Scope Triggering/Sampling Question
« on: October 15, 2025, 10:58:54 pm »
I have an Agilent 54831 scope and always seem to have issues with triggering/sampling that I don't understand.

Sometimes triggering issues can be attributed to auto trigger, but often times regardless of how I trigger, the waveform is always jumping.

Regardless, earlier I pieced together a simple circuit to explain a bridge rectifier circuit. It was just a function generator feeding 60Hz to a 1N4004, a 50k resistor, and a 220uF. The input and half wave signals looked good, I added the cap, selected AC on the scope, and expanded. The saw tooth looked horrible, full of noise.

I went into the menus and found 'sampling mode' was set to 'normal'. After I changed it to "high resolution' the wave seemed to look normal.

I also tinkered with memory depth and other stuff, but none of those settings seemed to make a difference.

As for the sampling mode, I would think 'normal' would have looked like a cleaner waveform some sort of decent waveform than 'high resolution', but apparently I'm wrong.

Which settings should I pay attention to in order to get cleaner looking waveforms?
« Last Edit: October 15, 2025, 11:03:40 pm by bostonman »
 

Online aeg

  • Frequent Contributor
  • **
  • Posts: 501
  • Country: us
Re: Scope Triggering/Sampling Question
« Reply #1 on: October 15, 2025, 11:14:08 pm »
Trigger source is line, it's not coherent to the function generator output. Trigger source should be something with a fixed phase relation to the signal of interest. I'm not sure why the high resolution traces look OK, maybe the mains cycle drifted into sync with the function generator while you were in the menus.
 

Online aeg

  • Frequent Contributor
  • **
  • Posts: 501
  • Country: us
Re: Scope Triggering/Sampling Question
« Reply #2 on: October 15, 2025, 11:19:17 pm »
Hm, scratch that, the channel 2 amplitude in normal mode is also implausably high. Maybe your scope is broken?
 

Offline Someone

  • Super Contributor
  • ***
  • Posts: 6124
  • Country: au
    • send complaints here
Re: Scope Triggering/Sampling Question
« Reply #3 on: October 15, 2025, 11:50:18 pm »
Hm, scratch that, the channel 2 amplitude in normal mode is also implausably high. Maybe your scope is broken?
Amplitudes seem to be excessively low??? 200mV/div
Pretty sure those scopes had the ability to scale probes.

Are those pdfs showing just an image? Use the forums image handling to make it easier to view/quote/comment on.
 

Offline bostonmanTopic starter

  • Super Contributor
  • ***
  • Posts: 2848
  • Country: us
Re: Scope Triggering/Sampling Question
« Reply #4 on: October 16, 2025, 12:56:36 am »
Quote
Amplitudes seem to be excessively low??? 200mV/div

I'll need to check, one probe should have automatically have set for x10, but I was more focused on trying to trigger the scope along with why I couldn't see a nice saw tooth.

Triggering on 'line' was while playing with the trigger because the wave wasn't remaining still. Sometimes I see waveforms jumping and try triggering at all points and can't seem to lock it. That's when I turn off auto trigger, set to line, etc...

The more confusing part is the blurry saw tooth that cleared up after I messed with the sampling mode. I'm not use to such scopes; most scopes I've used were always relatively low cost due to cheap companies not wanting to pay for expensive ones.

Now I'm using one that (I consider) pretty good and it seems very touchy. I would think higher sampling rate would improve the wave forms, but I'm not sure what's going on. I have two of the same models and in the past I've tried all four channels along with the other scope with similar results.
 

Offline Someone

  • Super Contributor
  • ***
  • Posts: 6124
  • Country: au
    • send complaints here
Re: Scope Triggering/Sampling Question
« Reply #5 on: October 16, 2025, 02:02:31 am »
The more confusing part is the blurry saw tooth that cleared up after I messed with the sampling mode.
If you show some examples of the problems that radically change with small differences then that would help (not downloading and opening up n pdf files and trying to link them back to your text).

