502AOK some more.
Deoxited the XY switch.. no joy, still requires to be in-between its two stable positions in order to get dots / anything on the screen. That's strange... it must be more complex than meets the eye... but what ?!

Then Deoxited the entire scope. No joy, nada. Zero change whatsoever to the way the scope behaves.
But that's not depressing... it's still progress ! First, now I can rule out dirty contacts... or else I would have had at least SOME kind of change, somewhere, one way or another.... but nope, not a sausage. So that means the issues I have are clearly actual electrical faults, most likely dead tubes all over the place... that purple/gassy 12AT7 was not a good start I guess... probably means there are a bunch more tubes that are cactus.
I guess I could brute force it like some would, and just pull all the tubes one by one to test them... but I don't have a tube tester anyway. Well I have my 575 transistor curve tracer, once it's restored I plan on making a little adapter to enable it to test vacuum tubes. But for now, no way to do mass tube testing...
The better, that means we actually get to use brain cells, look at schematics and probe stuff,... that's much more interesting and entertaining, to me

But before I go probe inside this scope, it first needs a good shower ! It's too dirty / dusty inside, with cog webs to boot, brrrr....
Just a quick shower to get it to a decent state of cleanliness.
I checked the power rails. It's all fine. At first they were all a bit high, like 104V instead of 100V, 370 for 350, and -157V instead of -150V, for the reference voltage. Let it warm up for an hour, voltages decreased slightly, getting closer to spec. Like 369 instead of 371, for example. So I adjusted the -150V reference rail as good as I could (I managed -149.9V, good enough....), and all other voltages where then, logically, to spec. So that's all good.
As for ripple... I was "hoping" one rail would be tragically out of spec, given I heard that cap "vent" yesterday. That would indicate me what cap died.
But.... ripple was well within spec on all rails ?!

But I think I found the bad cap anyway, see pictures. It's not one of the big can caps, no. On the UNDER side of the lower deck, soldered on ceramic strips, I see 4 small axial electrolytic caps, all identical, 6.25uF. They look similar to the bad ones in my HP 120B tube scope.... light grey/silver in colour, black text, transparent plastic wrapping...
I noticed one of them has the positive end looking bad... looks like the red seal is not in good shape eh ?!
The others are only marginally better...
I looked them up on the power supply schematic below, highlighted in yellow. They provide ADDITIONAL filtering for the vertical amplifiers... additional filtering, not primary. Since the ripple upstream is already within spec, these small caps going high ESR / drying out is not a big deal I think, ripple is still within reason.. it's not gonna cause the catastrophic failures and weird and wonderful stuff I see all over the place in this scope...
Worst case, these caps go short-circuit, overload the rail but... in this case I would see a voltage drop, but I don't. I measured the voltage directly on the terminals of all these caps, and it's spot on. So clearly they go high-impedance not low impedance.
So.... as much as it sucks to have these crappy caps dying in there... I highly doubt they are causing my numerous weird problems. Even more so because these caps as I said provide additional filtering for the vertical amplifiers, not for the time base or trigger, which I have problems with as well...
So looks like I have covered the basics then... visual inspection, contact cleaner, power supplies... time to start troubleshooting now ! It's converging toward bad tubes I feel, many of them....
So, I say somehow I AM making progress ?!

3H20AM, going to bed, good night people....