My detailed measurement process:
I carefully soldered each DUT to the wires going to the Lemo connector, then cleaned the solder joints to remove any flux residue. After soldering, I let the resistors sit for 3 days to allow them to recover from thermal and mechanical stress. I prepared them in batches of five.
Before starting any measurements, all instruments were powered on for 2 hours to stabilize. The MIT meters were logging data via their dedicated software throughout the entire process. The temperature calibrator was programmed to automatically step through the setpoints, so my only tasks were to process the data afterward and install the next resistor for testing.
For each temperature point, I took the average resistance value during the stable period (after the 30-minute soak) and entered it into the Excel table, where the TCR is calculated automatically.
I followed the calculation method described in the Fluke 742A Series Instruction Manual, which also uses the 18–28°C temperature range. This seemed like a well-established reference, so I adopted the same approach.
About the MIT meters:
The MIT precision temperature meters have two important features:
Thermoelectric EMF compensation and built-in constants that allow them to calculate how their internal references change with ambient temperature and apply real-time corrections automatically.
Why I didn't measure the actual temperature continuously:
I decided to rely on the calibrator's setpoint rather than measuring the temperature inside the block for every resistor. My reasoning is simple: I doubt any meter could resolve a resistance change caused by a 0.007°C temperature variation when measuring a precision resistor. This is clearly visible when measuring something like a VHP101T — the graph shows mostly instrument noise, not actual instability of the DUT. So I use a single temperature value for calculations, trusting the calibrator's stability data I provided earlier.
You may notice two sheets in the Excel file: "100 Ом EE (HP)" and "100 Ом EE". The "(HP)" suffix means I used the HP34401A for those measurements. The average result was 2.34 ppm with the HP versus 1.99 ppm with the MIT 8.30 — a difference of about 0.35 ppm.
The reason I did this comparison is that the MIT 8.30 has a 10 kΩ measurement limit, so for any resistor above 10 kΩ I'll have to use the HP34401A. I wanted to understand the offset between the two instruments.
I plan to continue testing various resistors and will update the table as new data comes in.