But perhaps first get a handle on selecting an appropriate trigger, trigger level, on obvious waveforms before working through the particulars of sampling/decimation/aliasing etc.
https://www.tek.com/en/documents/primer/oscilloscope-systems-and-controls
 

Offline moffy

  • Super Contributor
  • ***
  • Posts: 3088
  • Country: au
Re: Scope Triggering/Sampling Question
« Reply #6 on: October 16, 2025, 05:21:38 am »
I had similar problems when using my signal generator, an FY6600, with its original power supply, it was generating up to 80V AC on its -ve terminal wrt ground on the scope. I had to upgrade the PSU in the unit to eliminate the noise, now it's just fine.
 

Offline StillTrying

  • Super Contributor
  • ***
  • Posts: 2892
  • Country: se
  • Country: Broken Britain
Re: Scope Triggering/Sampling Question
« Reply #7 on: October 16, 2025, 07:10:23 am »
Are you sure the trigger is on the clean mid level of the sine, and not the 3 mV of noise on the sawtooth?
« Last Edit: October 16, 2025, 08:55:36 am by StillTrying »
.  That took much longer than I thought it would.
 

Offline bostonmanTopic starter

  • Super Contributor
  • ***
  • Posts: 2848
  • Country: us
Re: Scope Triggering/Sampling Question
« Reply #8 on: October 16, 2025, 01:20:33 pm »
Quote
Are those pdfs showing just an image? Use the forums image handling to make it easier to view/quote/comment on.

The reason I did PDFs (ignoring that my scope either prints to a default virtual PDF printer or saves as an image) is that it includes all the scope settings allowing anyone deciphering the waveform to see exactly how the scope was setup.

Quote
I had similar problems when using my signal generator, an FY6600, with its original power supply, it was generating up to 80V AC on its -ve terminal wrt ground on the scope. I had to upgrade the PSU in the unit to eliminate the noise, now it's just fine.

I didn't consider noise from the generator, however, recently I measured a full wave circuit in a stereo amp going through a diff probe, had a good ground connection, etc... and got the typical noise (see attached).

I don't remember if I had scaled the v/div correctly in the attachment because I was going through the diff probe, but I was paying more attention to how the waveform looked.

As mentioned, I'm so use to low grade scopes that produce somewhat pencil thick waveforms that using this higher end model raises questions on what I'm doing wrong. Since all four channels and two same model scopes produce similar looking waveforms, I somewhat ruled out a broken scope.

Typically triggering issues in one case may be not using another channel to trigger on a data bus, then another time it's that I triggered correctly, but then the wave decides to jump phase for some reason, now it was changing the sample mode to high resolution gave me a waveform I am familiar with.

Seems I haven't provided enough information, so let me take more measurements maybe this weekend and I'll post them. It's probably easier to examine this question with the scope present rather than a single image, so I understand the difficulty in assessing the images.
 

Offline makebit

  • Regular Restricted
  • *
  • Posts: 61
  • Country: gb
Re: Scope Triggering/Sampling Question
« Reply #9 on: October 16, 2025, 04:31:55 pm »
I wonder if some previous user to yourself was using it extensively with the  earth cut out of the plug and wrecked it, such as it now has your said problem.
Hi, I'm also known as Treez, Faringdon, 5ubtle, ocset, zenerbjt opampsmoker, and coffeefet. I've been a naughty boy, so I'm not allowed to start new threads any more.
I'm also not allowed to change this footer :-(
 

Offline bostonmanTopic starter

  • Super Contributor
  • ***
  • Posts: 2848
  • Country: us
Re: Scope Triggering/Sampling Question
« Reply #10 on: October 17, 2025, 02:53:50 am »
I have a few more captures, but posting two for now.

The grounding statement in the previous message gave me the idea to measure the ground wire on the probe to check whether it's open. I have different probes, one being an Agilent probe that automatically puts the scope in x10 mode; and the one I use for my main probe. It's in great shape, but oddly enough, measures approx. 2.1ohms from the ground lead to BNC shell whereas my other probes measure extremely low (approx. 0.3ohms).

One reason for only posting a few of my captures is: I measured a (semi good) 9V battery and got more noise on channel 1 than channel 2 when put on AC coupling; something I just realized. Before posting, I want to check whether channel 1 has more noise.

The measurements in the attached two looked the same (or very similar) on my other exact model Agilent scope (they were bought from different people a few years apart, so not one person could have applied similar damaging measuring tactics - or most likely didn't).

