« on: January 29, 2019, 04:21:44 pm »
I would like to buy a new fairly low voltage PSU with high adjustable resolution in the 100uV range or better, any advice?
« Last Edit: January 30, 2019, 02:32:14 am by FriedMule »
Even if I appear online is it not necessary so, my computer is on 24/7 even if I am not on.
Offline David Hess
  • DavidH
Re: What High resolution power supply to buy?
« Reply #1 on: January 29, 2019, 08:49:15 pm »
Even source meters from companies like Keithley do not have that kind of performance and they are expensive.  Did you have a budget in mind?
Offline colorado.rob
« Reply #2 on: January 29, 2019, 09:48:28 pm »
Keysight B2912A will do the business.
Offline DaJMasta
    • medpants.com
« Reply #3 on: January 29, 2019, 10:03:47 pm »
Definitely sourcemeter territory, the newer Keithley 2450 has a 500nV source resolution, but the noise figure is big for that setting level.  Also that's only on the lowest range which tops out at 20mV.
Are you sure you need nV accuracy in the source?  It's easy to get a meter with that kind of resolution, but getting a constant voltage source of that caliber is tricky.  Easier to do a precision current source (the same Keithley meter mentioned earlier can set currents with a 500fA resolution, 10^6 times finer steps).
« Reply #4 on: January 30, 2019, 02:34:18 am »
Wary sorry, did hit N where I wanted to hit U as in micro instead of wrongly nano or better.