Author Topic: Teardown / repair: Keithley 619 Dual channel multimeter  (Read 4269 times)
Offline free_electronTopic starter
    • SiliconValleyGarage
« on: October 19, 2014, 07:01:53 pm »
Found this on ebay for a nickle and a dime (well, 80$) : keitlhey 619 dual channel electrometer. I always wanted an electrometer so this was it. Known to broken but at that price worth the gamble.
Front panel
http://www.siliconvalleygarage.com/eevblog/619/619front.jpg
These are older machines but constructed very nice and modular so easy to work on
http://www.siliconvalleygarage.com/eevblog/619/619guts.jpg
GPIb is handled by a motorola 68488 in cerdil package.
http://www.siliconvalleygarage.com/eevblog/619/619gpib.jpg
The brains of the outfit is a 6802 and two 2716 eproms. Four static nibble-wide static rams of the  2114 type (1024 nibbles) make th emain memory.
http://www.siliconvalleygarage.com/eevblog/619/619cpu.jpg
The i/o controller board handles display muxing and keyboard scanning. it also creates an optical isolated serial interface for both channels.
http://www.siliconvalleygarage.com/eevblog/619/619io.jpg
Digital portion of the dual slope a/d convertor i smade using TTL and cms logic Nothing custom here. simple counters do the work.
http://www.siliconvalleygarage.com/eevblog/619/619digital.jpg
The analog portion of the A/d is partially built on teflon standoffs to guarantee low leakage. Das mashinen is nicht fur das gefingerpoken !  in plain english : keep your greasy fingers out !
http://www.siliconvalleygarage.com/eevblog/619/619analog.jpg
Oh joy ! The reference is made using the well known LM199 buried zener with built in thermostat.The metal can AD542 is the input opamp for the a/d convertor.
http://www.siliconvalleygarage.com/eevblog/619/619ref.jpg
Each channel has its own driver board and high voltage supply. they mke +260 and -260 volts for the feedback driver.
http://www.siliconvalleygarage.com/eevblog/619/619driver.jpg
the input boards employ guarding and double shielding techniques .
http://www.siliconvalleygarage.com/eevblog/619/619input.jpg
Teflon stnadoffs galore ! Even the reed relays are special constructions. with separate shielding cans The metal can AD515 electrometer-opamp does the work.
http://www.siliconvalleygarage.com/eevblog/619/619teflon_thumb.jpg[/img]
[url]http://www.siliconvalleygarage.com/eevblog/619/619teflon.jpg]http://www.siliconvalleygarage.com/eevblog/619/619electrometer.jpg[/url
Same warning as witht the A/d convertor : keep your greasy fingers out. This stuff is sensitive they even recommend waiting at least an hour after soldering in this section to allow thermal equalisation...
Backside with the two input channels
http://www.siliconvalleygarage.com/eevblog/619/619twin.jpg
THe machine was dead. at powerup it would show its serial number and firmware revision and then hang. a quick check of the internal supplies showed one at 1 volt that should have been 5 volt. Unplugging the boards fed by that rail ( i suspected a short in one of em ) made that rail shoot to 13 volts. oops. not good !
Turned out the to-3 can regulator for that real was dead. with no load it would simply pass its input voltage. hang 1 kiloohm of it and it collapsed to 1 volt. a quick dig in the scrap bin and i found a donor board that had one.
power up and all was well.
i checked readings against a calibarated 34401 and the machine is well within spec on the ranges i could check.
The following users thanked this post: Echo88, sipo75
Offline grumpydoc
Re: Teardown / repair: Keithley 619 Dual channel multimeter
« Reply #1 on: October 19, 2014, 09:58:06 pm »
Sigh. There are times I really, really wish I lived in the 'states.