First time on Keithley 2000
« on: June 29, 2016, 09:48:45 am »
Hi all,

Yesterday I receive my Keithley 2000, I buy it for 170€ shipped, the seller said no errors, so when it arrive I've plug the DMM and no errors on screen, good.

Cosmetic aspect:

- The bottom of the unit has some scratchs.
- It has no bench kit (so the bottom scratchs)
- The front face plate is all yellow
- The red Keithley has some scratchs
- The with 2000 Multimeter has some scratchs
- The display lens has some minor not deep scratchs
- The Local word is almost gone
- The rear cover of the Scan card is missing

Inside aspect:

- Very clean unit
- Very clean PCB
- No signal of soldering that I can see
- No rust

One thing I noted is there is no ATMEL chip inside, so this is what I've inside.

U165 - Orbit 6715A

The Electrolytic Capacitors are:
C104 - H9555 85º
C131 - H9644 105º
C146 - H9633 105º
C149 - H9644 105º
C156 - H9633 85º (The big 6800uF 16V)

The Main Board shows 2000-102-02F, guess this is the revision of the board.

Firmware Rev. A06 A02
Calibration date - 03/20/03
Calibration Date Due - 03/20/04
Count - 4

Pictures will come in shortly, no battery on the camera.

The repair log will be this

1 - Changing the 5 Electrolytic Capacitors for new ones (maybe Panasonic, Chemi-Con KMG series) ???? Any suggestion here.
2 - Cleaning the front plate with hydrogen peroxide (10 Volumes)
3 - Remove the scratchs from the display lens
4 - Buy some SST39SF020A-70 FLASH ROMs to burn the new A19 firmware (the ones I can find end with 70-4I-NHE or 70-4C-NHE) what is good for this?

After this maybe start looking for a Cal lab here in Portugal to send the meter to Calibration and upgrade to A20 firmware.

Pictures link < https://1drv.ms/f/s!ACbdz8jXIGKi3k0 >
« Last Edit: June 29, 2016, 10:26:18 am by Nuno_pt »
Nuno
CT2IRY
 

Offline TiN

  • Super Contributor
  • ***
  • Posts: 4543
  • Country: ua
    • xDevs.com
Re: First time on Keithley 2000
« Reply #1 on: June 29, 2016, 10:29:52 am »
Sounds good unit for decent price.

Your meter made in 1996, so capacitor replacement is very needed step ;).
Chemicon KMGs are ok, I use them often, had no issues so far.
4I ROMs are industrial temperature grade, 4C - commercial. It does not matter in this case for you, either will work.
YouTube | Metrology IRC Chat room | Let's share T&M documentation? Upload! No upload limits for firmwares, photos, files.
 

Online Kjelt

  • Super Contributor
  • ***
  • Posts: 6460
  • Country: nl
Re: First time on Keithley 2000
« Reply #2 on: June 29, 2016, 11:10:47 am »
There is one chip that can become very hot so place a small heatsink on it when you are at it.
Unfortunately I do not remember the number/function but it had to do with the communications :(
 

Offline Nuno_ptTopic starter

  • Frequent Contributor
  • **
  • Posts: 435
  • Country: pt
Re: First time on Keithley 2000
« Reply #3 on: June 29, 2016, 01:15:46 pm »
Thanks TiN for the explaining about the ROM's.
 
Kjelt I'll let TiN see if he knows what chip it is, since he's the Keithley Guru, if it worth the trouble since I'm on it I'll place the heatsink, also on the two IC's between the three black caps, thinking about arrange some way of putting one heatsink for both of the IC's, both connect to an alluminium L and the heatsink on top of the L, if it fits the spare space when the cover is put in.

The first link is update with new photos, one of the display showing some burn on the right side.


I did some cleaning with one product that I use to take the yellow of the plastic headlights from the cars, try it and I can see some good results, put the cream with the finger across all the display, the scrub with a toothbrush, no silkscreen gone or with less color, the buttons was scrub with water and soap and the toothbrush also, some results here https://1drv.ms/f/s!Aibdz8jXIGKi3lb0OYrGbaD-lyGh .


