Author Topic: Ultra Precision Reference LTZ1000  (Read 2136971 times)
0 Members and 128 Guests are viewing this topic.
Offline Andreas
Re: Ultra Precision Reference LTZ1000
« Reply #3175 on: March 04, 2022, 08:06:16 am »
Hello,
with a dual supplied OP-Amp or a OP-amp which can go below 0.5V at the output there is always risk to forward bias the substrate diodes in the LTZ1000. I would not want to remove this diode.
with best regards
Andreas
Offline julian1
« Reply #3176 on: March 04, 2022, 10:28:22 pm »
Hello,
with best regards
Andreas
Thanks, I think I agree.
Offline RoadRunner
« Reply #3177 on: March 07, 2022, 02:34:11 pm »
Hello to all,
I am looking to make a reference board with LTZ1000. I wonder where do people find these impossible to find resistors?
If any one already have these resistors for sale, I would love to buy them.
Is there a group buy planned? Are there other people looking for them? May be we can plan a group buy.
Regards
Offline Olivia
« Reply #3178 on: March 07, 2022, 03:57:28 pm »
Currently cou could get the LTZ1000CH#PBF from mouser.com, but not the 1000A
If we could organize a group buy for the Resistors and/or for the 1000A (once it is availible again from analog.com or other vendors),
I would very much be interested to participate.
Edit: The resistors are mostly from http://www.vishaypg.com/foil-resistors/hermetically-sealed/ for the hermetically sealed ones.
And yes the ordering from Vishay is not the simplest. I haven´t placed a order with them yet.
Edit2: The datasheet specifies a 13k/1k divider, does anyone know the resulting temperature from the upper resistor across multiple values?
Ole
« Last Edit: March 07, 2022, 04:15:01 pm by Ole »
*record scratch noise* Hey, you.
Yes, you. Have an awesome day!
Offline maat
« Reply #3179 on: March 07, 2022, 06:48:27 pm »
In Germany the Vishay distributor is BADER GmbH & Co. KG (https://www.bader.net/). Send them an email, be prepared for hefty prices and long wait times (14 Weeks or so).
Online Kleinstein
« Reply #3180 on: March 07, 2022, 07:46:20 pm »
At least for the first test one does not need the absolute highest stability resistors.  E.g. Mouser has 1 K and 13 K resistors in the S102 version (plastic case metal foil).
Some of the divider ratios can also be obtained with resistor networks: e.g. 1:12.1 from a dual 10K/1K. This is at the lower temperature end, but may still be OK for the non A version. A smaller resistor for R2 can shift the temperature up a little if needed.
The 8 or more equal resistor networks can also do some rations like 1:12 and 1:15 and 1:14 (AFAIR) and maybe a fre more.
Offline martinr33
« Reply #3181 on: March 07, 2022, 08:15:31 pm »
I have had good results from these folks:
http://webdirect.texascomponents.com/SearchResults.asp?Cat=1842
You have to call / email to get quotes, and they do have a small premium for single parts. They are a Vishay licensee. They also used to have VHP devices, but they do not show them on their web page which means they could have a supply problem. Also, if you call them, they may have parts not on the web site. Generally, I have had them ship in a couple of weeks.
Also:
https://www.ebay.com/itm/121777259014?hash=item1c5a7d8206:g:x8wAAOSwVL1WD3Em
for example.
Offline Andreas
« Reply #3182 on: March 07, 2022, 09:25:03 pm »
Hello,
according to the datasheet:
"With the values given in the applications, temperature is normally 60°C."
From there you can extrapolate the resistors with the typical -2mV/deg C of the base emitter diode.
with best regards
Andreas
Offline branadic
  • www.qoool-sensing.org
« Reply #3183 on: March 07, 2022, 09:39:35 pm »
And even more ratios: https://xdevs.com/guide/rnet_ratio/ just pick the one you need.
-branadic-
Measuring is like guessing, but more advanced.
