Author Topic: Multimeter spreadsheet (80 manufacturers and over 400 meters listed)  (Read 836941 times)

bkmit and 3 Guests are viewing this topic.

Online rsjsouza

  • Super Contributor
  • ***
  • Posts: 6002
  • Country: us
  • Eternally curious
    • Vbe - vídeo blog eletrônico
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #275 on: November 28, 2015, 12:17:59 pm »
Wytnucls, any reason (other than age, perhaps) the Keithley 191 is out of the list? I think it is a great meter on paper for a dirt cheap price. The reason is that I was looking for one and found it missing from the table.

BTW, regarding free/non-free viewers, Openoffice (free office package) opens the spreadsheet just fine.

Also, as I said before, great work on the spreadsheet!

(edit) added 191's specsheet
« Last Edit: November 28, 2015, 12:29:29 pm by rsjsouza »
Vbe - vídeo blog eletrônico http://videos.vbeletronico.com

Oh, the "whys" of the datasheets... The information is there not to be an axiomatic truth, but instead each speck of data must be slowly inhaled while carefully performing a deep search inside oneself to find the true metaphysical sense...
 

Offline WytnuclsTopic starter

  • Super Contributor
  • ***
  • Posts: 3045
  • Country: be
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #276 on: November 28, 2015, 02:03:55 pm »
No particular reason, except that I wasn't familiar with it. It will be listed at the next revision.

« Last Edit: November 28, 2015, 02:07:43 pm by Wytnucls »
 

Offline ez24

  • Super Contributor
  • ***
  • Posts: 3082
  • Country: us
  • L.D.A.
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #277 on: November 28, 2015, 07:03:25 pm »
No particular reason, except that I wasn't familiar with it. It will be listed at the next revision.



I have been looking at this one.  Why does this one show a reading with nothing connected to it?  Is it bad?  Several on ebay also show readings of some sort with no wires connected?  thanks
YouTube and Website Electronic Resources ------>  https://www.eevblog.com/forum/other-blog-specific/a/msg1341166/#msg1341166
 

Offline allikat

  • Contributor
  • Posts: 30
  • Country: gb
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #278 on: December 04, 2015, 06:27:37 am »
Meters often show some slight measurements when open circuit. It's just electrical noise and induced voltages as far as I know. Those will often show up, but it will settle down as soon as you put an actual circuit on the end of the probes.
Any engineer can readily identify 3 smells:
1: Coffee, 2: Escaped magic smoke, 3: Bullshit
(from an original post by John Coloccia)
 

Online rsjsouza

  • Super Contributor
  • ***
  • Posts: 6002
  • Country: us
  • Eternally curious
    • Vbe - vídeo blog eletrônico
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #279 on: December 04, 2015, 12:15:32 pm »
No particular reason, except that I wasn't familiar with it. It will be listed at the next revision.
U da man. Thanks!
Vbe - vídeo blog eletrônico http://videos.vbeletronico.com

Oh, the "whys" of the datasheets... The information is there not to be an axiomatic truth, but instead each speck of data must be slowly inhaled while carefully performing a deep search inside oneself to find the true metaphysical sense...
 

Offline FireFlower

  • Regular Contributor
  • *
  • Posts: 130
  • Country: fi
  • Wish had a cabin in Lappland
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #280 on: December 05, 2015, 12:09:11 am »
What is with that updated spreadsheet download with EEVBlog DM1 branded multimeter?

Google doesn't return any results for it.
 

Offline tautech

  • Super Contributor
  • ***
  • Posts: 28613
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #281 on: January 01, 2016, 09:23:55 pm »
Wytnucls, has there been any further developments of the OWON DMM's?
You stated you added them to your spreadsheet: https://www.eevblog.com/forum/testgear/need-advise-for-owon-multimeter/msg675501/#msg675501

Latest in the OP does not show them or am I blind?
Avid Rabid Hobbyist.
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 

Offline WytnuclsTopic starter

  • Super Contributor
  • ***
  • Posts: 3045
  • Country: be
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #282 on: January 02, 2016, 04:14:48 am »
Owon B35T is listed in the downloadable spreadsheet. The image on post one hasn't been updated in a while.
« Last Edit: January 02, 2016, 04:17:42 am by Wytnucls »
 

Offline kevbry

  • Contributor
  • Posts: 19
  • Country: ca
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #283 on: January 02, 2016, 04:41:20 am »
What is with that updated spreadsheet download with EEVBlog DM1 branded multimeter?