Now to order some:
Chemi-Con KMG caps, all 105º, should I keep the same voltage of the caps or put some high voltage, the originals are 16V, 50V and 63V?

Some LCC32 ROM's http://pt.farnell.com/microchip/sst39sf020a-70-4i-nhe/memory-flash-2mbit-parl-32plcc/dp/1829978?selectedCategoryId=&exaMfpn=true&categoryId=&searchRef=SearchLookAhead&searchView=table&iscrfnonsku=false

One TL866CS version http://www.ebay.com/itm/High-Performance-USB-TL866CS-Universal-Programmer-with-9-Adapters-for-12000-ICs-/121405177087?hash=item1c444ffcff:g:3ygAAOSw-vlVi~W1


One Bench Kit support.

   
« Last Edit: June 29, 2016, 08:21:20 pm by Nuno_pt »
Nuno
CT2IRY
 

Offline TiN

  • Super Contributor
  • ***
  • Posts: 4543
  • Country: ua
    • xDevs.com
Re: First time on Keithley 2000
« Reply #4 on: June 30, 2016, 04:31:16 am »
Newer versions of 2000 (and some other Keithley gear) are using custom ASIC instead of ALTERA CPLD. Functionally both are same.

The hottest chip should be LM399 reference (one with white round plastic cover on it, in center of the board), which is due it's internal 90° oven. It's a bad idea to put heatsink on it, it supposed to be running hot to remove impact from ambient temperature change to reference output.

LDOs are usually around 60-65c in closed case. Everything else is in 50ish region. I had some thermal images of 2000 here.

I like Nuno_pt's way of thinking. Installing larger heatsinks on LDOs might help, but just a bit (since no forced airflow in the unit). Don't combine them or short to each though, as tab's are not always ground, so connecting all tabs together will cause a short!

YouTube | Metrology IRC Chat room | Let's share T&M documentation? Upload! No upload limits for firmwares, photos, files.
 

Online Kjelt

  • Super Contributor
  • ***
  • Posts: 6460
  • Country: nl
Re: First time on Keithley 2000
« Reply #5 on: June 30, 2016, 08:46:09 am »
No it was not the reference chip, it was some communications/glue chip at the end of the board.
I had two of these for work, now at home and one failed on this chip, when replaced we noticed it became rather warm.
Since it was quite expensive ($70 orso incl shipping) we have put a heatsink on it as precaution.

Perhaps I will open it up someday and make a picture.
 

Offline Nuno_ptTopic starter

  • Frequent Contributor
  • **
  • Posts: 435
  • Country: pt
Re: First time on Keithley 2000
« Reply #6 on: June 30, 2016, 01:19:11 pm »
Hi TiN and all,

I've take another look at the 3 LDO's (U110, U124, U125), but on U110 & U125 been so close about 3mm between the two and with the 3 caps (2200uF & 1000uF) so close two of them above and one below, there is no good way to fit some heatsink on both of them, or individually, without some metal work.

With U124 being alone is different, I can take out the small heatsink that he has, and fit an larger heatsink, it's worth the trouble?

Seems that Keithley face the same problem that I'm facing regarding the heatsink's in U110 & U125.

I don't have a thermal cam so can't see if the large heatsink will make some large difference, like TiN said no airflow inside the meter.

Also after TiN suggestion that the tab of the LDO's are not always ground, I've check that with continuity test from another DMM, and none of the tabs of the 3 LDO's beep relative to the 2000 ground of the PCB, so this would require separate heatsinks for U110 & U125.

- For the ROM's I order the 4I version Farnell 1829978.

The caps are same values as the original ones (or I should have look at the next voltage for all?), but I look at some better ones, like more hours at 105º or at 125º, this are the ones I found with stock in EU, basically Chemi-Con doesn't have stock in EU or I would do like TiN and would go to the KMG series road, and the ones on Epay are from unknown provenience.

