximum input values of +30 dB only apply if all attenuators are present in the device are raised to maximum."
https://www.microcontroller.net/topic/521714
I read on the internet that the direction of attenuators also matters and that you shouldn't buy cheap Chinese parts. Regardless of the fact that the expensive equipment sold in Europe is also produced in China, my "Siglent ATT-20 20 dB Attenuator" (https://www.siglent.eu/product/1141970/siglent-att-20-20-db-attenuator) has no label which direction is the correct one. Similar issue with my BLK-222+ DC blocker.
The Tracking generator of the Siglent SSA2021X
Output level -20 dBm~0 dBm
Output level resolution 1 dB
Output flatness +/-3 dB
should not harm when connect to the input ?
« Last Edit: January 13, 2024, 08:41:44 pm by fnaumann »
« Reply #3 on: January 13, 2024, 09:18:40 pm »
There should be no problem connecting the output of the SSA2021X's tracking gen to the SSA2021X's input. Usually this is how you calibrate it anyway.
I can see direction being an issue with some high power attenuators, but not with a 1W one or less.
For example, if the input is on the left you can see there'd be more power dissipated in R1 than in R3. So on some larger attenuators I can see them making R1 higher wattage than R3. But on a low power one, they are usually the same wattage and thus 100% bi-directional:
« Reply #4 on: January 13, 2024, 09:54:57 pm »
That was my first thought, but the many warnings I found on the Internet about spectrum analyzers confused me.
Here (see attached picture) are two of my 20 dB, 50Ω  which covers a frequency range till 4GHz. One slim (with BNC connectors) for 1W (30 dBm ?) and a fat one for 10W. If I understand you right, I should not use the 10W attenuator as input protection for my SSA2021X?
based on https://www.siglenteu.com/wp-content/uploads/dlm_uploads/2017/10/SSA3000X_DataSheet_DS0703X-E04A.pdf
Maximum average RF power 30 dBm, 3 minutes, fc≥10 MHz, attenuation >20 dBm, preamp off
Maximum damage level 33 dBm, fc≥10 MHz, attenuation >20 dBm, preamp off
The output level of the Tracking Generator looks save
I know my questions are at Newbi level. Could we compare it with the DC area. Two 1Ω resistors, one with 5W and the other with 50W. If the voltage and therefore the current flow becomes too high, the resistor burns out, depending on its respective power value. When this burns down, the bond is dead and no damage can be done to the other side of the resistor. If the same applies to attenuator, what's the problem? Or does the attenuator then switch fully on without any attenuation?
« Last Edit: January 13, 2024, 10:55:58 pm by fnaumann »
« Reply #5 on: January 13, 2024, 10:47:16 pm »
The 10W probably doesn't matter which way you connect it. If it did matter, they would label the input and output. It's spec sheet would also tell you if it mattered.
It should be safe to use for situations where you expect less than 10W of RF power.
Here's an example of two attenuators that have a designated input and output:
The following users thanked this post: fnaumann
« Reply #6 on: January 14, 2024, 09:19:02 am »
As you have the analyzer on front of you, you need take closer note of the info on the display.
Input attenuation is by default on Auto and normally 20dB or more depending on input settings.
The analyzer automatically deducts internal attenuation settings from settings to display the correct signal value even when the internal attenuation is set manually.
Add external attenuation and you need change the Reference level to account for it for the measurements to be correct.
Best advise is to have a play with the TG signal which of course you can set the level and then add external attenuation and change the reference level to get a correct result.
It will soon appear to all be quite simple and straightforward to get reliable results.
Once you go to just RF In signals you need be more careful as even a mobile phone nearby handshaking with a cell tower can send input levels beyond safe levels where you will be warned by the internal overload buzzer and OSD warnings. This is especially true when you have the Preamp set to ON which is when you need work with care.
Little steps and have fun getting to know your instrument.
Avid Rabid Hobbyist
Offline blackdog
  • Please stop pushing bullshit...
« Reply #7 on: January 14, 2024, 12:28:36 pm »
And there are other things to consider when using analyzers.
