So, I finally got around to doing a thing! As stated earlier (I think), I've noticed my THOR 002 V1.1.2 is reading way off (high) - like, my bedspread was reading 80F in equilibrium with the rest of my room, where a thermometer read 71F. TL;DR, I did a calibrated measurement and, with the measurement parameters all set correctly, the camera does in fact read within 1C. Mine came with an implausibly low ambient temperature setting from the factory.
I had a project from a while back that's a cheap implementation of a blackbody temperature target. It's a PCB with a big thick pad of copper, a ring of resistors on the edge well thermally coupled to that pad, and two TMP117 high-accuracy PCB thermometers, one well-coupled to the center of the pad and one on the corner. It's also got an apparent surface of calibrated/known emissivity - a patch of Acktar Metal Velvet, which specifies e>0.985.

A microcontroller runs a PID controller to drive the temperature at the center of the target using the resistors. The TMP117 has a datasheet accuracy of 0.1C, plus whatever delta there is across the coupling vias (negligible), plus the controller error (known).
As measured, the controller held the target center to 50C within 0.2C, and the corner is pretty consistently 0.7C above that.

Initially, I noticed my THOR002 was only a little off, closer than I'm used to, but I knew I needed to set the emissivity correct for a good test anyway. While browsing Settings>Measurement Parameters to do that, I noticed that the setting for ambient temperature was set implausibly low - something like -8C. I took the opportunity to set that correctly for the current room temp. I couldn't key in 0.985 for emissivity - I had to select the closest stock option of "Skin" at 0.98. The "distance" setting also doesn't go any lower than 0.5m, so I tried to take my measurement from about that distance.
Lo and behold, with settings all correct, the measurement was just about spot on: 49.8/50.0 on center, and 50.3/50.8ish on the corner.

My Thermal Master P3, with all settings dialed in, was also basically bang-on: 50.4/50, 50.8/50.8.

A few minutes later, I'm now measuring 27.4 when the target is bang-on 30C, so perhaps there is still some room for improvement, but again it's technically just about within the error bars.