- For the 6800uF 16V the Farnell P/N are:
2472006 (Vishay 6000h 125º)
1744933 (Pana 5000h 105º)

- For the 2200uF 16V the Farnell P/N are:
8126313 (Rubycon 10000h 105º)
2346454 (Rubycon 10000h 105º)
1800641 (Panasonic 10000h 105º)
2479889 ( Panasonic 10000h 105º)
2079157 ( Panasonic 10000h 105º)

- For the 1000uF 50V the Farnell P/N are:
1855192 (Panasonic 5000h 105º)

- For the 100uF 63V the Farnell P/N are:
2079295 (Panasonic 6000h 105º)
1144638 (Rubycon 7000h 105º)

I've order the Vishay, and the rest all from Panasonic, was this the good choice?
This is to leave some ideas for the new ones repairing the K2000, what caps they could look for.

Let's see when the capacitors arrive, I'll change them, but I'm off on vacations in 1 week time, and only after when I get home will order the TL866CS (that usually takes 15 to 20 days to get here) and burn the new A19 firmware, meanwhile will take another look at the 3 LDO's and the heatsinks.
« Last Edit: June 30, 2016, 02:31:03 pm by Nuno_pt »
Nuno
CT2IRY
 

Online Kjelt

  • Super Contributor
  • ***
  • Posts: 6460
  • Country: nl
Re: First time on Keithley 2000
« Reply #7 on: June 30, 2016, 03:45:50 pm »
If I am not mistaken , using a newer firmware will invalidate the calibration correct? What is the gain of using a newer firmware?
 

Offline Nuno_ptTopic starter

  • Frequent Contributor
  • **
  • Posts: 435
  • Country: pt
Re: First time on Keithley 2000
« Reply #8 on: June 30, 2016, 04:53:46 pm »
Kjelt from what I've read in other post's about 2000 you can upgrade firmware till version A19, for the last version witch is A20you need a new calibration since the A20 firmware has one more step in the calibration processe, so if you upgrade the 2000 to A20 and don't perform a new calibration you lose all the data.

I'll let TiN explain what one gain in upgrading to new firmware.
Nuno
CT2IRY
 

Online Kjelt

  • Super Contributor
  • ***
  • Posts: 6460
  • Country: nl
Re: First time on Keithley 2000
« Reply #9 on: June 30, 2016, 04:59:08 pm »
Ah that is cool, I might do that also if I knew what I would gain with the latest firmware ?  :-//
One of mine is A05 02 the other A06 02.
 

Online Kjelt

  • Super Contributor
  • ***
  • Posts: 6460
  • Country: nl
Re: First time on Keithley 2000
« Reply #10 on: June 30, 2016, 05:21:26 pm »
So to come back to that "hot" ic, I opened the Keithley and it is the one in the upper right corner marked U165 (see picture).
My roms are ST 27C2001 that is different from yours or do you want to replace them with flash roms on purpose?
My elco's still look ok, nothing wrong with them, you advise to replace them just in case?
thanks.
 

Offline Nuno_ptTopic starter

  • Frequent Contributor
  • **
  • Posts: 435
  • Country: pt
Re: First time on Keithley 2000
« Reply #11 on: June 30, 2016, 06:06:22 pm »
Kjelt, I want to replace my U156 (EVEN) and U157 (ODD), in PLCC32, that are old EEPROM's style, with the newer FLASH EEPROM's SST39SF020A-70-4I-NHE.

See here on TiN's website https://xdevs.com/fix/kei2000/#firmware.

See also here post #20 https://www.eevblog.com/forum/testgear/keithley-2000-rs-232-wtf/

See also here on the Keithley forum https://forum.tek.com/viewtopic.php?f=32&t=6600

About your Elco caps, see the date code, if they are like my Nichicon '95 and '96 year, the reference is YYWW, the first two are the year of the caps and the last two are the week of that year, so for my caps with 20 years I sure don't want that they leak on the board causing damage.

