As a last step, therefore, I only ask you to confirm me that you are reasonably convinced that my unit is affected by a malfunction.

I'd say your unit is malfunctioning compared to mine and I would not be satisfied with that.  I would be interested to see what that noise looks like.  Try setting it up so that you have the fat trace and then speed up the timebase to 5nS/div, then do a single shot.  Also do the single shot in dots mode.

But that's just curiosity.  I think your scope is defective.
A 3.5 digit 4.5 digit 5 digit 5.5 digit 6.5 digit 7.5 digit DMM is good enough for most people.
 

Offline ExaLab

  • Regular Contributor
  • *
  • Posts: 86
  • Country: it
Re: Siglent Technical Support join in EEVblog
« Reply #1451 on: December 30, 2020, 07:24:32 pm »

This scope have unknown problem. Also it looks like problem is not in DC offset system. Because it disappear when ADC switch to interleaved mode.  (in 4 channel mode ADCs run in 500MSa/s mode and when there is in this case CH1 and CH4 these both ADC run 1GSa/s mode).  Why it affect so much when it change this ADC working mode is unknown to me, of course can speculate this and that but it do not solve this problem at all.

I agree, very strange behavior... the noise increases linearly with the DC offset (this is a fact).
It could be as follows: the DC offset reference is affected by a noise that is correlated (i.e. synchronous) to the 1 GSa/s and somehow incorrelated (i.e. asynchronous) to the 500MS/s.
If so, the superimposed noise would only be evident in the second case.

Anyway, in the absence of further developments, I'll proceed with the return of the instrument!

What I find a bit strange is the fact that although I sent a detailed report to Siglent's customer support, until now they were unable to give me any valid solution to this issue. I hope it ends well!
 

Offline ExaLab

  • Regular Contributor
  • *
  • Posts: 86
  • Country: it
Re: Siglent Technical Support join in EEVblog
« Reply #1452 on: December 31, 2020, 02:15:13 am »
On the image 2 vertical pixel is one ADC step.
Many times noise affect so that it change randomly between two steps some times 3 steps when we are in this voltage level.
If you turn it to faster time scale and set diplay mode dots you see these real ADC samples. This is normal for this kind of ADC and front end also.

Tomorrow, I'll take some acquisitions in order to measure the real amount of the noise. Just to avoid misunderstandings, when you speek about 1 step you refer to two distinct levels whose difference is 1LSB. So, with two steps you refer to three distinct levels whose total width is 2LSB and so on...   Is it correct?
 

Offline Bicurico

  • Super Contributor
  • ***
  • Posts: 1722
  • Country: pt
    • VMA's Satellite Blog
Re: Siglent Technical Support join in EEVblog
« Reply #1453 on: December 31, 2020, 09:15:26 am »
Hi.

Due to the fact that mu SSA3021X got bricked recently, I had to open it in order to get access to the SD card reader. In the process I had to cut the warranty sticker - at the time I was so upset with the fact my device was not working I simply did not have the patience to apply Dave's technique of sticker removal.

My device is out of warranty for quite some while, so the broken sticker doesn't matter. But somehow my ADS kicks in and it deeply annoys me that the sticker is not in it's virgin state.

Hence I would like to know if someone could perhaps provide me with a new sticker. Of course I pay whatever reasonable costs for sending me the sticker.

Again, there is no other reason than having everything as new. My device still has the screen sticker, too. Also, now harm done warranty-wise, either, as it expired.

Thanks,
Vitor

Offline tautech

  • Super Contributor
  • ***
  • Posts: 28608
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Siglent Technical Support join in EEVblog
« Reply #1454 on: December 31, 2020, 09:41:14 am »
Good luck getting Siglent warranty stickers. Wanted some for years for the very few warranty repairs I do but asked several times and none were ever supplied. So now I just make sure my paperwork records are perfect should they ever be needed.

And.....pull that screen protector off !  :horse:
Avid Rabid Hobbyist.
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 

Offline Bicurico

  • Super Contributor
  • ***
  • Posts: 1722
  • Country: pt
    • VMA's Satellite Blog
Re: Siglent Technical Support join in EEVblog
« Reply #1455 on: December 31, 2020, 09:49:50 am »
Nope! The screen protector stays on!   :)

Well, asking doesn't cost money so I had some hopes.

