So, at present, it seems to me that there is no way to permanently enable the software options. Correct?
If you want to take it further from here, to have the correct Serial number, you'll also need a way to generate the correct license info for your specific device's Serial + Host ID.
This can be done with some help from the forum, but the procedure is not public yet.
Hi guys,
my unit shows exactly 100% same issue on 1-PORT calibration procedure as described by steve1515 and eloso.
If you have a look at the review of SVA1032X by The Signal Path (https://youtu.be/ToVJTKCyIU8) you can see at 33:30 of the video that his unit shows the same issue of missing calibration status "---" on left top of his screen.
So, it seems to be definitely a buggy issue of the FW in the pro-mode.
TauTech:
Is there any chance that Siglent could fix that in an update?
Why have your Smith plots for OPEN and SHORT a long "line" response and not a "spot" response as should be? You get clean spot result instead when using the calibration menu SHORT or OPEN.
Thanks!
TauTech:
Is there any chance that Siglent could fix that in an update?
Why have your Smith plots for OPEN and SHORT a long "line" response and not a "spot" response as should be? You get clean spot result instead when using the calibration menu SHORT or OPEN.
Thanks!
So, it seems to be definitely a buggy issue of the FW in pro-mode.
Great! Would be perfect if this issue will be solved with next FW update.
My question refers to screenhots in #688 of tautech.
I know that "arcs" occure due to capacitance and inductance effects of physical connections.
However, when doing same calibration in simple "OPEN" and "SHORT" mode instead of "1-PORT" mode I get perfect nice spots.
So, it seems to be definitely a buggy issue of the FW in pro-mode.
I wouldn't call it a "bug"... You're outside the envelope.
Hi guys,
my unit shows exactly 100% same issue on 1-PORT calibration procedure as described by steve1515 and eloso.
If you have a look at the review of SVA1032X by The Signal Path (https://youtu.be/ToVJTKCyIU8) you can see at 33:30 of the video that his unit shows the same issue of missing calibration status "---" on left top of his screen.
So, it seems to be definitely a buggy issue of the FW in the pro-mode.
TauTech:
Is there any chance that Siglent could fix that in an update?
Why have your Smith plots for OPEN and SHORT a long "line" response and not a "spot" response as should be? You get clean spot result instead when using the calibration menu SHORT or OPEN.
Thanks!New FW for SVA/SSA-X-R is under construction and we're still sorting a few issues with it.
Did the product ID change in this update? If not, couldn't you just use the same telnet ADS file and update the startup script to re-enable telnet again?
upgrade 3.2.2.4->3.2.2.5 done, all hacks still working!!!
Now VNA number of points is up to 10001, that's wonderful! Unfortunately SA number of points is still not changeable and fixed to 751.
telnet is reenabled successfully with script SVA1000X_telnet_11410, followed by the usual editing to activate it on boot.
:SYSTem:TIME
:SYSTem:TIME?
:SYSTem:DATE
:SYSTem:DATE?
:SYSTem:PRESet
:SYSTem:FDEFault
REBOOT
:SYSTem:PRESet:TYPE
:SYSTem:PRESet:TYPE?
:SYSTem:PON:TYPE
:SYSTem:PON:TYPE?
:SYSTem:COMMunicate:LAN:TYPE
:SYSTem:COMMunicate:LAN:TYPE?
:SYSTem:COMMunicate:LAN:IPADdress
:SYSTem:COMMunicate:LAN:IPADdress?
:SYSTem:COMMunicate:LAN:SMASk
:SYSTem:COMMunicate:LAN:SMASk?
:SYSTem:COMMunicate:LAN:GATeway
:SYSTem:COMMunicate:LAN:GATeway?
:SYSTem:COMMunicate:LAN:MAC
:SYSTem:COMMunicate:LAN:MAC?
:SYSTem:COMMunicate:LAN:HOSTname
:SYSTem:COMMunicate:LAN:HOSTname?
NET_BRIDGE_ADDR?
SRLN
SRLN?
SGLT-UPGRADE
SGLT-UPGRADE_CFG
SGLT-UPGRADE_PATH
SGLT-UPGRADE_CFG_PATH
SGLT-TEST-VAST-DATA
SGLT-TEST-VAST-DATA?
SCDP
IDN-SGLT-PRI?
LOAD:CALI:FILE?
CHIP
CHIP?
DMESg?
SVNALL?
$$SY_FP
MMEMory:UPLoad?
LICENSE
LICENSE?
LICENSE_KEY
LICENSE_PR?
LICENSE_ID?
INNEr:PRODuction:MODE
SYSTem:FileSystem:Type?
BOARDTEST?
WEB:PSW
WEB:PSW?
FACTORYDATA
FACTORYDATA?
:DISPlay:WINDow#:TRACe#:Y[:SCALe]:RLEVel
:DISPlay:WINDow#:TRACe#:Y[:SCALe]:RLEVel?
:DISPlay:FSCan:VIEW#:WINDow#:TRACe:Y[:SCALe]:RLEVel
:DISPlay:FSCan:VIEW#:WINDow#:TRACe:Y[:SCALe]:RLEVel?
:DISPlay:WINDow#:TRACe#:Y[:SCALe]:PDIVision
:DISPlay:WINDow#:TRACe#:Y[:SCALe]:PDIVision?
:DISPlay:FSCan:VIEW#:WINDow#:TRACe:Y[:SCALe]:PDIVision
:DISPlay:FSCan:VIEW#:WINDow#:TRACe:Y[:SCALe]:PDIVision?
:DISPlay:WINDow#:TRACe#:Y[:SCALe]:RPOSition
:DISPlay:WINDow#:TRACe#:Y[:SCALe]:RPOSition?