Look here to what the caps can do https://www.eevblog.com/forum/repair/repair-old-keithley-2000-teardown-and-fix/

My U165 is mark 2000-802A02 ORBIT 6715A - 9629 (Year 96, Week 29) - 9512G476G, instead of the old ALTERA like here on TiN 's board http://dev.xdevs.com/projects/kei2000/repository/entry/photos/k2000_assy.jpg
« Last Edit: June 30, 2016, 08:22:54 pm by Nuno_pt »
Nuno
CT2IRY
 

Online Kjelt

  • Super Contributor
  • ***
  • Posts: 6460
  • Country: nl
Re: First time on Keithley 2000
« Reply #12 on: June 30, 2016, 08:03:07 pm »
Thanks there is a lot to read  :-+
 

Offline Nuno_ptTopic starter

  • Frequent Contributor
  • **
  • Posts: 435
  • Country: pt
Re: First time on Keithley 2000
« Reply #13 on: June 30, 2016, 08:20:21 pm »
No problem Kjelt, lot's of reading here on EEVblog and also on TiN xdevs website (The Keithley Guru) and also on the Tektronix forum.

FYI also:
U156 & U157 are both EEPROM's PLCC32 (this are the ones I will replace)
U151 & U152 are both the RAM's
U136 is an NVRAM

- Page 21 of the repair manual states that:

"Memory circuits
- ROMs U156 and U157 store the firmware code for instrument operation.
U157 stores the D0-D7 bits of each data word, and U156 stores the D8-D15 bits.

- RAMS U151 and U152 provide temporary operating storage.
U152 stores the D0-D7 bits of each data word, and U151 stores the D8-D15 bits.

- Semi-permanent storage facilities include NVRAM U136.
This IC stores such information as instrument setup and calibration constants.
Data transmission from this device is done in a serial fashion."
Nuno
CT2IRY
 
The following users thanked this post: Kjelt

Online Kjelt

  • Super Contributor
  • ***
  • Posts: 6460
  • Country: nl
Re: First time on Keithley 2000
« Reply #14 on: June 30, 2016, 08:22:55 pm »
Ok might be worthwhile to retrieve and store the calibration info as well?
 

Offline Nuno_ptTopic starter

  • Frequent Contributor
  • **
  • Posts: 435
  • Country: pt
Re: First time on Keithley 2000
« Reply #15 on: June 30, 2016, 08:26:30 pm »
That is way I'll by new Flash EEPROM's, and burn them with A19 firmware that is on TiN's website.
I'll save the old ones if something go wrong I'll still have the old EEPROM's with A06, to put the DMM back to work.

I forgot about U165, I'll see if I can measure the temperature in some way, and if I can I'll post the running temperature after leaving the meter on for 1 or 2 hours.
« Last Edit: June 30, 2016, 09:46:30 pm by Nuno_pt »
Nuno
CT2IRY
 

Offline TiN

  • Super Contributor
  • ***
  • Posts: 4543
  • Country: ua
    • xDevs.com
Re: First time on Keithley 2000
« Reply #16 on: July 01, 2016, 04:28:46 am »
That's good idea to save calibration constants from NVRAM U136 (which is simple I2C EEPROM, TL866 supports it, but you will need to carefully desolder it from the board, read and then put back).
This way you can always revert original ROMs/CAL back.

Replacing firmware ROMs with FLASH is just to make firmware upgrade easy and quick, without need of getting new OTP ROM or waiting with UV light to erase UVEPROM.
I'm not quiet familiar with different 2000's version, as I got 2000 just to be complete in Keithley DMM family. My main focus was/is on 2001/2002 meters which are different beasts with different hardware.

Replace capacitors, time wasted for discussion if doing it or not is not worth a risk. Spend 10$ and you good to go.

Quote
U165 is mark 2000-802A02 ORBIT 6715A
As I mentioned, functionally is equally same custom gate-array. Earlier meters use ALTERA CPLD, newer - smaller ASIC. This chip is digital part of multislope integrating ADC with switch control logic, SPI interface, triggering, mux control and edge detectors/counters. Heatsinking it not required, nor helping. :)



YouTube | Metrology IRC Chat room | Let's share T&M documentation? Upload! No upload limits for firmwares, photos, files.
 
The following users thanked this post: Kjelt

Online Kjelt

  • Super Contributor
  • ***
  • Posts: 6460
  • Country: nl
Re: First time on Keithley 2000
« Reply #17 on: July 01, 2016, 07:46:19 am »
Thanks TiN and Nuno_pt, I agree replacing the elcos now is way better than cleaning up in a couple of years time  :)

I wish there was a way to calibrate this meter myself, that is, adjusting it a tiny bit against another calibrated meter.