A good after sales support is important and details like stickers are not to be neglected.

If Siglent wants to compete with the big ones they should consider calibration stickers, too.

Offline tautech

  • Super Contributor
  • ***
  • Posts: 28608
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Siglent Technical Support join in EEVblog
« Reply #1456 on: December 31, 2020, 09:56:43 am »
If Siglent wants to compete with the big ones they should consider calibration stickers, too.
Products now come with proper Cal certs.
Avid Rabid Hobbyist.
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 

Offline Bicurico

  • Super Contributor
  • ***
  • Posts: 1722
  • Country: pt
    • VMA's Satellite Blog
Re: Siglent Technical Support join in EEVblog
« Reply #1457 on: December 31, 2020, 09:58:39 am »
Yes. But they certainly have an expiry date. Are Siglent distributors able to perform calibration services and issue new certificates?

Offline tautech

  • Super Contributor
  • ***
  • Posts: 28608
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Siglent Technical Support join in EEVblog
« Reply #1458 on: December 31, 2020, 10:04:22 am »
Are Siglent distributors able to perform calibration services and issue new certificates?
If they have a NIST certified Cal lab....I don't.

Something best left to the professional Cal labs......however most service manuals list a performance verification procedure so for those that have calibrated equipment they can check any instrument still performs to original specification.
Avid Rabid Hobbyist.
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 

Offline Bicurico

  • Super Contributor
  • ***
  • Posts: 1722
  • Country: pt
    • VMA's Satellite Blog
Re: Siglent Technical Support join in EEVblog
« Reply #1459 on: December 31, 2020, 10:15:15 am »
 :-+
Thank you.

Offline ebastler

  • Super Contributor
  • ***
  • Posts: 6637
  • Country: de
Re: Siglent Technical Support join in EEVblog
« Reply #1460 on: December 31, 2020, 10:40:22 am »
Good luck getting Siglent warranty stickers.

Maybe there's an opportunity for a gray market here?  ;)
Owners of out-of-warranty Siglent instruments could carefully remove and resell their intact stickers. ::)
 
The following users thanked this post: tautech

Offline tautech

  • Super Contributor
  • ***
  • Posts: 28608
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Siglent Technical Support join in EEVblog
« Reply #1461 on: December 31, 2020, 11:11:00 am »
Happy 2021 to all.
12.08am Jan 1 2021 downunder in NZ.
Avid Rabid Hobbyist.
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 
The following users thanked this post: Performa01, Martin72

Offline Bicurico

  • Super Contributor
  • ***
  • Posts: 1722
  • Country: pt
    • VMA's Satellite Blog
Re: Siglent Technical Support join in EEVblog
« Reply #1462 on: December 31, 2020, 11:39:46 am »
Good luck getting Siglent warranty stickers.

Maybe there's an opportunity for a gray market here?  ;)
Owners of out-of-warranty Siglent instruments could carefully remove and resell their intact stickers. ::)

I pay 5 Euro for your intact sticker! :)

Of course the sticker is not neededk, especially after warranty runs out. But like when I put my cups in the cupboard, all symetrically arranged, I do like to have all stickers intact and my devices in prestine condition.

It's just one of many oddities of my self.

Happy 2021 everyone!

Offline KVM2007

  • Contributor
  • Posts: 10
  • Country: ge
Re: Siglent Technical Support join in EEVblog
« Reply #1463 on: December 31, 2020, 12:55:08 pm »
Quote
I pay 5 Euro for your intact sticker! :)
...
Dave's technique of sticker removal.
This Dave's technique?
 
I have a new Siglent with sticker, and I have no warranty. But I don't like Dave's technique  :palm:
If it's a David Copperfield, it doesn't work for 5 Euros
 

Offline Martin72

  • Super Contributor
  • ***
  • Posts: 5982
  • Country: de
  • Testfield Technician
Re: Siglent Technical Support join in EEVblog
« Reply #1464 on: December 31, 2020, 01:24:51 pm »
Best way to remove a sticker complete is to heat it up with a fan - but not too hot, of course.