I normally have -6dB or -10dB wide bandwidth attenuators mounted on the two connections of my spectrum analyzer.
This is for two reasons, that is some protection, I almost never measure on high power amplifiers and secondly
to make sure that the connections of the HF analyzer become nicer 50 Ohm.
But then if I do need less attenuation, I adjust the attenuation or remove the extra attenuator, if it cannot harm the instrument.
With HF analyzers, it is better to think first, then think again and can I do what I thought of?
That usually keeps your equipment in one piece.  ;)
Kind regards,
Bram
Necessity is not an established fact, but an interpretation.
« Reply #8 on: February 04, 2024, 11:53:17 pm »
This means that if I want to use my Siglent SSA to measure the RF signals in my apartment (e.g. local WiFi and 5G signals), connecting an antenna directly to the SSA can destroy the spectrum analyzer if a smartphone is nearby? Is a 10 dB attenuator enough to avert this danger?
Offline blackdog
« Reply #9 on: February 05, 2024, 10:36:29 am »
Hi fnaumann,  :)
If a 10dB attenuator is enough to keep the signals you want to see from dropping into noise, then I would use the 10dB attenuator.
I don't know your environment, if you have a hobby room surrounded by your family who each have a cell phone then the power you get in at an unfavorable time can be quite high.
Also if you have radio amateurs nearby the power if you mount an antenna on your SA can be a problem.
Your 10dB attenuator attenuates the power 10x, with a 20dB attenuator it is 100x, but it may be if you were to use a 20dB attenuator that the signals you want to see disappear into the noise.
The point is, that you work "wisely", better to work with -20dB first, see if you can make sensible measurements, too much in the noise, then if you do not see very strong signals, use the -10dB attenuator.
Kind regards,
Bram
« Reply #10 on: February 05, 2024, 08:47:47 pm »
A few iPhones are flying around near me. And the neighbors next to and below me will also use smartphones. No radio amateurs as far I know.
This annoys me as a spectrum analyzer beginner - the thing is apparently so sensitive that you have to handle it with kid gloves. An oscilloscope is a much more robust part, especially if you use a differential probe (and do not need to cover a common ground). I'm honestly disappointed with the SSA use cases shown in YouTube videos. I can do a lot of antenna-related things much more relaxed with my NanoVNA-F V2, and other things with my SDR boxes. Above all, the FM and AM demodulation function on the SSA is a joke (completely unnecessary in my opinion). And then I now have a TinySA Ultra - which in one of the videos even offers a smaller noise floor. I know I can not compare a TinySA Ultra with a Siglent SSA2021X , but ....
Offline blackdog
« Reply #11 on: February 05, 2024, 09:18:11 pm »
Hi fnaumann,
The point is to be careful, my spectrum analyzer has a limit of 100mW at its input.
If you put a pickup antenna next to a mobile phone, you can get over this limit.
The phone transmits not only when you make a call, it does this just when it feels like it. :-)
But don't get paranoid, it's all about being careful with your measuring equipment.
The 10 0f 20dB attenuator at the input you start with during a measurement session normally makes the SA indestructible.
( if you ar not doin measurement on tranmitters!)
The field of an antenna connected to a transmitter decreases quadratically with distance.
So if the distance to a phone is 4-meters, and you double this to 8 meters, then only 1/4 is left.
The manufacturers of this kind of measuring equipment have to perform tricks to get a good noise number, good distortion value and then also have a "decent protected" input, it is not easy.
Greetings,
Bram
Offline EggertEnjoyer123
« Reply #12 on: February 05, 2024, 09:20:31 pm »
You can get a lower noise floor by turning the resolution bandwidth down. (The tradeoff is that it'll take longer to sweep).
For low frequency stuff (like maybe under 50 MHz), a scope is usually better, but once you get to high frequencies scopes become prohibitively expensive. Also, high frequency scopes have 50 ohm inputs too and most of them can take even less power than your spectrum analyzer.
If you want to analyze noise around your house you should look at making a near field probe (There's a EEVBlog video on making your own out of coax). Connecting that directly to the spectrum analyzer should be perfectly fine (as long as you don't put it in the microwave or something).
