Just in case anyone is wondering about the claimed measurement accuracy of thermal imaging cameras that were used during the Covid-19 pandemic…..
Thermal imaging was seen as a fast and effective way to check individuals or groups of humans for evidence of fever (elevated body temperature). Whilst this whole approach was deeply flawed and not particularly reliable, efforts were made to improve the accuracy of thermal imaging systems to make their measurement of humans more accurate. I will not detail the challenges of estimating a humans core temperature using extremities and “hot spots” on the face (tear duct region of eyes) but needless to say, the thermal imaging of humans approach faced serious challenges if it were to be effective.
In order to improve upon the nominal +/-2 Celsius or +/-2% (whichever is greater) measurement tolerance of a good quality thermal imaging camera, manufacturers had to significantly improve their camera calibration routines, camera temperature stability and FFC event accuracy. Even after making these improvements, the measurement tolerance of even the better human temperature measurement thermal imaging cameras was +/-0.5 Celsius. The Chinese Government had set a tolerance target of just +/-0.3 Celsius for their domestic thermal camera manufacturers to meet. In order to improve the measurement accuracy to that required by the Government, manufacturers introduced Blackbody thermal references to their human temperature measurement solutions. The temperature and Emissivity of the Blackbody emission plate was known and the Blackbody was positioned in the cameras field of view. The required measurement corrections could then be calculated in software to ensure the best possible accuracy of thermal camera measurement. There will always be an error introduced by the varying Emissivity of the target but the +/-0.3 Celsius measurement tolerance was achieved.
Before anyone gets too excited about this. The cameras that were adapted from general use models to accurate human temperature measurement models were modified in both hardware, firmware and calibration. To achieve the +/-0.5 Celsius measurement accuracy, most used improved FFC hardware/routines to better self calibrate the camera during use. The calibration at the factory was greatly enhanced with more reference points over a narrow range of temperatures. A human temperature measurement camera usually has a restricted measurement range of around 20 Celsius to 45 Celsius over which its specifications for measurement accuracy are valid. The early attempts to improve the accuracy of standard general purpose thermal cameras failed miserably and were rejected by the Chinese Government. The models that actually passed the Government testing were greatly enhanced over the standard general purpose models. This is one of the reasons why these human temperature measurement cameras cost significantly more than general purpose models (there were other, profit orientated, reasons of course!) Some thermal camera manufacturers tried to “cash-in” on the Pandemic by releasing general purpose thermal cameras that used a Blackbody” reference source to improve measurement accuracy via software. Some of those cameras will pass the +/-0.3 Celsius measurement tolerance specification, whilst others will not. YMMV.
So in Précis, do not expect the measurement accuracy of a high quality human temperature measurement thermal camera from a general purpose camera, even if it’s menus include a human temperature measurement mode ! Read the manufacturers measurement accuracy specification and work with that when using the camera.
Fraser