Offline rf-loop

  • Super Contributor
  • ***
  • Posts: 4126
  • Country: fi
  • Born in Finland with DLL21 in hand
Re: Siglent Technical Support join in EEVblog
« Reply #1465 on: December 31, 2020, 03:08:59 pm »
On the image 2 vertical pixel is one ADC step.
Many times noise affect so that it change randomly between two steps some times 3 steps when we are in this voltage level.
If you turn it to faster time scale and set diplay mode dots you see these real ADC samples. This is normal for this kind of ADC and front end also.

Tomorrow, I'll take some acquisitions in order to measure the real amount of the noise. Just to avoid misunderstandings, when you speek about 1 step you refer to two distinct levels whose difference is 1LSB. So, with two steps you refer to three distinct levels whose total width is 2LSB and so on...   Is it correct?

Just this is enough reason for ask change or just return. Why you do not send it back to seller, he can talk with Siglent what to do. This is not some perhaps problem. This image tell enough. This is failure. Period.  ;)

Do not wait this case go old..  this is clear FOA case (Failed on Arrival)
I drive a LEC (low el. consumption) BEV car. Smoke exhaust pipes - go to museum. In Finland quite all electric power is made using nuclear, wind, solar and water.

Wises must compel the mad barbarians to stop their crimes against humanity. Where have the wises gone?
 

Offline KVM2007

  • Contributor
  • Posts: 10
  • Country: ge
Re: Siglent Technical Support join in EEVblog
« Reply #1466 on: December 31, 2020, 04:55:27 pm »
Best way to remove a sticker complete is to heat it up with a fan - but not too hot, of course.
All the world's experience on the EEVblog
https://www.eevblog.com/forum/blog/eevblog-799-how-to-remove-warranty-void-security-stickers/  :)

Last post - super!  :-DD
Quote
In an amusing bit of irony, several companies in the United States are being urged by the Federal Trade Commission to remove the “Warranty Void if Removed” stickers and scrap other onerous policies that appear on their products because they are unenforceable garbage.
 

Offline ExaLab

  • Regular Contributor
  • *
  • Posts: 86
  • Country: it
Re: Siglent Technical Support join in EEVblog
« Reply #1467 on: January 20, 2021, 10:51:11 pm »
On the image 2 vertical pixel is one ADC step.
Many times noise affect so that it change randomly between two steps some times 3 steps when we are in this voltage level.
If you turn it to faster time scale and set diplay mode dots you see these real ADC samples. This is normal for this kind of ADC and front end also.

Tomorrow, I'll take some acquisitions in order to measure the real amount of the noise. Just to avoid misunderstandings, when you speek about 1 step you refer to two distinct levels whose difference is 1LSB. So, with two steps you refer to three distinct levels whose total width is 2LSB and so on...   Is it correct?

Just this is enough reason for ask change or just return. Why you do not send it back to seller, he can talk with Siglent what to do. This is not some perhaps problem. This image tell enough. This is failure. Period.  ;)

Do not wait this case go old..  this is clear FOA case (Failed on Arrival)


Thank you so much for your valuable support!

As anticipated, I did some acquisitions on the amounth of noise of channel two in the worst conditions: it was between six and seven bits (as you predicted...)
More, the nature of this noise seems to be correlated to the sampling frequency, in the sense that it occours with an apparent frequency of 250 MHz...

Anyway, I have returned the product and am waiting for the seller to send me a new one... I should receive it by the end of the month.
I'll let you know how it ends.

In the meantime, thanks again!!