The following users thanked this post: MiDi
Offline Olivia
« Reply #3184 on: March 07, 2022, 10:42:48 pm »
I would presume that the average voltage across R4 and R5 would be 7,7V which would give UR5 with 13k on R4 a value of 550mV approximately.
Since I don´t yet know the normal UBE of Q2 nor the normal temperature, I am gonna presume a temperature of 20°C and 0.65V.
My target oven temperature is between 40°C and 50°C which would mean that the UBE of Q2 would be at 0.6V so a 1k/12k Divider should do the job.
Although I might want to add a spot for an SMD resistor besides each part of the divider os that I can adjust the temperature by addding higher resistance values in paralell.
Ole
Offline julian1
« Reply #3185 on: March 08, 2022, 03:12:57 am »
Are there any reports of improvement in low frequency noise (current leakage, da, acoustic pickup) with film versus c0g/x7r  - for the 22nF or 100nF compensation cap (if used)?
« Last Edit: March 08, 2022, 03:19:20 am by julian1 »
Offline Andreas
« Reply #3186 on: March 08, 2022, 05:16:27 am »
No the output voltage is 7.2 V per data sheet.
by the way I published all operating points of some of my references here:
https://www.eevblog.com/forum/metrology/ultra-precision-reference-ltz1000/msg874560/#msg874560
the 0.53x V give around 52 deg C heater temperature which is enough for about 40 deg C environment temperature.
Note that the "environment temperature" of the reference internal of a housing is higher than room temperature.
with best regards
Andreas
Online Kleinstein
« Reply #3187 on: March 08, 2022, 07:06:57 am »
The usual choice is using film caps - this was at least before higher capacitance C0G were available. The reference is static and dielectric absorbtion would have essentially no effect.  By now the price for C0G caps even with 100 nF is at an acceptable level and there is no real need to take the risk with X7R or similar with possible acoustic pickup and leakage.   I don't think X7R was ever a good choice.
Offline Olivia
« Reply #3188 on: March 20, 2022, 12:44:40 pm »
So I´ve worked on a version of the LTZ1000(A) Reference Voltage Circuit that gives +10V and -10V as output.
Most of the circuit originates from the datasheet, I just added the ouput stage.
Edit: The added file is for the Falstad Simulator, I hope that is works
« Last Edit: March 20, 2022, 12:54:31 pm by Ole »
Online Kleinstein
« Reply #3189 on: March 20, 2022, 01:05:20 pm »
That circuit looks horrible:
The compensation for the temperature regulation is wrong in the main reference part.
The added part to generate the +-10 V is made would have poor performance. There is a good intention to get some kelvin sensing, but the difference ampfier would just add way too much drift, making things only worse.
The more practical way is to avoid the extra stage and use some compensation of the ground current to get a good apprixmation to kelvin sensing. Depending on the rest of the circuit / supply, one could also just have a seprate ground for the supply and output side, meeting at a star ground point near the LTZ1000 ground.
Offline Andreas
« Reply #3190 on: March 20, 2022, 03:04:52 pm »
Hello,
the scaling resistors are rather high-ohmic.
Depending on OP-Amp you may get high current noise.
No further comments as the schematic is near un-readable.
with best regards
Andreas
The following users thanked this post: TiN, Dr. Frank
Offline Olivia
« Reply #3191 on: June 10, 2022, 06:34:01 am »
I´ve got a question about R2 and R3:
Can the values of these resistors theoretically be reduced to approx. 50kOhms?
Also, I´ve heard somewhere that R2 and R3 can theoretically be 5ppm/°C as well.
Is that an option or rather not?
Online Kleinstein
« Reply #3192 on: June 10, 2022, 07:27:14 am »
A different value for R2 and R3 is definitely an option. The data-sheet already has circuits with 30 K and 50 K as examples. There is also no real need to have the same value and type for R2 and R3 - the main reason is a simpler BOM. More current at the transistors gives slightly less temperature effect from the transistors (which is not good), but with only 50 K the differente is minute ( some 1% for the temperature sens and some 3 ppm/K for the unheated TC). On the upside the noise is marginally lower. A somewhat higher values (e.g. 100 K) would work as well, but chances are that there is not much difference. With a much different value (e.g. 300 K or 20 K ) one may have to adjust some of the capacitors too.
