When constructing parallelized LM399 arrays, is it good to separate the array into small subunits (e.g. 8 = 4*2) and have each unit be buffered by an extra op-amp before the average voltage is given to a final gain op amp?
Or is directly feeding the 8 LM399 voltages into one op amp sufficient (and have that op amp, with a transistor drive stage, give the current feedback to the LM399s)
My suggestion,
use 8 independant units with own output buffer and isolated power supply.
The heater currents are rather large (and change with environment temperature) and may influence the output voltage depending on routing of traces.
Star grounding is much easier with completely independant units.
Each unit has floating supply with only one connection between heater GND and zener GND.
Other possibility would be to use negative voltage for heater supply to get the GND currents away from precision GND.
If one unit begins to drift due to some reason you can also exchange it more easily.
Note: zener current is too high with currently used resistors.
The output voltage pulled down by 200 Ohms may not reach 10V with only 15V supply.
I would place a small capacitor between Pin1 + Pin2 of the Op-Amp U6A to make it more stable for capacitive loads.
And: LM399 with 6860 mV have the lowest T.C. according to experience of some users here in the forum.
with best regards
Andreas