Could you tell us what type of construction you have made and at what temperature you run the oven?
The construction is pretty close to the evaluation board. It's an FR4 PCB with cutouts around the ADR1001 with some additional components, a plastic cap covering that section and the entire PCB placed in a Hammond enclosure. Oven setpoint is 60 degC.
I would also monitor the humidity during the ageing measurement.
And with seasonal changes you will see the final result after minimum 1 year.
And please show photos from PCB and mounting method of the chip on the PCB.
Agreed, this is still very preliminary, longer term testing will tell. Next tests steps will include climate chamber tests and switching the Duts off and back on to determine hysteresis.
Since this is a commercial project i'm afraid I can't share photos, as mentioned though, it's not that different from the evaluation board.
The graph shows that there is ambient temperature influence, but not the simple, immediate one described by the TC parameter. One can see delayed compensation, typical for a meter (more complex than a voltage reference).
Is the meter exposed to similar ambient temperature variations?
Is the meter new as well? Is it powered continously or was it turned on for this experiment?
Apart from that some ppm of initial drift is normal for a zener reference, let's say within first week.
Regards, Dieter
I think that's hard to say from these hourly averaged plots. The reference is covered with a plastic cap and inside an aluminium enclosure, which I guess can also cause some delay.
The meter is indeed exposed to the same ambient. Meter has been in use for prolonged time and was allowed to stabilize for at least 24 hours before these tests. But here there could definitely be some drift by DMM7510 aswell to take into account.
Agreed that the initial drift is as expected, even more inline with datasheet spec than the drift of the eval board as recorded by Andreas.
The datasheet states a typical drift of -3 ppm (15 µV for 5 V) in the first 200 hours (~8 days) and -5 ppm (25 µV for 5 V) for 1000 hours (~42 days) for the raw zener core (6.6 V).
Additional drift from the divider creating the 5 V out of the 6.6 V is to be expected as well. So besides the temperature influence, that can be seen in your graph and that dieter1 already pointed out, the drift you are seeing is not too surprising (carefully read the datasheet). Other people already pointed out before, that the divider and/or the buffer amplifier are not superior and rather noisy as well.
-branadic-
Yes agreed, comparing to the datasheet the drift is not far from the typical values, looking at the plots though, f.e. plot 12, the initial drift for my 4 prototypes is below the -1 sigma line, at least on the first 150 hours. Time will have to tell if it's just faster drift or also a bigger drift.
My surprise was not so much comparing to the datasheet but more to the data of Andreas. So I guess I'm just happy that I didn't find that pretty much linear looking drift of say 4 ppm over 50 days his plot shows initially.