About the firmware, I see in the schematics that there are many revisions, so the pcb's or layouts have changed a bit over time, is it certain that the latest A19 firmware will run perfectly on the older revision pcb's?
I mean if it are only software bugfixes this would be good, but if it is due to different components or layout we might have trouble with the older revisions/pcb's, is there any documentation available that states which firmwares are the latest for certain pcb versions?

Found the answer: https://www.eevblog.com/forum/repair/keithley-2000-firmware-compatibility/
« Last Edit: July 01, 2016, 07:51:15 am by Kjelt »
 

Offline Nuno_ptTopic starter

  • Frequent Contributor
  • **
  • Posts: 435
  • Country: pt
Re: First time on Keithley 2000
« Reply #18 on: July 01, 2016, 08:44:15 am »
Kjelt, yes it's better replace them now, then have a mess later on, I agree with TiN on this for the value of the order.

2 - 1829978 = 3.52€
2 - 2472006 = 4.16€
2 - 2346454 = 2.10€
2 - 1855192 = 3.92€
2 - 1144638 = 1.47€

Total order with 23%VAT is 18.66€, I've just bought 2 of each.

I'm afraid to become Keithley addicted like TiN :-)

TiN I'll take a closer look at the NVRAM U136, and see if I can desolder it.

TiN I'm also looking at the 2001, but much more expensive, and like you said different beast with very different hardware.

I should have my caps and the flash eeprom's in the beginning of next week, the caps will be replace then but the eprom's only I get back from vacations.


Nuno
CT2IRY
 

Offline TiN

  • Super Contributor
  • ***
  • Posts: 4543
  • Country: ua
    • xDevs.com
Re: First time on Keithley 2000
« Reply #19 on: July 01, 2016, 01:51:44 pm »
Kjelt
With 2000 it's probably worth to get unit fully functional, with latest firmware (A19 or A20), all cleaned, caps replaced and fully assembled, leave it running few days to make sure all is okayish and no self-test errors or any suspicious readings, then take it to local cal facility and ask them nicely to calibrate it. Should be doable for something in 100-150$ range or less, if you lucky. Making standards for your own calibration will be way more expensive (2-10 times than the meter at least).

Nuno_pt
Beware this road, it's dark and full of terrors. You will find addiction to seeing that last 9th digit wobble faster than you expect...
If you patient, it's possible to get bad broken 2001 in 150-200$ range, just keep watching for offerings.
And I'm pretty sure in 95% of broken old 2001 (almost each one with capacitor electrolyte damage) all problems are fixable with enough dedication.
Here is one for example: http://www.ebay.com/itm/Keithley-2001-7-5-digit-digital-multimeter-/152146004454?hash=item236c9bc9e6:g:~WwAAOSwmtJXaE3O with megavolt writing on top  :scared: :wtf:
It's likely to go for sale in 400-500$ region though, as still 2d 20h till auction end. Also official calibration of these are in range of 500$.
YouTube | Metrology IRC Chat room | Let's share T&M documentation? Upload! No upload limits for firmwares, photos, files.
 

Offline Nuno_ptTopic starter

  • Frequent Contributor
  • **
  • Posts: 435
  • Country: pt
Re: First time on Keithley 2000
« Reply #20 on: July 01, 2016, 02:05:48 pm »
TiN I'm looking at it also, shipping is not expensive since it's near, let's see for how much it goes.

I know it's a little narrow and very dark road, leading to 8.5 and praying for 9.5 or 10.5 digits.

Or I can stay on 6.5 and flipping some till I've raise founds to get an 7.5. :-)
Nuno
CT2IRY
 

Online Kjelt

  • Super Contributor
  • ***
  • Posts: 6460
  • Country: nl
Re: First time on Keithley 2000
« Reply #21 on: July 01, 2016, 02:14:47 pm »
Hi TiN, I was not suggesting building or buying my own standard.
I was just thinking on my work they have calibrated meters so if I bring my meter and see what the difference is for instance in the DC volts range, it would be nice if by changing the calibration values in the nvram I could tweak it so it is more accurate. But then I need to know what the exact locations and contents of the NVRAM are, I do not think such a table exists?
 