Google doesn't return any results for it.

Any news on this? I found a post from November where Dave stated that a rebadged Brymen would be out "this month", but no updates since.
Edit: Found another post stating that the Brymen collaboration would be coming out in January. Looking forward to it, specs in the spreadsheet looked nice.
« Last Edit: January 02, 2016, 04:58:34 am by kevbry »
 

Offline WytnuclsTopic starter

  • Super Contributor
  • ***
  • Posts: 3045
  • Country: be
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #284 on: January 02, 2016, 05:32:59 am »
Tentative specs. To be updated once meter is released.
 

Offline fanOfeeDIY

  • Supporter
  • ****
  • Posts: 412
  • Country: jp
    • YouTube Channel
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #285 on: January 03, 2016, 03:34:41 pm »
Hi,
I have HIOKI DT4256 and I would like to contribute to the chart.

This is the list I could investigated so far.

Count
  6000
AC+
  Yes
BW
  1k
DUT
 (i am not sure about the meaning)
AC volt
  100uV
DC Accuracy
  30/-/3
DC amp min
  10nA
DC amp max
  10A
bw
 (i am not sure about the meaning)
m?
  100mohm
nS/M?
  60M
Cnt
  very fast
pF
  x (1nF min)
mF
  10mF
Dio
  5
?Hz
  100k
%
  yes (DT4256 has it)
W
  x

  x
PC
  yes
Kit
  option
Int Log
  x
Clock
  x
Disp
  2
Light
  light blue
'I'''I
 (i am not sure about the meaning)
'M/m
  yes
Peak
  x
Hld
  yes (it has auto hold, identical to Flue touch hold)
dB
  x
LoZ
  x
VFD
 (i am not sure about the meaning)
Evt
  x
Batt
  4AAA
Life
  130hour
F G
  x
E Pwr
  x
Jack
 (i am not sure about the meaning)
Fuse
  1000
CAT
  III 1000>
P/F
  x
4-20
  (i am not sure about the meaning)
NCV
  x
UL
  ?
Price
  $200


I could not figure out some of the abbreviation.
There might be some mistakes or misunderstanding, so I welcome any kind of comments or questions.

 

Offline ez24

  • Super Contributor
  • ***
  • Posts: 3082
  • Country: us
  • L.D.A.
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #286 on: January 04, 2016, 05:55:01 am »
4-20
  (i am not sure about the meaning)
--------------
Reads 4ma to 20ma as a percentage

at 4ma meter reads 0 %

at 20ma meter reads 100%

at 12ma meter reads  50%

etc  nice feature
YouTube and Website Electronic Resources ------>  https://www.eevblog.com/forum/other-blog-specific/a/msg1341166/#msg1341166
 

Offline WytnuclsTopic starter

  • Super Contributor
  • ***
  • Posts: 3045
  • Country: be
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #287 on: January 04, 2016, 07:40:01 am »
The new Hioki DT4256 will be included in the next revision.
No AC+DC, 1mV AC (100uV DC) and 10uA DC (100uA AC) resolution only, no logic level frequency %, but has non contact high voltage detection.
 

Offline fanOfeeDIY

  • Supporter
  • ****
  • Posts: 412
  • Country: jp
    • YouTube Channel
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #288 on: January 04, 2016, 10:35:36 am »
Thank you ez24,

I found that all the abbreviations were addressed as comments on the excel file.

Hi Wytnucls,

Would like to correct some mistakes on my previous HIOKI DT4256 post.

BW
  1kHz

Volt
  1mV

DC uA
  10uA

A
  10A

bw
  1k Hz (also DT4252 and DT4253 are the same)

'I'''I
  Yes, it has bar graph.

m/M
  It has average.

VFD
  It has low pass filter, selectable on 100Hz and 500Hz.

NCV
  Selectable between, AC40V-AC600V or AC60V-AC600V.

4-20
  Yes (on DT4256 and 4253, and not on 4252)
 

Offline fanOfeeDIY

  • Supporter
  • ****
  • Posts: 412
  • Country: jp
    • YouTube Channel
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #289 on: January 04, 2016, 10:43:20 am »
This is about Yokogawa.

All TY720, TY710, TY530, TY520 have:

Jack
  Physical port shutter.
 

Offline fanOfeeDIY

  • Supporter
  • ****
  • Posts: 412
  • Country: jp
    • YouTube Channel
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #290 on: January 04, 2016, 11:17:21 am »
I found that

UL
   All TY720, TY710, TY530, TY520 are UL listed.
 

