First order of business, the fan:
The original fan must have died so someone mickey-moused a replacement on the back, with its own power cord and switch. The impeller is gone and the fan board looks a bit cooked. Someone must have been taking that "Do not remove covers" tag seriously. Wire got broke off the switch and needed to be soldered back on. If I ever get this thing working I'll see about wiring it up inside to the power switch.
I wonder if the fan seized and no one noticed until some parts cooked and died. It appears to have seen some service before...
Tektronix did in fact publish a good general "how to troubleshoot analog oscilloscopes for noobs" guide:
http://www.sphere.bc.ca/test/tek-parts/troubleshooting-scopes.pdfPower rails look okay excepting the -8v power rail was at -6.2v, and now seems to be down to -6v as I measure it again. That's not right. Dug through the service manual for guidance, found something about checking resistor R4202, which looks okay but it is connected to what I assume is supposed to be the -8v power regulator, Q4102 but it seems to only have 1.5v going in and 0.9v coming out. What would it be doing with that low a voltage? The other regulators seem to be for +55v and +15v. And that's all I had time for today.
Still waiting on some scope probes to show up so I can check ripple with the LeCroy. Looks like they just passed through New York last night, so they took the
long way around the globe for some reason...
Spent a lot of time fiddling with the knobs and feeding in test signals. It appears to sweep and trigger just fine. Seems to be some kind of 60hz noise bleeding in to both the horizontal and vertical. What is supposed to be a dot is far from it.
The beam really doesn't want to go the right, except for that one spike. So I'm thinking there's some problem relating to the right deflection plate. Wires look connected to the tube near as I can tell. I suppose physical damage to the tube is possible, but thumping the thing around on the table doesn't seem to effect the trace in any way so I hold out hope that it's electrical...