Tried making some adjustments to make the waveforms easier to read. First, I recorded rundown waveforms from 0.002 PLC to 10 PLC. For a specific board, the waveform consistency is very high, with no significant differences visible. Then a basic comparison between Board #1 and Board #2 is conducted at 0.01PLC.
The main difference lies in the -S1 stage. From the slope_Amp waveform, it can be seen that Board #2 is relatively flat, while Board #1 has a slight rising slope, though not very steep. Then, I removed the 1M resistor connected in parallel with the 24K on Board #1. Under this condition, Error 114 is reported 20-30 seconds after startup. At this point, the waveform of Board #1 shows a steeper slope in the -S1 stage (still need to look at slope_Amp).
Here's the counterintuitive part: Board #2's relatively flat -S1 stage seems somewhat unreasonable, yet it can run continuously for over ten hours without any Error 114. On the other hand, Board #1 (after removing the 1M resistor) has a rising slope in -S1 that looks more reasonable, but it can't last 30 seconds before throwing Error 114.
Can we find something else important from these charts?