Author Topic: ADR1399 reference  (Read 292746 times)
Offline Kleinstein
Re: ADR1399 reference
« Reply #525 on: August 26, 2026, 08:37:19 am »
A seprate generate 5 V reference could cause an issue if the reference voltage changes. So far the reports for the ADR1399 are more on the lower half for the voltage.
The graph for the 1 PLC 0 V measurement looks like quite some values are off to both the positive and negative side. So a histogram that does not looks like it would have 3 peaks. The extra noise with 1 PLC compared to 10 PLC points to an issue with the auxiliaray (µC internal) ADC.
So maybe the auxiliary ADC scale is not adjusted with the normal DC cal.
The run-up part allows for 4 cases (2 each for the input and zero reading). A scale factor for the auxiliarx ADC that is off would separate the cases.
One may get an even simpler picture from the 1 PLC non AZ mode for a histogram.
Offline iMo
  • Country: li
« Reply #526 on: August 26, 2026, 09:06:20 am »
..in my 34401P the 5V_Vcc is 4.902V and the voltage at the +5V_REF divider before the 706 is 4.931V (I have not measured DC at the MCU's +5V_REF as it is jumping there a lot). With the original 399 it would be aprox 4.96V.
In the 34401 the +5V_REF "follows" the x399 voltage..
I have not seen a difference here in the noisiness before/after "adjusting" the 10V range when replaced the x399.
With the constant +5V_REF in case of the 34970 the AUX_ADC readings might be say -XXmV (? how much ?) off when the HP399 is replaced with a 1399 (assuming typically the 399=7V and 1399=6.9V).
@Alex: did you try to always adjust the 1V range you are using after the x399 replacements?
PS: the good message might be you may fix that potential issue simply with a 100ohm resistor as it is done in the 34401, thus make the AUX_ADC "ratiometric"..  :D
« Last Edit: August 26, 2026, 10:12:59 am by iMo »
Offline Alex Nikitin
  • Femtoampnut and Tapehead.
    • A.N.T. Audio
« Reply #527 on: August 26, 2026, 12:55:06 pm »
Yes, it appears there is at least some correlation with the reference voltage from my measurements. I did calibrate the DCV after two of the replacements, with no obvious effect on the noise though. Need to look at this again, probably in a little while as I don't have time right now  :( .
Cheers
Alex
Offline Kleinstein
« Reply #528 on: August 26, 2026, 03:02:28 pm »
It may need and extra calibration step to adjust the scale factor for the µC internal ADC. There is the extra step for the +10 and -10  V linearity correction. There is a way to read the parameter:
https://www.eevblog.com/forum/testgear/hp-agilent-34401a-hidden-menu/msg4636828/#msg4636828
But I don't know which way to get it adjusted - a kind of factory calibration call.
With the µC ref. set from a separate zener, one could just trim that voltage the analog way, to keep the same ratio to the main reference or to get minimal noise for the 1PLC and faster modes.
Offline iMo
  • Country: li
« Reply #529 on: August 26, 2026, 03:48:39 pm »
I did the "special -10V" adjustment (the step needed after replacing some ADC parts or messing with cal memory) couple of times, hoping it makes the "34401P anomaly" smaller (SM p.82). It did not help.
It is described in the service manual p.85 for 34970 and it is more complex than with 34401, because of the multiplexer.
« Last Edit: August 26, 2026, 04:07:17 pm by iMo »
    • A.N.T. Audio
    • A.N.T. Audio
« Reply #531 on: August 29, 2026, 08:24:18 pm »
Now I have to measure this lot   8)
Cheers
Alex
The following users thanked this post: exe
Offline exe
  • self-educated hobbyist
« Reply #532 on: August 29, 2026, 10:42:24 pm »
Where did you get it? And, sorry for derailing the thread a litle bit, is there a cheap source for adr1399/lm399? I couldn't find much( I think the absolute best price I found is lm399 for like 20euro/piece.