« Last Edit: February 05, 2024, 09:22:38 pm by EggertEnjoyer123 »
« Reply #13 on: February 06, 2024, 12:00:16 am »
I know this. Watch this starting at minute 16.
« Reply #14 on: February 08, 2024, 11:54:15 pm »
I did my own testing with my SSA2021X and TinySA Ultra using a test signal from my SDG 1032X.
The signals were clearly visible on the TinySA Ultra out of the box. With the SSA2021X, the signal was also visible after various adjustments, but with significantly more noise. The signal/noise ratio seems scary to me as a layman. I tried to set the SSA2021X using various options so that I could see the test signal better on the display. As a test signal I used a simple 10 MHz sine wave signal with 1 to 5 dBm output level (SDG 50 ohm output mode).
To protect the SSA, I installed a 30 dB attenuator (10 W) plus a DC blocker. I reduced the internal attenuation to 0 dB (after the signal-to-noise ratio didn't improve much). Which doesn't make it much better. I also tried different bandwidth settings.
Then when I activated the internal preamp, the SSA beeped and warned of an overload situation. What crap. :-- The SSA is specified as a maximum of 30 dBm and 50 V DC. Of course, these values are only given for special settings. But - I had used a maximum of a 5dBm signal with 30dB of external attenuation, which corresponds to a -25dBm input signal? How can a -25dBm signal produce an overload warning on the SSA? I only activated the PreAmp option under “Amplitude” without setting a manual value. Shouldn't the software prevent overload?
« Last Edit: February 08, 2024, 11:58:44 pm by fnaumann »
« Reply #15 on: February 09, 2024, 12:00:26 am »
To protect the SSA, I installed a 30 dB attenuator (10 W) plus a DC blocker. I reduced the internal attenuation to 0 dB (after the signal-to-noise ratio didn't improve much). Which doesn't make it much better. Then when I activated the internal preamp, the SSA beeped and warned of an overload situation. What crap. The SSA is specified as a maximum of 30 dBm and 50 V DC. Of course, these values are only given for special settings.
BUT - I had used a maximum of a 5dBm signal with 30dB of external attenuation, which corresponds to a -25dBm signal. How can a -25dBm signal produce an overload warning on the SSA? I only activated the PreAmp option under “Amplitude” without setting a manual value. Shouldn't the software prevent overload?
TBH this is incorrect methodology.
Instead you should let the SA manage internal attenuation and add external attenuators when/if necessary.
External attenuation can then be nulled/corrected with use of the amplitude offset feature.
Post a screenshot up and lets compare settings.
Avid Rabid Hobbyist
« Reply #16 on: February 09, 2024, 12:54:23 am »
I assumed that the spectrum analyzer internal attenuation is one of the tunable parameters of a spectrum analyzer, which should be adjusted by the User if necessary, and the external attanuators are intended to protect the spectrum analyzer from power overload (worst case destruction). Since in many cases you don't know exactly how high the signal level to be measured is (dynamic range of the signal), I assumed that you always have to attenuate a little more generously in order to protect the device.
« Last Edit: February 09, 2024, 01:52:49 am by fnaumann »
« Reply #17 on: February 09, 2024, 01:26:10 am »
Certainly so depending on what you are doing however don't underestimate how sensitive a SA is, easily equivalent to any radio.
My first a SSA3032X, scared me regarding this sensitivity and how careless use could set the overload buzzer off so now the cellphone is never nearby when sensitivity is increased or the Preamp is on.
Later I got a SVA1015X and then upgraded to SVA1032X which remains my fav RF instrument where its SA capability is identical to SSA models.
In the early days I was terrified of front end damage yet now on my 3rd none have ever been damaged however anything that might offer a strong signal is first assessed with a simple non-contact RF loop.
Apart from those early days until I found my feet I no longer use external attenuators but have a small selection of N, SMA and BNC types should the need for them arise.