[:SENSe]:POWer[:RF]:ATTenuation
[:SENSe]:POWer[:RF]:ATTenuation?
[:SENSe]:POWer[:RF]:ATTenuation:AUTO
[:SENSe]:POWer[:RF]:ATTenuation:AUTO?
:DISPlay:WINDow:TRACe:Y[:SCALe]:RLEVel:OFFSet
:DISPlay:WINDow:TRACe:Y[:SCALe]:RLEVel:OFFSet?
:DISPlay:FSCan:VIEW:WINDow:TRACe:Y[:SCALe]:RLEVel:OFFSet
:DISPlay:FSCan:VIEW:WINDow:TRACe:Y[:SCALe]:RLEVel:OFFSet?
[:SENSe]:POWer[:RF]:GAIN[:STATe]
[:SENSe]:POWer[:RF]:GAIN[:STATe]?
:DISPlay:WINDow:TRACe:Y[:SCALe]:SPACing
:DISPlay:WINDow:TRACe:Y[:SCALe]:SPACing?
:UNIT:POWer
:UNIT:POWer?
[:SENSe]:CORRection:IMPedance[:INPut][:MAGNitude]
[:SENSe]:CORRection:IMPedance[:INPut][:MAGNitude]?
[:SENSe]:CORRection:OFF
[:SENSe]:CORRection:CSET#:OFF
[:SENSe]:CORRection:CSET:ALL[:STATe]
[:SENSe]:CORRection:CSET:ALL[:STATe]?
[:SENSe]:CORRection:CSET#[:STATe]
[:SENSe]:CORRection:CSET#[:STATe]?
[:SENSe]:CORRection:CSET#:DATA
[:SENSe]:CORRection:CSET#:DATA?
[:SENSe]:CORRection:CSET#:ADD
[:SENSe]:CORRection:CSET#:DELete
[:SENSe]:CORRection:CSET#:ALL:DELete
:DISPlay:WINDow#:TRACe#:Y[:SCALe]:AUTO
:TRACe#:Y[:SCALe]:RLEVel
:TRACe#:Y[:SCALe]:RLEVel?
:TRACe#:Y[:SCALe]:PDIVision
:TRACe#:Y[:SCALe]:PDIVision?
:TRACe#[:Y]:AUToscale
[:SENSe#]:BWIDth[:RESolution]
[:SENSe#]:BWIDth[:RESolution]?
[:SENSe]:BWIDth:VIDeo
[:SENSe]:BWIDth:VIDeo?
[:SENSe]:BWIDth[:RESolution]:AUTO
[:SENSe]:BWIDth[:RESolution]:AUTO?
[:SENSe]:BWIDth:VIDeo:AUTO
[:SENSe]:BWIDth:VIDeo:AUTO?
[:SENSe]:BWIDth:VIDeo:RATio
[:SENSe]:BWIDth:VIDeo:RATio?
[:SENSe]:BWIDth:VIDeo:RATio:CONfig?
[:SENSe#]:BANDwidth[:RESolution]
[:SENSe#]:BANDwidth[:RESolution]?
[:SENSe]:BANDwidth:VIDeo
[:SENSe]:BANDwidth:VIDeo?
[:SENSe]:BANDwidth[:RESolution]:AUTO
[:SENSe]:BANDwidth[:RESolution]:AUTO?
[:SENSe]:BANDwidth:VIDeo:AUTO
[:SENSe]:BANDwidth:VIDeo:AUTO?
[:SENSe]:BANDwidth:VIDeo:RATio
[:SENSe]:BANDwidth:VIDeo:RATio?
[:SENSe]:BANDwidth:VIDeo:RATio:CONfig?
[:SENSe]:AVERage:TYPE
[:SENSe]:AVERage:TYPE?
[:SENSe]:FILTer:TYPE
[:SENSe]:FILTer:TYPE?
[:SENSe]:BWIDth:SHAPe
[:SENSe]:BWIDth:SHAPe?
[:SENSe]:DDEMod:FFT:WINDow:TYPE
[:SENSe]:DDEMod:FFT:WINDow:TYPE?
[:SENSe]:FSCan:SCAN:BWIDth[:RESolution]
[:SENSe]:FSCan:SCAN:BWIDth[:RESolution]?
[:SENSe]:FSCan:SCAN:BWIDth[:RESolution]:AUTO
[:SENSe]:FSCan:SCAN:BWIDth[:RESolution]:AUTO?
[:SENSe]:DEMod
[:SENSe]:DEMod?
[:SENSe]:DEMod:TIME
[:SENSe]:DEMod:TIME?
[:SENSe]:DEMod:VOLume
[:SENSe]:DEMod:VOLume?
[:SENSe]:DEMod:EPHone
[:SENSe]:DEMod:EPHone?
[:SENSe#]:FREQuency:STARt
[:SENSe#]:FREQuency:STARt?
[:SENSe#]:FREQuency:CENTer
[:SENSe#]:FREQuency:CENTer?
[:SENSe#]:FREQuency:STOP
[:SENSe#]:FREQuency:STOP?
[:SENSe#]:FREQuency:SPAN
[:SENSe#]:FREQuency:SPAN?
[:SENSe]:FREQuency:CENTer:STEP[:INCRement]
[:SENSe]:FREQuency:CENTer:STEP[:INCRement]?
[:SENSe]:FREQuency:CENTer:STEP:AUTO
[:SENSe]:FREQuency:CENTer:STEP:AUTO?
:CALCulate:MARKer:TRCKing[:STATe]
:CALCulate:MARKer:TRCKing[:STATe]?