    • A.N.T. Audio
« Reply #533 on: August 29, 2026, 10:55:33 pm »
A fortunate eBay BIN. I was pretty convinced I’m buying a lot of Chinese pulls with welded legs, they certainly looked like it on the photos  :palm: , however the price was irresistible, I better not to say it.
Received the lot today, looks completely legitimate, measures too, two date codes, both from 2015 (if my memory is right, in 2015 the LM399H was priced around $10 a piece. My only worry now is these are rejects from someone’s selection process. We shall see, I’m measuring first chips right now.
Cheers
Alex
« Last Edit: August 29, 2026, 10:57:34 pm by Alex Nikitin »
Offline iMo
  • Country: li
« Reply #534 on: August 30, 2026, 06:30:46 am »
..do wire 16 or 20 of them in parallel and you will get a much better reference than a single 1399 (except the 0.5A heater)..
Everything else would be wasting the time, imho..
 :D
    • A.N.T. Audio
« Reply #535 on: August 30, 2026, 09:34:20 am »
Everything else would be wasting the time, imho…
That thought DID cross my mind ! It is not so much 400mA continuous, but 3A start-up current that bothers me. In any case I need to measure all chips and see if there are any really horrible ones.
Cheers
Alex
Offline iMo
  • Country: li
« Reply #536 on: August 30, 2026, 10:33:10 am »
..you may limit the startup current, from my experience with limiting that current (ie some 30-40mA w/ the 723 w/ "a fold back" I did) it takes much longer the 399 heats-up fully, like several minutes..
So perhaps 4-5 groups, each group powered from a dedicated voltage regulator set to, say, 100-150mA limit..
PS: the 78L15 has 100mA DS current limit (in reality it allows a bit more, like 150mA for some time till overheats)..
« Last Edit: August 30, 2026, 11:12:33 am by iMo »
    • A.N.T. Audio
« Reply #537 on: August 30, 2026, 12:41:50 pm »
I am thinking about switching the heaters in groups, that might be more efficient and faster to reach the right temperature.
Cheers
Alex
Offline bsw_m
  • Country: by
« Reply #538 on: August 30, 2026, 03:59:41 pm »
Some ADR1399 for experiments )
The following users thanked this post: d-el
Offline exe
« Reply #539 on: August 31, 2026, 10:08:01 am »
Meanwhile, I closed the window in the lab. This further improved the noise.
Now, according to AI, with linear temperature compensation, the voltage is within +-1uV about >94% of time. Though there is only one outlier about 6 days ago.
I tried to analyze the outlier, but I can't find any correlation with environmental factors. It does look unusual. Like, I can see what is "normal" noise, and normal drift, that dip doesn't fit it. It also could be that temperature correction has its limits. I think I should start a separate thread for that.
Offline iMo
  • Country: li
« Reply #540 on: August 31, 2026, 12:38:49 pm »
..an occasional jump +/- 1-2uV up or down is normal with 1399, afaik. What outlier do you mean? How do you make the temperature compensation?
Offline exe
« Reply #541 on: August 31, 2026, 06:31:18 pm »
On the plot there are two curves: raw voltage, and voltage with applied ~1.9 uV/C correction that I called "temperature compensation".
By outlier I meant that dip that I marked on plot attached.
Offline dietert1
  • Country: br
    • CADT Homepage
« Reply #542 on: August 31, 2026, 07:09:36 pm »
It's easy to get a 1 uV deviation from thermal EMF, even when using low thermal EMF connectors and copper wires. In the same sense: There are temperature effects proportional to the speed of temperature change.
Trying to work at < 1 ppm requires a lot of patience, really.
Our ADR1399 is a double oven. Maybe one day in the future i will replace its ADR1399 evaluation module by a 6x ADR1399 array - parts are there. I once made a 6x LM399 array with good results.
Regards, Dieter