Avid Rabid Hobbyist
« Reply #18 on: February 09, 2024, 01:54:51 am »
A measuring device that declares a max input level of 30 dBm and then issues an overload warning when the signal is -55 dB (E -5) weaker is suspect to me. And if a internal preamplifier activation in a menu can kill the device - such an option should not be able to be activated or should turn off by itself if it get critical. Surely that should be possible with such an expensive measuring device?
And let's assume that I'm too stupid to be able to display the signal sensibly on the SSA2021X , then the automatic function "Auto Tune" - offered in a separate button - should be able to display a measurement signal reasonably sensibly (signal-to-noise ratio) and not as a small bump surrounded by a sea of noise. And why is this all possible out of the box for a device that only costs a fraction?
« Last Edit: February 09, 2024, 02:06:21 am by fnaumann »
« Reply #19 on: February 09, 2024, 02:01:23 am »
Some simple examples with 10dB attenuation on RF In using the default -20dB TG signal which of course will display ~-30dB sweep.
Then with an offset correction added and nothing else changed to now display a -20dB sweep....the TG level and with the 10dB attenuation corrected/nulled.
This is a simple and quick trick using the TG source to check you have set things correctly before proceeding to sniff anything of higher power.
When done this way Markers will display the correct amplitude values without any need for 'in the head' calculations.
Avid Rabid Hobbyist
« Reply #20 on: February 09, 2024, 02:30:49 am »
Before I use newly purchased attenuator, I checked them with the TG. Especially their attenuation over the entire available frequency range. I don't do that before every measurement. And I know the meaning of Ref Offset.
The advantage of fixed attenuators is safety. With variable attenuators, I am always afraid that they will accidently change their attenuator value and then kill the SA during a measurement.
« Reply #21 on: February 09, 2024, 03:35:14 am »
As yet we have not seen any screenshots from you showing settings that might trigger warnings.
FYI max continuous signal is specified as: 30 dBm, 3 minutes, fc ≥10 MHz, att > 20 dBm, preamp off
This is of course without external attenuators.
As yet I prefer to control any instrument rather than let an Autotune process second guess my requirements therefore I have never used it, maybe I should get some experience with it.  :-//
Avid Rabid Hobbyist
The following users thanked this post: Chernobyl
Offline fourfathom
« Reply #22 on: February 09, 2024, 08:01:17 am »
In many cases an "overload" indication doesn't mean a damaging-level input, but just that the input level with the current attenuation and gain settings results in a signal that is causing the detector to clip or distort.  It's not an input power problem, just a dynamic range problem that can be corrected by adjusting the attenuation or gain settings.  No harm done.
We'll search out every place a sick, twisted, solitary misfit might run to! -- I'll start with Radio Shack.
The following users thanked this post: Performa01, pdenisowski
Offline pdenisowski
  • Product Management Engineer, Rohde & Schwarz
    • Test and Measurement Fundamentals Playlist on the R&S YouTube channel
« Reply #23 on: February 09, 2024, 05:02:35 pm »
If I remember correctly, max UE transmit power in LTE and 5GNR is +23 dBm.  I lot of things would have to go wrong (so to speak) at the same time in order to see even +20 dBm at the analyzer input.
And it's not just handshaking (I assume you mean RACH): theoretically the base station could power control the UE up to max transmit power at any time, although this is also very uncommon. I used to do cellular drive test measurements and I'm not sure I ever saw a UE power controlled up to max Tx power.
That said, a nearby UE can absolutely overload the front end of an analyzer and cause distortion / intermod:  I saw that many times while doing interference hunting in the field.  But I've never heard of an analyzer being damaged by OTA signals from a nearby UE.
Damage from overload when directly connected with cable?  Oh yes :)
Test and Measurement Fundamentals video series on the Rohde & Schwarz YouTube channel:  https://www.youtube.com/playlist?list=PLKxVoO5jUTlvsVtDcqrVn0ybqBVlLj2z8
Offline pdenisowski
« Reply #24 on: February 09, 2024, 05:04:00 pm »
Yes, exactly this.  In fact, often I've seen what is clearly distortion caused by an overdriven input even BEFORE any kind of "overload" indication appears.