[:SENSe]:FREQuency:SPAN:FULL
[:SENSe]:FREQuency:SPAN:ZERO
[:SENSe]:FREQuency:SPAN:HALF
[:SENSe]:FREQuency:SPAN:DOUBle
[:SENSe]:FREQuency:SPAN:PREVious
[:SENSe]:FREQuency:CENTer:SET:STEP
[:SENSe]:FREQuency:OFFSet
[:SENSe]:FREQuency:OFFSet?
:DISPlay:WINDow:TRACe:X[:SCALe]:SPACing
:DISPlay:WINDow:TRACe:X[:SCALe]:SPACing?
[:SENSe]:FREQuency:TUNE:IMMediate
[:SENSe#]:FREQuency:DATA?
[:SENSe]:FSCan:RANGe
[:SENSe]:FREQuency:MIDSpan
[:SENSe]:FREQuency:MIDSpan?
:INNer:SET:MODE:REMOte
:INNer:VERS:MATCh:ID?
:INNer:ALL:DP:DATA?
:INNer:ALL:EMI:DATA?
:INNer:AMPT:PARM?
:INNer:AMPT:CORR:PARM?
:INNer:FREQuency:PARM?
:INNer:BW:PARM?
:INNer:ALL:TRACE:PARM?
:INNer:PEAK:PARM?
:INNer:SWEEP:PARM?
:INNer:TRIG:PARM?
:INNer:LIMit:PARM?
:INNer:TG:PARM?
:INNer:MEAS:PARM?
:INNer:MODE:PARM?
:INNer:MARKer#:PARM?
:INNer:LOAD:FILE
:INNer:EXPort?
DEBug:HWACcess:FPGA
DEBug:HWACcess:FPGA?
DEBug:HWACcess:CPLD
DEBug:HWACcess:CPLD?
DEBug:HWACcess:PLL1
DEBug:HWACcess:PLL1?
DEBug:HWACcess:PLL2
DEBug:HWACcess:PLL2?
DEBug:HWACcess:FILTer:RF
DEBug:HWACcess:TG
DEBug:HWACcess:TG?
DEBug:HWACcess:ATT:TG
DEBug:HWACcess:ATT:TG?
DEBug:HWACcess:TEST:SWITch
DEBug:HWACcess:PLL3
DEBug:HWACcess:PLL3?
DEBug:HWACcess:RTFPga
DEBug:HWACcess:RTFPga?
DEBug:HWACcess:REFPLL
DEBug:HWACcess:REFPLL?
DEBug:HWACcess:ADC
DEBug:HWACcess:ADC?
DEBug:HWACcess:VGA
DEBug:HWACcess:VGA?
DEBug:HWACcess:LOAD:RTFPGA
DEBug:HWACcess:INIT:PLL
DEBug:HWACcess:DAC
DEBug:HWACcess:DAC?
DEBug:HWACcess:IFDAc
DEBug:HWACcess:IFDAc?
DEV:INNersorce
DEBug:TEMPrt:BOARd?
DEBug:TEMPrt:ZCPU?
DEBug:ERASe:CALibration:DATA
DEBug:FACTory:MODE
DEBug:IF:FILTer
:TEMPreature:COMP:VALue?
:DEBug:INNersorce:DEFault:VALue?
:BOARdtest:FPGA:VERSion?
:BOARdtest:RTSA:FPGA:VERSion?
:BOARdtest:CPLD:VERSion?
:BOARdtest:PCBid:VERSion?
:BOARdtest:FPGA:RW?
:BOARdtest:DDR3:TEST?
:BOARdtest:FLASh:TEST?
:BOARdtest:LAN:TEST?
:BOARdtest:USBHost:TEST?
:BOARdtest:PLL40lock:TEST?
:BOARdtest:EXTTrig:TEST?
:BOARdtest:EXTRefsource:TEST?
DEBug:FLASh:DATA
DEBug:FLASh:DATA?
DEBug:ERASe:FLASh:SECTor
DEBug:ERASe:FLASh:BULK
DEBug:S11:CALFile:SYNC:TO:FLASh
:CALCulate:LLINe#:STATe
:CALCulate:LLINe#:STATe?
:CALCulate:FSCan:LLINe#:DISPlay
:CALCulate:FSCan:LLINe#:DISPlay?
:CALCulate:FSCan:LLINe#:MARGin
:CALCulate:FSCan:LLINe#:MARGin?
:CALCulate:FSCan:LLINe#:MARGin:STATe
:CALCulate:FSCan:LLINe#:MARGin:STATe?
:CALCulate:FSCan:LLINe#:TRACe
:CALCulate:FSCan:LLINe#:TRACe?
:CALCulate:LLINe#:TYPE
:CALCulate:LLINe#:TYPE?
:CALCulate:LLINe#:MODE
:CALCulate:LLINe#:MODE?
:CALCulate:FSCan:LLINe#:MODE
:CALCulate:FSCan:LLINe#:MODE?
:CALCulate:LLINe#:Y
:CALCulate:LLINe#:Y?
:CALCulate:FSCan:LLINe#:Y
:CALCulate:FSCan:LLINe#:Y?
:CALCulate:LLINe#:DATA
:CALCulate:LLINe#:DATA?
:CALCulate:FSCan:LLINe#:DATA
:CALCulate:FSCan:LLINe#:DATA?
:CALCulate:LLINe#:OFFSet:X
:CALCulate:LLINe#:OFFSet:X?
:CALCulate:LLINe#:OFFSet:Y
:CALCulate:LLINe#:OFFSet:Y?
:CALCulate:LLINe#:ADD
:CALCulate:LLINe#:DELete
:CALCulate:LLINe#:ALL:DELete
:CALCulate:LLINe:TEST:STARt
:CALCulate:LLINe:TEST:STOP
:CALCulate:LLINe:TEST:STATe?
CALCulate:LLINe:TEST
:CALCulate:LLINe:TEST?
