I'm striving to build a practical, inexpensive 10.000000V DC reference for my Kelvin Varley Divider Calibrator project (ala poor man's Fluke 732A/752B) for my Mini Metrology Lab
My thanks to Conrad Hoffman for the inspiration :
http://conradhoffman.com/mini_metro_lab.htmlSeveral Vref ICs are being evaluated, this weekend's project focuses on a less than $30 USD eBay LM399H based / LT1001 buffered multi Voltage stable reference module which arrived last week. Preliminary testing showed it to be very stable, but also very inaccurate (no surprise with 5% variability of actual output Voltage between samples)
This Asian version's design opted for a (not) precision fixed resistance divider scheme to supply any single output, jumper selected from 2.5, 5.0, 7.5, 10.0. The actual individual "calibration" chart showed all sorts of jumper selected values,but very few zeros present. Quite useless for my 10.00000 + KVD project. (see Fig-1)
9.85773V (/=) 10.00000V
The other V selections held no interest to my project so the jumpers were set for 10V and the plan to add a trimmer to get the accuracy needed began. Basically I want to duplicate the "Portable Calibrator" from the National Semiconductors data sheet for the LM399 series. (see Fig-6)
Pictures of the hack's first crude Proof-of-Concept (PoC) test, data sheets and the module's jumper schedule located at:
http://www.qsl.net/k/k0ff/VOLT_NUTS/LM399H-LT1001_10.00_Vref/Updates will arrive in that same folder and eventually a project PDF.
Before anything else, all the unneeded resistors and jumpers will be removed and grounding paths cleaned up as required. Once 10.0 V is achieved, all the other desired Vrefs will be selected by the KVD.
The needed R7 resistance turned out to be 19.182K Ohms (vs. the original fixed 20K), the next step will be to build the trimmer as a low resistance precision 10-turn panel potentiometer ( I like the Spectrol 10T 1K WW and Beckman 15 turn counter for this) with a stable ~18.5K main resistor, and box the whole thing for the next round of experiments and testing.
Have fun.
George Dowell