 

Offline CharlotteSwiss

  • Frequent Contributor
  • **
  • Posts: 792
  • Country: ch
Re: Siglent Technical Support join in EEVblog
« Reply #1468 on: January 26, 2021, 12:32:18 am »
Siglent sds1000 (my is 1202x-e)
The oscilloscope crashes many times when press print key  :--
I tried at least 10 sticks, even small ones, always the same story.
I read that someone returning the penultimate firmware had no more problems.
Looking at the firmware updates, I saw that the creation (unnecessary) of the PNG folder has been implemented.
From tests done, the bug happens much less (but it happens the same anyway) if when you save the png files on the pc, then you don't delete the PNG folder from the stick.
However, I am convinced that if the Siglent engineers wanted to review the updates from the penultimate to the latest firmware (in png files, the Print button, the USB key) they could provide customers with a bug-free firmware!
thanks
 :popcorn:
 

Offline blurpy

  • Regular Contributor
  • *
  • Posts: 233
  • Country: no
Re: Siglent Technical Support join in EEVblog
« Reply #1469 on: February 06, 2021, 03:03:04 pm »
Is it safe for the Siglent instruments to connect a ground strap to them (to ground me) or do I need something else to connect the strap to?

I have a SDG2042X with an earth terminal in the back and a SPD3303X-E with a ground terminal in the front. Since the ground terminal in the front of the PSU is so accessible it seems like the best option for me to connect the ground strap, but I don't know if I can damage something that way? Is it the same with the function generator? The strap has a 1M resistor.
 

Offline tautech

  • Super Contributor
  • ***
  • Posts: 28608
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Siglent Technical Support join in EEVblog
« Reply #1470 on: February 06, 2021, 08:34:31 pm »
Is it safe for the Siglent instruments to connect a ground strap to them (to ground me) or do I need something else to connect the strap to?

I have a SDG2042X with an earth terminal in the back and a SPD3303X-E with a ground terminal in the front. Since the ground terminal in the front of the PSU is so accessible it seems like the best option for me to connect the ground strap, but I don't know if I can damage something that way? Is it the same with the function generator? The strap has a 1M resistor.
If you can trust the integrity/safety of you mains supply earth I see no problems with what you propose.
However if a fault occurs somewhere and your mains earth 0V becomes elevated your ground strap will be too.

Be careful.
Avid Rabid Hobbyist.
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 

Offline blurpy

  • Regular Contributor
  • *
  • Posts: 233
  • Country: no
Re: Siglent Technical Support join in EEVblog
« Reply #1471 on: February 07, 2021, 11:39:17 am »
Is it safe for the Siglent instruments to connect a ground strap to them (to ground me) or do I need something else to connect the strap to?
If you can trust the integrity/safety of you mains supply earth I see no problems with what you propose.
However if a fault occurs somewhere and your mains earth 0V becomes elevated your ground strap will be too.

Be careful.

Oh, that's an interesting point. What is a safer way to connect these ground straps?
 

Offline ebastler

  • Super Contributor
  • ***
  • Posts: 6637
  • Country: de
Re: Siglent Technical Support join in EEVblog
« Reply #1472 on: February 07, 2021, 01:10:32 pm »
What is a safer way to connect these ground straps?

A water pipe or your room heater (assuming it is a water-filled radiator connected to central heating), if they are within reach.
 

Offline blurpy

  • Regular Contributor
  • *
  • Posts: 233
  • Country: no
Re: Siglent Technical Support join in EEVblog
« Reply #1473 on: February 07, 2021, 05:00:38 pm »
A water pipe or your room heater (assuming it is a water-filled radiator connected to central heating), if they are within reach.
I tested my radiator against mains earth and it's 2 ohm. What makes the radiator safer than directly using mains earth?
 

Offline ebastler

  • Super Contributor
  • ***
  • Posts: 6637
  • Country: de
Re: Siglent Technical Support join in EEVblog
« Reply #1474 on: February 07, 2021, 05:26:00 pm »
Whether it is safer or not will probably depend on the details of you hme's wiring. In my house (Germany, late 1960s), the water supply provides the "real" protective earth: There is no separate, dedicated earth for the electrical system, and its PEN is tied to the water pipes. Hence, one could envision scenarios where that connection goes bad or is forgotten by a sloppy electrician.

(On the other hand, I have also read about cases where part of the copper water lines were replaced by modern plastic tubes, resulting in broken earth connections there...)
 
The following users thanked this post: blurpy