:CALCulate:LLINe:FAIL?
:CALCulate:FSCan:LLINe:FAIL?
:CALCulate:LLINe:CONTrol:DOMain
:CALCulate:LLINe:CONTrol:DOMain?
:CALCulate:LLINe:CONTrol:BEEP
:CALCulate:LLINe:CONTrol:BEEP?
:CALCulate:LLINe:FAIL:STOP
:CALCulate:LLINe:FAIL:STOP?
:CALCulate:LLINe#:CONTrol:INTerpolate
:CALCulate:LLINe#:CONTrol:INTerpolate?
:CALCulate:FSCan:LLINe#:CONTrol:INTerpolate
:CALCulate:FSCan:LLINe#:CONTrol:INTerpolate?
[:SENSe]:DDEMod:RAWiq:COLLect
[:SENSe]:DDEMod:RAWiqdata?
[:SENSe]:MA:SAMPle:MAX?
[:SENSe]:MA:RAWiq:COLLect
[:SENSe]:MA:RAWiqdata?
[:SENSe]:MA:PARSe:RATE?
[:SENSe]:MA:PARSe:SIZE?
[:SENSe]:MA:BUFFer:SIZE?
:CALCulate:MARKer:TABLe[:STATe]
:CALCulate:MARKer:TABLe[:STATe]?
:CALCulate[:SELected]:MARKer:AOFF
:CALCulate#:MARKer#:STATe
:CALCulate#:MARKer#:STATe?
:CALCulate:IFBW:INDEx?
:CALCulate:EQLPf:INDEx?
:CALCulate:MARKer#:MODE
:CALCulate:MARKer#:MODE?
:CALCulate:FSCan:MARKer#:MODE
:CALCulate:FSCan:MARKer#:MODE?
:CALCulate:MARKer#:TRACe
:CALCulate:MARKer#:TRACe?
:CALCulate:FSCan:MARKer#:TRACe
:CALCulate:FSCan:MARKer#:TRACe?
:CALCulate:MARKer#:REFerence
:CALCulate:MARKer#:REFerence?
:CALCulate:FSCan:MARKer#:REFerence
:CALCulate:FSCan:MARKer#:REFerence?
:CALCulate:MARKer#:RELative:TO:MARKer
:CALCulate:MARKer#:RELative:TO:MARKer?
:CALCulate:MARKer#:X:READout
:CALCulate:MARKer#:X:READout?
:CALCulate:MARKer#:FUNCtion
:CALCulate:MARKer#:FUNCtion?
:CALCulate:MARKer#:BANDwidth:NDB
:CALCulate:MARKer#:BANDwidth:NDB?
:CALCulate:MARKer#:BWIDth:NDB
:CALCulate:MARKer#:BWIDth:NDB?
:CALCulate:MARKer#:BANDwidth:RESult?
:CALCulate:MARKer#:BANDwidth:RLEFt?
:CALCulate:MARKer#:BANDwidth:RRIGht?
:CALCulate:MARKer#:BWIDth:RESult?
:CALCulate:MARKer#:BWIDth:RLEFt?
:CALCulate:MARKer#:BWIDth:RRIGht?
:CALCulate:MARKer#:FCOunt[:STATe]
:CALCulate:MARKer#:FCOunt[:STATe]?
:CALCulate:MARKer#:FCOunt:X?
:CALCulate#:MARKer#:X
:CALCulate#:MARKer#:X?
:CALCulate#:FSCan:MARKer#:X
:CALCulate#:FSCan:MARKer#:X?
:CALCulate:MARKer#:X:POSition
:CALCulate:MARKer#:X:POSition?
:CALCulate#:MARKer#:Y
:CALCulate#:MARKer#:Y?
:CALCulate#:FSCan:MARKer#:Y?
:CALCulate:MARKer#[:SET]:CENTer
:CALCulate:MARKer#[:SET]:STEP
:CALCulate:MARKer#[:SET]:STARt
:CALCulate:MARKer#[:SET]:STOP
:CALCulate:MARKer#[:SET]:RLEVel
:CALCulate:MARKer#[:SET]:DELTa:SPAN
:CALCulate:MARKer#[:SET]:DELTa:CENTer
:CALCulate#[:SELected]:MARKer:COUPle
:CALCulate#[:SELected]:MARKer:COUPle?
:CALCulate#[:SELected]:MARKer:DISCrete
:CALCulate#[:SELected]:MARKer:DISCrete?
:CALCulate#[:SELected]:MARKer:REFerence[:STATe]
:CALCulate#[:SELected]:MARKer:REFerence[:STATe]?
:TRACe#:MARKer#:X
:TRACe#:MARKer#:X?
:TRACe#:MARKer#:Y?
:TRACe#:MARKer#:REFerence
:TRACe#:MARKer#:REFerence?
:TRACe:MARKer:COUPle
:TRACe:MARKer:COUPle?
:TRACe#:MARKer#:TYPE
:TRACe#:MARKer#:TYPE?
:TRACe#:MARKer#:ENABle
:TRACe#:MARKer#:ENABle?
:CALCulate:FSCan:MARKer#[:SET]:SLISt
:CALCulate:FSCan:MARKer#[:SET]:METer
:CALCulate:FSCan:MARKer#:TO:METer
INSTrument:MEASure
INSTrument:MEASure?
:DISPlay:VIEW[:SELect]
:DISPlay:VIEW[:SELect]?
[:SENSe]:MEASure:REFLction#:RETUrnloss?
[:SENSe]:MEASure:REFLction#:COEFficient?
[:SENSe]:MEASure:REFLction#:VSWR?
[:SENSe]:CHPower:FREQuency:CENTer
[:SENSe]:CHPower:FREQuency:CENTer?