Also, I realized my generator was set for 50ohms output impedance which took me a minute to figure out why I initially got higher than 1V RMS, but now it's set for high Z.

Note the measurements in the attachments are in with the scope set to high resolution and not normal. In normal resolution mode the noise is much worse. As mentioned above, the wave forms look the same on both scopes.
 
The following users thanked this post: Someone

Offline bostonmanTopic starter

  • Super Contributor
  • ***
  • Posts: 2848
  • Country: us
Re: Scope Triggering/Sampling Question
« Reply #11 on: October 21, 2025, 12:42:03 am »
I wanted to get a better idea of the noise floor and whether a channel had more noise than another.

One channel had 20MHz BW limit on and the other didn't. Both are enabled, so the noise is much less but saw some slight variation with the probes.

I'm still uncertain why the Agilent probe has approx. 3ohms between the alligator clip lead and the BNC end.

In any case, attached are two images, none of the settings were changed, except the sampling mode (normal versus high resolution). Due to the high value resistor (i.e. low load), it seems the saw tooth waveform isn't as noticeable, so I decreased the frequency and increased amplitude. Now it's much clearer, but I still don't understand what that sampling mode setting is doing.

I would think 'normal' would give a relatively good looking waveform, but, as the images show, it looks horrible in 'normal' mode and far less noise in 'high resolution'.

Does anyone know why this is?
 

Offline Someone

  • Super Contributor
  • ***
  • Posts: 6124
  • Country: au
    • send complaints here
Re: Scope Triggering/Sampling Question
« Reply #12 on: October 21, 2025, 01:01:19 am »
I would think 'normal' would give a relatively good looking waveform, but, as the images show, it looks horrible in 'normal' mode and far less noise in 'high resolution'.

Does anyone know why this is?
Do you not have a copy of the manual for this scope? The same acquisition modes are explained in the manuals for similar scopes (Keysight 3000) if you need to look for others.

Tek have an explainer of their modes:
https://www.tek.com/en/blog/sample-processing-digital-oscilloscope

But the 54830 is old enough that the intensity graded modes are separate?
 

Offline bostonmanTopic starter

  • Super Contributor
  • ***
  • Posts: 2848
  • Country: us
Re: Scope Triggering/Sampling Question
« Reply #13 on: October 21, 2025, 03:06:17 am »
I have a quick start guide, but didn't find a manual.

 

Offline jonpaul

  • Super Contributor
  • ***
  • Posts: 4023
  • Country: fr
  • Analog, magnetics, Pwr high voltage, Audio, Cinema
    • IEEE Spectrum
Re: Scope Triggering/Sampling Question
« Reply #14 on: October 21, 2025, 08:10:38 am »
read Tektronix Circuits Concepts and Measurement concepts book on probes, sampling, trigger etc.

https://w140.com/tekwiki/wiki/Concepts_Series

The 1980s bible.

Have the full printed set.

Enjoy

Jon
The Internet Dinosaur..
passionate about analog electronics since 1950s
 

Offline Someone

  • Super Contributor
  • ***
  • Posts: 6124
  • Country: au
    • send complaints here
Re: Scope Triggering/Sampling Question
« Reply #15 on: October 21, 2025, 08:31:54 am »
I have a quick start guide, but didn't find a manual.
Quickstart guide:
https://www.keysight.com/us/en/assets/9018-04651/quick-start-guides/9018-04651.pdf
Quote
Most of the documentation for using the Infiniium Oscilloscope is
included in its built-in information system. The built-in information
system contains the information generally found in a User’s Guide.
Use the “windows” help.
 

Offline StillTrying

  • Super Contributor
  • ***
  • Posts: 2892
  • Country: se
  • Country: Broken Britain
Re: Scope Triggering/Sampling Question
« Reply #16 on: October 21, 2025, 10:11:49 am »
I think the noise is about normal myself.

Hi Resolution  Real-time boxcar averaging reduces random noise and increases resolution.
Vertical Resolution 8 bits,  ≥12 bits with averaging.


https://www.testequipmenthq.com/datasheets/Agilent-54831B-Datasheet.pdf
.  That took much longer than I thought it would.
 