Offline fanOfeeDIY

  • Supporter
  • ****
  • Posts: 412
  • Country: jp
    • YouTube Channel
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #291 on: January 04, 2016, 11:35:26 am »
This is on Yokogawa.

VFD
  On TY720, TY530, TY520 but not on TY710.
 

Offline fanOfeeDIY

  • Supporter
  • ****
  • Posts: 412
  • Country: jp
    • YouTube Channel
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #292 on: January 04, 2016, 11:48:12 am »
LoZ
  TY720 has it. Other TY710, TY530, TY520 don't.
 

Offline fanOfeeDIY

  • Supporter
  • ****
  • Posts: 412
  • Country: jp
    • YouTube Channel
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #293 on: January 04, 2016, 11:51:11 am »
This is for HIOKI.

Accuracy
  DT4252 and DT4253 -> 30/-/5
  DT4256 -> 30/-/3
 

Offline WytnuclsTopic starter

  • Super Contributor
  • ***
  • Posts: 3045
  • Country: be
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #294 on: January 06, 2016, 06:45:12 am »
This is on Yokogawa.

VFD
  On TY720, TY530, TY520 but not on TY710.
Not according to the Yokogawa manuals. Quote a reference with your corrections.
 

Offline WytnuclsTopic starter

  • Super Contributor
  • ***
  • Posts: 3045
  • Country: be
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #295 on: January 06, 2016, 06:46:55 am »
LoZ
  TY720 has it. Other TY710, TY530, TY520 don't.
Again, no LoZ on TY720. Read the manual.
 

Offline WytnuclsTopic starter

  • Super Contributor
  • ***
  • Posts: 3045
  • Country: be
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #296 on: January 06, 2016, 06:54:35 am »
I found that

UL
   All TY720, TY710, TY530, TY520 are UL listed.
530, 520 are not UL listed.

http://tmi.yokogawa.com/files/uploaded/IMTY530_S01_010.pdf
 

Offline WytnuclsTopic starter

  • Super Contributor
  • ***
  • Posts: 3045
  • Country: be
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #297 on: January 06, 2016, 07:03:02 am »
This is for HIOKI.

Accuracy
  DT4252 and DT4253 -> 30/-/5
  DT4256 -> 30/-/3
4252 600mV 20/-/5
4253 600mV 50/-/5
 

Offline fanOfeeDIY

  • Supporter
  • ****
  • Posts: 412
  • Country: jp
    • YouTube Channel
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #298 on: January 07, 2016, 12:28:15 pm »
This is on Yokogawa.

VFD
  On TY720, TY530, TY520 but not on TY710.
Not according to the Yokogawa manuals. Quote a reference with your corrections.

Hi,

I did refer to original Japanese manuals and materials, since I am Japanese and could read in Japanese.
Many Japanese companies are not so good at making English web-sites in general.
Iwatsu and Leader do not look good compare to Japanese web-sites, but even the Japanese one is not so great.
Hioki andYokogawa are mostly OK but I see some places it could have been improved.

This is community forum and I just wanted to help improving the chart.
I thought having many people from the world would benefit on this forum and also good place to put the knowledge together.
I really did not have any other intention.

I own both TY530 and DT4256.

This is the manual about low pass filter.




 

Offline fanOfeeDIY

  • Supporter
  • ****
  • Posts: 412
  • Country: jp
    • YouTube Channel
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #299 on: January 07, 2016, 12:31:24 pm »
I found that

UL
   All TY720, TY710, TY530, TY520 are UL listed.
530, 520 are not UL listed.

http://tmi.yokogawa.com/files/uploaded/IMTY530_S01_010.pdf

English web site is not always up to date.
I thought it is good to verify who could read the original Japanese.

This is dated on 2011.
http://tmi.yokogawa.com/files/uploaded/IMTY530_S01_010.pdf

This is the official UL list from the UL web-site which specifies about TY700 and TY500 series.

http://database.ul.com/cgi-bin/XYV/template/LISEXT/1FRAME/showpage.html?name=PICQ.E214270&ccnshorttitle=Measuring,+Testing+and+Signal-generation+Equipment&objid=1075210338&cfgid=1073741824&version=versionless&parent_id=1073991173&sequence=1

This is the pictures regarding UL.




« Last Edit: January 07, 2016, 02:02:39 pm by fanOfeeDIY »
 


Share me