[:SENSe]:CHPower:BWIDth:INTegration
[:SENSe]:CHPower:BWIDth:INTegration?
[:SENSe]:CHPower:BANDwidth:INTegration
[:SENSe]:CHPower:BANDwidth:INTegration?
[:SENSe]:CHPower:FREQuency:SPAN
[:SENSe]:CHPower:FREQuency:SPAN?
[:SENSe]:CHPower:FREQuency:SPAN:POWer
:MEASure:CHPower?
:MEASure:CHPower:CHPower?
:MEASure:CHPower:DENSity?
[:SENSe]:ACPRatio:BANDwidth:INTegration
[:SENSe]:ACPRatio:BWIDth:INTegration
[:SENSe]:ACPRatio:BANDwidth:INTegration?
[:SENSe]:ACPRatio:BWIDth:INTegration?
[:SENSe]:ACPRatio:OFFSet[:FREQuency]
[:SENSe]:ACPRatio:OFFSet[:FREQuency]?
[:SENSe]:ACPRatio:OFFSet:BANDwidth[:INTegration]
[:SENSe]:ACPRatio:OFFSet:BWIDth[:INTegration]
[:SENSe]:ACPRatio:OFFSet:BANDwidth[:INTegration]?
[:SENSe]:ACPRatio:OFFSet:BWIDth[:INTegration]?
:MEASure:ACPRatio:MAIN?
:MEASure:ACPRatio:LOWer:POWer?
:MEASure:ACPRatio:LOWer?
:MEASure:ACPRatio:UPPer:POWer?
:MEASure:ACPRatio:UPPer?
[:SENSe]:OBWidth:METHod
[:SENSe]:OBWidth:METHod?
[:SENSe]:OBWidth:PERCent
[:SENSe]:OBWidth:PERCent?
[:SENSe]:OBWidth:XDB
[:SENSe]:OBWidth:XDB?
:MEASure:OBWidth?
:FETCh:OBWidth?
:MEASure:OBWidth:OBWidth?
:FETCh:OBWidth:OBWidth?
:MEASure:OBWidth:CENTroid?
:FETCh:OBWidth:CENTroid?
:MEASure:OBWidth:OBWidth:FERRor?
:FETCh:OBWidth:OBWidth:FERRor?
:FETCh:TPOWer?
:MEASure:TPOWer?
[:SENSe]:TPOWer:FREQuency:CENTer
[:SENSe]:TPOWer:FREQuency:CENTer?
[:SENSe]:TPOWer:LLIMit
[:SENSe]:TPOWer:LLIMit?
[:SENSe]:TPOWer:RLIMit
[:SENSe]:TPOWer:RLIMit?
:MEASure:TOI?
:MEASure:TOI:IP3?
[:SENSe]:SPECtrogram:STATe
[:SENSe]:SPECtrogram:STATe?
[:SENSe]:SPECtrogram:RESTart
[:SENSe]:HARMonics:FREQuency:FUNDamental
[:SENSe]:HARMonics:FREQuency:FUNDamental?
[:SENSe]:HARMonics:FREQuency:FUNDamental:AUTO
[:SENSe]:HARMonics:FREQuency:FUNDamental:AUTO?
[:SENSe]:HARMonics:FREQuency:STEP[:INCRement]
[:SENSe]:HARMonics:FREQuency:STEP[:INCRement]?
[:SENSe]:HARMonics:FREQuency:STEP[:INCRement]:AUTO
[:SENSe]:HARMonics:FREQuency:STEP[:INCRement]:AUTO?
[:SENSe]:HARMonics:NUMBer
[:SENSe]:HARMonics:NUMBer?
:FETCh:HARMonics:AMPLitude:ALL?
:MEASure:HARMonics:AMPLitude:ALL?
:FETCh:HARMonics:AMPLitude?
:MEASure:HARMonics:AMPLitude?
:FETCh:HARMonics:FREQuency:ALL?
:MEASure:HARMonics:FREQuency:ALL?
:FETCh:HARMonics:FREQuency?
:MEASure:HARMonics:FREQuency?
[:SENSe]:CNRatio:BANDwidth:INTegration
[:SENSe]:CNRatio:BANDwidth:INTegration?
[:SENSe]:CNRatio:BANDwidth:NOISe
[:SENSe]:CNRatio:BANDwidth:NOISe?
[:SENSe]:CNRatio:OFFSet
[:SENSe]:CNRatio:OFFSet?
:MEASure:CNRatio?
:MEASure:CNRatio:CARRier?
:MEASure:CNRatio:NOISe?
[:SENSe]:HARMonics:SELect
[:SENSe]:HARMonics:SELect?
:CALCulate#:PARameter#:DEFine
:CALCulate#:PARameter#:DEFine?
:CALCulate#[:SELected]:FORMat
:CALCulate#[:SELected]:FORMat?
[:SENSe#]:CORRection:EXTension[:STATe]
[:SENSe#]:CORRection:EXTension[:STATe]?
[:SENSe#]:CORRection:EXTension:PORT#:TIME
[:SENSe#]:CORRection:EXTension:PORT#:TIME?
[:SENSe#]:CORRection:RVELocity:COAX
[:SENSe#]:CORRection:RVELocity:COAX?
[:SENSe#]:CORRection:EXTension:AUTO:PORT
[:SENSe#]:CORRection:COLLect[:ACQuire]:LOAD
[:SENSe#]:CORRection:COLLect[:ACQuire]:OPEN
[:SENSe#]:CORRection:COLLect[:ACQuire]:SHORt
[:SENSe#]:CORRection:COLLect[:ACQuire]:THRU
[:SENSe#]:CORRection:COLLect:CKIT:LABel
[:SENSe#]:CORRection:COLLect:CKIT:LABel?
