OP's requirements are a bit unclear, if it's to measure the temperature (surface temperature I suppose?) of some object rather than ambient temp, then the sensor will need to be thermally coupled to the object and that somewhat limits your options - for instance, that is likely making it inconvenient to use a small SMD sensor, that are available with a variety of interfaces (mostly I2C though) and accuracy, and would otherwise be the easiest.
You need to also define what distance there needs to be between the sensor and device that reads it, if you can make your own PCB, etc.
If you were considering NTCs, and there is some distance between the sensor and reading board, I would rather recommend using RTD sensors (like Pt100), which are much more accurate without requiring much calibration, but require a proper front-end. And they have a response curve that is not linear, but it's actually standard and coefficients are defined per a standard, and if your front-end is decent, they may not require calibration or maybe just the offset. In your case, a Pt1000 sensor is probably more adapted. While they require a dedicated front-end, the benefit is that the front-end can be at some distance from the sensor itself, and so the front-end won't be affected by the sensed temperature, making things also more accurate / not requiring fancy compensation. Just a thought. I have used Pt100 sensors with a relatively simple front-end, using a ADS1118 ADC and applying the response curve from the standard in software, and even without any kind of extra calibration, I was getting an error within +/-1°C. That was the least painful experience I have had with temperature sensing as far as I'm concerned, YMMV.