5 ppm/C, 10 ppm/C and maybe even 25 ppm/C would still be acceptable for the resistors. The relevant question is more the long term stability and a low TC is used as a overall parameter to judge resistor quality, as specs for the long term drift are rare and vague. The drift is attenuated by a factor of some 100 to 500. So 10 ppm/C for the resistor would lead to some 0.02 to 0.1 ppm/C for the voltage and thus likely only a small contribution. If the same resistor series is available in 2, 5 and 10 ppm/K grades it may not make a big difference to use the best selection there.
Offline Olivia
« Reply #3193 on: June 10, 2022, 08:15:10 am »
The reason why I would differ from the ususal 70k value is because there are 50k SMD resistors with a near zero tempco,
and those are in stock. Theoretically one could place two resistors in series but that could introduce thermal EMF.
Offline Andreas
« Reply #3194 on: June 10, 2022, 06:35:12 pm »
Hello,
50k is possible.
Keep in mind that this will increase the sensitivity of setpoint temperature resistor ratio changes (1K/12K5) by about 10%.
So about 1.1 ppm / 100 ppm instead of 1 ppm/100 ppm ratio change.
Measured values should be somewhere at the beginning of this thread when I made measurements on my LTZ1000A #1 and #2.
The 70K resistor changes have low influence on overall T.C. so put your money mainly on the setpoint temperature resistor ratio.
with best regards
Andreas
edit: see here:
https://www.eevblog.com/forum/metrology/ultra-precision-reference-ltz1000/msg436981/#msg436981
« Last Edit: June 10, 2022, 08:32:36 pm by Andreas »
The following users thanked this post: Dr. Frank
Offline cellularmitosis
« Reply #3195 on: August 13, 2022, 10:08:14 am »
If we look at just the refamp portion of the 7V LTZ circuit, we have this:
This got me thinking, could we just add one additional resistor to produce 10V?  Has anyone tried this?
LTZs: KX FX MX CX PX Frank A9 QX
Online dietert1
  • Country: br
    • CADT Homepage
« Reply #3196 on: August 13, 2022, 10:22:17 am »
The proposed circuit will work, yet it will be useless.
Ube of the transistor adds a -2 mV/K temperature coefficient. This is to compensate the TC of the zener, so it vanishes (mostly). Now you multiply that by 600R/120R = 5 and there will be a TC of -10 mV/K in 10 V = -1000 ppm/K.
Now you can argue the LTZ1000 has a near perfect oven, but still this won't work well. Unless you  find a way to compensate the TC once more, maybe using the temperature sensor inside the LTZ1000.
Regards, Dieter
The following users thanked this post: cellularmitosis
Offline KT88
« Reply #3197 on: August 13, 2022, 01:42:24 pm »
No. Vbe is changing slightly with temperature which results in a change of Iz. There is one resistance value where the temperature drift will be compensated. You find this application example in the DS for an unheated LTZ1000. If the resistance is increased the over all drift worsons…
Also with the heated device there will be slight changes in the zener current which will be exaggerated by the additional resistance.
Online dietert1
  • Country: br
    • CADT Homepage
« Reply #3198 on: August 13, 2022, 02:35:44 pm »
No, the resistor value for fine tuning TC is around 10 or at most 20 Ohms.
The intention of the proposed circuit was to raise the reference output to 10 V.
Regards, Dieter
Offline KT88
« Reply #3199 on: August 13, 2022, 02:58:21 pm »
Increasing the resistor further than 20ohms would overcompensate the drift. You would need about 600 Ohms at 5mA Iz to get to 10V. That would mean you are overcompensating by 30 - 60x…