[:SENSe#]:CORRection:COLLect:CKIT:LABel:CATalog?
[:SENSe#]:CORRection:COLLect:CKIT:GENDer
[:SENSe#]:CORRection:COLLect:CKIT:GENDer?
[:SENSe#]:CORRection:COLLect:METHod[:RESPonse]:THRU
[:SENSe#]:CORRection:COLLect:METHod:SOLT1
[:SENSe#]:CORRection:COLLect:METHod:TYPE?
[:SENSe#]:CORRection:COLLect:CLEar
[:SENSe#]:CORRection:COLLect:SAVE
CALCulate:TRANsform:DISTance:STARt
CALCulate:TRANsform:DISTance:STARt?
CALCulate:TRANsform:DISTance:STOP
CALCulate:TRANsform:DISTance:STOP?
CALCulate:TRANsform:DISTance:UNIT
CALCulate:TRANsform:DISTance:UNIT?
CALCulate:TRANsform:DISTance:WINDow
CALCulate:TRANsform:DISTance:WINDow?
[:SENSe]:CORRection:LOSS:COAX
[:SENSe]:CORRection:LOSS:COAX?
[:SENSe]:DDEMod[:FORMat]:RLENgth
[:SENSe]:DDEMod[:FORMat]:RLENgth?
[:SENSe]:DDEMod[:FORMat]:SRATe
[:SENSe]:DDEMod[:FORMat]:SRATe?
[:SENSe]:DDEMod:MODulation
[:SENSe]:DDEMod:MODulation?
[:SENSe]:ADEMod:STYLe
[:SENSe]:ADEMod:STYLe?
[:SENSe]:DDEMod:FILTer[:MEASurement]
[:SENSe]:DDEMod:FILTer[:MEASurement]?
[:SENSe]:DDEMod:FILTer:ABT
[:SENSe]:DDEMod:FILTer:ABT?
[:SENSe]:DDEMod:FILTer:REFerence
[:SENSe]:DDEMod:FILTer:REFerence?
[:SENSe]:DDEMod:FILTer:RLENgth
[:SENSe]:DDEMod:FILTer:RLENgth?
[:SENSe]:STATistic:STATe
[:SENSe]:STATistic:STATe?
[:SENSe]:DDEMod[:FORMat]:SYMBol:POINts
[:SENSe]:DDEMod[:FORMat]:SYMBol:POINts?
:READ:DDEMod?
:READ:ADEMod?
:DISPlay:VIEW:DENSity:PERSistence
:DISPlay:VIEW:DENSity:PERSistence?
[:SENSe]:CAT:RST
[:SENSe]:CAT:FREFlect:TYPE
[:SENSe]:CAT:FREFlect:TYPE?
[:SENSe]:CAT:FREFlect:OPEN
[:SENSe]:CAT:FREFlect:SHORt
[:SENSe]:CAT:FREFlect:LOAD
:DISPlay:VIEW:DENSity:PERSistence:INFinite
:DISPlay:VIEW:DENSity:PERSistence:INFinite?
:DISPlay:VIEW:SPECtrogram:TRACe:SELection
:DISPlay:VIEW:SPECtrogram:TRACe:SELection?
:DISPlay:VIEW:SPECtrogram:TRACe:STARt
:DISPlay:VIEW:SPECtrogram:TRACe:STARt?
:DISPlay:VIEW:SPECtrogram:TRACe:NUMber
:DISPlay:VIEW:SPECtrogram:TRACe:NUMber?
:CALCulate:RESTart
[:SENSe]:DDEMod:EVM:RMS?
[:SENSe]:DDEMod:EVM:PK?
[:SENSe]:DDEMod:ASK:ERR?
[:SENSe]:DDEMod:ASK:DEPTh?
[:SENSe]:DDEMod:FSK:ERR?
[:SENSe]:DDEMod:FSK:DEVI?
TEST
[:SENSe]:FSCan:SEQuence
[:SENSe]:FSCan:SEQuence?
[:SENSe]:FSCan:SCAN#:POINts?
[:SENSe]:FSCan:SCAN#:TIME
[:SENSe]:FSCan:SCAN#:TIME?
:DISPlay:METer#[:STATe]
:DISPlay:METer#[:STATe]?
[:SENSe]:METer#:DETector:DWELl
[:SENSe]:METer#:DETector:DWELl?
[:SENSe]:METer#:DETector
[:SENSe]:METer#:DETector?
:CALCulate:METer#:LIMit[:DATA]
:CALCulate:METer#:LIMit[:DATA]?
:CALCulate:METer#:LIMit:STATe
:CALCulate:METer#:LIMit:STATe?
:CALCulate:METer#:LIMit:ULLine
:CALCulate:SLISt:MARK:SIGNal
:CALCulate:SLISt:MARK:CLEar:SIGNal
:CALCulate:SLISt:MARK:ALL
:CALCulate:SLISt:MARK:CLEar:ALL
:CALCulate:SLISt:DELete:SIGNal
:CALCulate:SLISt:DELete:ALL
:CALCulate:SLISt:SORT:TYPE
:CALCulate:SLISt:SORT:ORDer
:CALCulate:SLISt:SORT:ORDer?
:CALCulate:METer:POWer[:CURRent]?
[:SENSe]:METer:PHOLd:RESet
:SEQ:RUN?
:READ:DDEMod:BIT?
