Point taken.
I have just purchased the cheap chinese LM399 in the plastic case. Realistically, this will most likely more than challenge the 4.5 digit meter that I will end up with. (4.5D seems to be the upper limit for compact portable meters.)
My biggest problem is calibration.
The seller claims 10v and Vz are checked with a Keithley 7510. We will see.
enut11
FYI, as a comparison of the above mentioned references.
I have the 10V out of my CircuitValley ADR1399 ref connected to the front inputs on my Keysight 34465A (last cal in 2015 at factory, 9.999,985,9) and my more trusted 34401A (Agilent version with round buttons). The Emeco LM399 ref 10V output is connected to the rear inputs.
The CV ADR1399 ref reads 10.000,274 on the 34465A and 10.000,16 on the 34401A. On my older Agilent 34401As it reads 10.000,86 and 10.000,33. The 34465A using stats, 10V range, and 10,000+ readings at 10PLC, std.dev 0.000,002.
The Emeco LM399 ref reads 10.000,111 on the 34465A, and 10.000,01 on the 34401A. On my older Agilent 34401A's it reads 10.000,69 and 10.000,16. The 34465A using stats, 10V range, and 700+ readings at 10PLC, std.dev 0.000,004.
Note the consistent approx 160uV difference between the two references across all four meters, so I don't think that is due to Front/rear inputs or meters. Essentially it appears that one of the 34401A's is out of 1-year spec, but the other are all likely good. My total uncertainty with these measurements is probably too high to be sure, but it gives me some confidence for those three other meters. That one with a reading much higher is my only SG manufactured one (approx 2005), all the others are MY manufactured (approx 2014, 2007, 2001).
This was just a quick check I did this morning. I am just using common banana plug leads, daisy chained between meters, not low EMF cabling although I have bought some to (eventually) set up some longer term logging. The weather yesterday and today have been hotter than usual, but all of the meters have been powered on constantly for months. I need to check my notes if I can find them, but I seem to recall that these readings match quite well with those from about 6 months ago.
Both references were bought in December 2024, and have been powered on almost continuously since then. The CV ADR1399 was one of the early batch, and it didn't get a label of its initial 10V reading (only Vz). I wasn't too worried as I had planned to leave it to age and eventually get a reading on a local 7.5+d calibrated meter. The Emeco LM399 10V output was labelled as 9.99962 using a DMM7510, but it likely has drifted somewhat in nearly 2 years (no idea how much burn in they do).
BTW, if you are passing through Sydney at some point in your travels I'd be happy to do a comparison measurement for you.