The specs say: <-145dBm, typ. -150dBm for >10MHz center frequency, 10Hz RBW, 1Hz VBW and 0dB attenuation.
You still have 10dB attenuation, so without that the displayed noise floor would be somewhere near -150dBm. Since you’re using 1Hz RBW, we’d expect <-155dBm - maybe there’s still some other setting left to optimize.
The 1st order dynamic range is specified as 165dB for 1Hz RBW and the 1dB compression point is +10dBm (which is 10~15dB higher than the usual candidates and this is what makes the FSE analyzers so great), which confirms that the noise floor should indeed not be higher than -155dBm in this scenario.
But even if your particular unit doesn’t quite meet its specs and the noise floor is some 5dB higher than expected and the 1st order dynamic range is “only” 160dB – don’t worry. There are still not many other analyzers that could compete.
The usual Chinese budget SA have some -5dB CP – and consequently would have to display a noise floor of -165dBm/Hz (or -170dBm/Hz to meet the R&S specs for the 1st order dynamic range) with preamp off.
Not to mention the FSEA30 phenomenal 115dB 3rd order dynamic range (above 50MHz), which is the one that really counts. And this doesn’t require the lowest possible noise floor, but exceptional linearity in the entire signal path up to the point, where the narrow band filtering occurs, which for narrow RBWs is the digital processing after the ADC in the final IF…