CALCulate:MARKer#:MAXimum
CALCulate:MARKer#:MINimize
CALCulate:FSCan:MARKer#:MAXimum
CALCulate1:MARKer#:MAXimum
CALCulate1:MARKer#:MINimize
CALCulate:MARKer#:MAXimum[:SET]:CENTer
CALCulate:MARKer#:MAXimum:NEXT
CALCulate:MARKer#:MAXimum:LEFT
CALCulate:FSCan:MARKer#:MAXimum:LEFT
CALCulate:MARKer#:MAXimum:RIGHt
CALCulate:FSCan:MARKer#:MAXimum:RIGHt
CALCulate:MARKer#:PTPeak
CALCulate:MARKer#:CPSearch[:STATe]
CALCulate:MARKer#:CPSearch[:STATe]?
CALCulate:MARKer#:CPEak[:STATe]
CALCulate:MARKer#:CPEak[:STATe]?
CALCulate:MARKer#:CVSearch[:STATe]
CALCulate:MARKer#:CVSearch[:STATe]?
CALCulate:MARKer:PEAK:TABLe[:STATe]
CALCulate:MARKer:PEAK:TABLe[:STATe]?
CALCulate:PEAK:TABLe?
CALCulate:MARKer:PEAK:EXCursion
CALCulate:MARKer:PEAK:EXCursion?
CALCulate:MARKer:PEAK:THReshold
CALCulate:MARKer:PEAK:THReshold?
CALCulate:MARKer:PEAK:SEARch:MODE
CALCulate:MARKer:PEAK:SEARch:MODE?
CALCulate:MARKer:PEAK:SORT
CALCulate:MARKer:PEAK:SORT?
[:SENSe]:SWEep:TIME
[:SENSe]:SWEep:TIME?
[:SENSe]:SWEep:TIME:AUTO
[:SENSe]:SWEep:TIME:AUTO?
[:SENSe]:ACQuisition:TIME
[:SENSe]:ACQuisition:TIME?
[:SENSe]:ACQuisition:TIME:AUTO
[:SENSe]:ACQuisition:TIME:AUTO?
[:SENSe]:SWEep:SPEed
[:SENSe]:SWEep:SPEed?
:INITiate#:CONTinuous
:INITiate#:CONTinuous?
:INITiate#[:IMMediate]
:INITiate#:RESTart
[:SENSe]:FSCan:SCAN:ABORt
[:SENSe]:SWEep:MODE
[:SENSe]:SWEep:MODE?
[:SENSe]:SWEep:COUNt
[:SENSe]:SWEep:COUNt?
[:SENSe]:FSCan:SCAN#:PRBW
[:SENSe]:FSCan:SCAN#:PRBW?
[:SENSe]:FSCan:SCAN#:PRBW:AUTO
[:SENSe]:FSCan:SCAN#:PRBW:AUTO?
[:SENSe]:QPD:DWELl:TIME
[:SENSe]:QPD:DWELl:TIME?
[:SENSe#]:SWEep:POINts
[:SENSe#]:SWEep:POINts?
:INITiate:PAUSe
:INITiate:RESume
ABORt
*OPC
*OPC?
*WAI
*TRG
*IDN?
DEBTTT
DEBug:CALibration:MODE
[:SENSe]:CALibration:STATe
*RST
:SYSTem:CLEAr
:SYSTem:RESTart
:SYSTem:POWer:OFF
SCOPeid?
DISconnectDEVice
:SYSTem:CONFigure:SYSTem?
HCOPy:SDUMp:DATA?
:KEY:CONTrol
:MMEMory:PATH
:MMEMory:STORe
:MMEMory:LOAD
:MMEMory:DELete
:SYSTem:LKEY
:SYSTem:LKEY?
:SYSTem:LKEY:DELete
:SYSTem:OPTions?
:SYSTem:LKEY:TIME:RESet
:SYSTem:MODE:FACTory?
:SYSTem:LANGuage
:SYSTem:LANGuage?
:SYSTem:COMMunicate:LAN:CONFig:APPLy
:INSTrument[:SELect]
:INSTrument[:SELect]?
DEBug:CALibration:FILe:SAVe
DEBug:SWEEp:STOP
DEBug:SWEEp:RESTart
:DISPlay:WINDow:TRACe:Y:DLINe:STATe
:DISPlay:WINDow:TRACe:Y:DLINe:STATe?
:DISPlay:WINDow:TRACe:Y:DLINe
:DISPlay:WINDow:TRACe:Y:DLINe?
:DISPlay:WINDow:TRACe:GRATicule:GRID:BRIGhtness
:DISPlay:WINDow:TRACe:GRATicule:GRID:BRIGhtness?
:SYSTem:PRESet:USER#:SAVE
:SYSTem:PRESet:USER#:LOAD
[:SENSe]:EXTReference:AUTO
[:SENSe]:EXTReference:AUTO?
:DISPlay:WINDow:SELect
:DISPlay:WINDow:SELect?
:DEV:SVA?
*OPT?
INST:CAT?
:SYSTem:INFO:VERSion?
:RELease:VERSion?
:SYSTem:OVERload:TIME?
:OUTPut[:STATe]
:OUTPut[:STATe]?
:SOURce#:POWer[:LEVel][:IMMediate][:AMPLitude]
:SOURce#:POWer[:LEVel][:IMMediate][:AMPLitude]?
:SOURce:CORRection:OFFSet
:SOURce:CORRection:OFFSet?
:CALCulate:NTData[:STATe]
:CALCulate:NTData[:STATe]?
:DISPlay:WINDow:TRACe:Y[:SCALe]:NRLevel
:DISPlay:WINDow:TRACe:Y[:SCALe]:NRLevel?
:DISPlay:WINDow:TRACe:Y[:SCALe]:NRPosition
:DISPlay:WINDow:TRACe:Y[:SCALe]:NRPosition?
:DISPlay:WINDow:NTTRace[:STATe]
:DISPlay:WINDow:NTTRace[:STATe]?
:TRACe#:MODE
:TRACe#:MODE?
:TRACe#:FSCan:TYPE
:TRACe#:FSCan:TYPE?
[:SENSe]:AVERage:TRACe#:COUNt
[:SENSe]:AVERage:TRACe#:COUNt?
[:SENSe]:AVERage:TRACe#:CLEar
[:SENSe]:DETector:TRACe#[:FUNCtion]
[:SENSe]:DETector:TRACe#[:FUNCtion]?
[:SENSe]:FSCan:DETector:TRACe#
[:SENSe]:FSCan:DETector:TRACe#?
[:SENSe]:DETector:TRACe:PVTime
[:SENSe]:DETector:TRACe:PVTime?
[:SENSe]:DETector:TRACe:SPECtrogram
[:SENSe]:DETector:TRACe:SPECtrogram?
[:SENSe]:AVERage:TRACe#?
:CALCulate#[:SELected]:MATH:FUNCtion
:CALCulate#[:SELected]:MATH:FUNCtion?
:TRACe:MATH:TYPE
:TRACe:MATH:TYPE?
:TRACe:MATH:Z
:TRACe:MATH:X
:TRACe:MATH:Y
:TRACe:MATH:Z?
:TRACe:MATH:X?
:TRACe:MATH:Y?
:TRACe:MATH:OFFSet
:TRACe:MATH:OFFSet?
:TRACe#[:DATA]?
:TRACe:REF[:DATA]?
:TRACe#:FSCan[:DATA]?
:TRACe:ADC[:DATA]?
:TRACe[:DATA]:SPECtrum?
:TRACe[:DATA]:PVT?
:FORMat[:TRACe][:DATA]
:FORMat[:TRACe][:DATA]?
:CALCulate#:PARameter:COUNt
:CALCulate#:PARameter:COUNt?
:CALCulate#:PARameter#:SELect
:TRACe:SELect?
:CALCulate#[:SELected]:MATH:MEMorize
[:SENSe#]:AVERage:TRACe#:STATe
[:SENSe#]:AVERage:TRACe#:STATe?
:DISPlay:WINDow#:TRACe#:STATe
:DISPlay:WINDow#:TRACe#:STATe?
:DISPlay:WINDow#:TRACe#:MEMory[:STATe]
:DISPlay:WINDow#:TRACe#:MEMory[:STATe]?
:CALCulate#[:SELected]:DATA:FDATa
:CALCulate#[:SELected]:DATA:FDATa?
:CALCulate#[:SELected]:DATA:FMEMory
:CALCulate#[:SELected]:DATA:FMEMory?
[:SENSe#]:AVERage:COUNt
[:SENSe#]:AVERage:COUNt?
[:SENSe#]:AVERage[:STATe]
[:SENSe#]:AVERage[:STATe]?
:TRACe#:HOLD
:TRACe#:HOLD?
:TRACe#:DATA:NAME
:TRACe#:DATA:NAME?
:TRACe#:FORMat[:Y]
:TRACe#:FORMat[:Y]?
:DISPlay:LAYout
:TRACe:DEMod:EYE:LENGth
:TRACe:DEMod:EYE:LENGth?
:TRACe:DEMod:TABLe:FORMat
:TRACe:DEMod:TABLe:FORMat?
:TRACe:COPY
:TRACe:COPY:REF
:TRIGger[:SEQuence]:SOURce
:TRIGger[:SEQuence]:SOURce?
:TRIGger[:SEQuence]:VIDeo:LEVel
:TRIGger[:SEQuence]:VIDeo:LEVel?
:TRIGger[:SEQuence]:RFBurst:SLOPe
:TRIGger[:SEQuence]:RFBurst:SLOPe?
:TRIGger[:SEQuence]:RF:LEVel
:TRIGger[:SEQuence]:RF:LEVel?
:TRIGger[:SEQuence]:LEVel:LEVel
:TRIGger[:SEQuence]:LEVel:LEVel?
:TRIGger[:SEQuence]:LEVel:DELay
:TRIGger[:SEQuence]:LEVel:DELay?
:TRIGger[:SEQuence]:EXTernal:SLOPe
:TRIGger[:SEQuence]:EXTernal:SLOPe?
:TRIGger[:SEQuence]:EXTernal:DELay
:TRIGger[:SEQuence]:EXTernal:DELay?
TRIGger[:SEQuence]:FMT:MASK:BUILd
:TRIGger[:SEQuence]:FMT:TCRiteria
:TRIGger[:SEQuence]:FMT:TCRiteria?
:TRIGger[:SEQuence]:FMT:STATe
:TRIGger[:SEQuence]:FMT:STATe?
:TRIGger[:SEQuence]:FMT:ACTion
:TRIGger[:SEQuence]:FMT:ACTion?
:TRIGger[:SEQuence]:FMT:TRACe
:TRIGger[:SEQuence]:FMT:TRACe?
:TRIGger[:SEQuence]:FMT:MASK:DELete:ALL
DEBug:S11:CALibration:STATe
Hi,
what I found out for a unit in Pro mode (SVA3021x+) regarding VNA calibration & firmware 3.2.2.5.0.r8.
1) 1-port cal - does not really work suffiently.
2) Enhanced response calibration is working quite well. It does not change the start & stop frequencies, nor the number of points. After a performed calibration it is saving the data internally, Calibration on/off is only working from the enhanced response menu. This is true for a cold boot, too. Only downside you need to memorize the number of points you performed for the calibration. As soon as you apply the "right" number of points, start & stop freq., the calibration is used and showing "COR" in the upper left corner.
3) you can save calibration data as ".CAL" for example on the internal flash. When you load the data, you only need to apply points, start & stop - and again the "COR" shows up. (note them in the file name!)
4) For me it seems all preset and user Calibration Standards are not working. Port extension is not saved, either.