Offline TiN

  • Super Contributor
  • ***
  • Posts: 4543
  • Country: ua
    • xDevs.com
Re: First time on Keithley 2000
« Reply #22 on: July 01, 2016, 02:33:22 pm »
Quote
praying for 9.5 or 10.5 digits.
Unless you've got 50K$-200K$ for JJA voltage standard and unlimited liquid helium supply (15-30$USD/liter) , prayers will not help much  :popcorn:

Precision gear for hobbyist, is rarely an one time buy tool, it's the way of thinking. When you get habit to move slowly and smoothly in the room to minimize airflow around probes and meters to reduce thermal EMF noise  :o >:D.

Kjelt
The problem is not finding calibrated meter, but transferring it's value into your undertest test meter. You need to have source stable enough to do such a transfer. And to get decent TUR at 6.5 digits, with all those ranges which K2000 wants for calibration - it's usually something big, heavy and expensive like Fluke/Datron calibrators or array of sources. Of course there are darker ways to get something suitable, but that will take years of usual hobby-time to test and validate. So sending 6.5digit meter out for 200$ calibration is cheapest option, unless you have access to calibrator.

Your idea is interesting, but it have caveats down the road.
A. You need to ensure that source does not drift away between your reference measurement and adjustment. It's near impossible with just one calibrated meter to be certain. :)
B. Making table is easy task, most of calibration values are stored in standard floating point format with fixed offsets. You will need just to figure out checksum algorithm and make a map. It will take you a day with hex-editor even without experience. I did this for 2001 while ago.
« Last Edit: July 01, 2016, 02:39:02 pm by TiN »
YouTube | Metrology IRC Chat room | Let's share T&M documentation? Upload! No upload limits for firmwares, photos, files.
 

Offline saturnin

  • Regular Contributor
  • *
  • Posts: 116
  • Country: cz
Re: First time on Keithley 2000
« Reply #23 on: July 01, 2016, 05:13:02 pm »
it would be nice if by changing the calibration values in the nvram I could tweak it so it is more accurate. But then I need to know what the exact locations and contents of the NVRAM are, I do not think such a table exists?

Actually, you can use :cal:prot:data? GPIB command to get calibration data from your K2000 too (if I remember correctly, it does not work with RS-232 interface). I attached dumped cal data from my K2000 unit.

I also think it should be possible to change calibration constants via GPIB too. Why should it be possible? I found out K2000 firmware contains secret commands that were most likely used by Keithley's engineers during development as backdoors (see attached part of K2000 FW - A07 version). By comparing known commands from documentation with strings in FW, I found two groups of secret commands:

1/ :diag:keit: group

:diag:keit:warmboot
-> causes restart of the unit via GPIB

:diag:keit:ehraccess
-> the command is accepted by multimeter, but it freezes, does not respond then and must be restarted (is it some kind of enhanced read access to multimeter memory??)

:diag:keit:isn
-> not sure it is correct form of command, I have not tried it yet

2/ :cal:unlock: group
-> I suspect that commands from this group can be used to define cal constants via GPIB
-> their syntax should be something like:

:cal:unlock:dc:<const_name>:<value>
:cal:unlock:ac:<const_name>:<value>

I have not tried them yet too since my unit has good calibration and I don't want to spoil it if I don't know exactly what I am doing...

!WARNING! Commands mentioned above are not documented in the official documentation, may be inaccurate, misleading and you use them at your own risk  ;)
 

Online Kjelt

  • Super Contributor
  • ***
  • Posts: 6460
  • Country: nl
Re: First time on Keithley 2000
« Reply #24 on: July 06, 2016, 06:42:51 pm »
2 - 1829978 = 3.52€
2 - 2472006 = 4.16€
2 - 2346454 = 2.10€
2 - 1855192 = 3.92€
2 - 1144638 = 1.47€ 
Thanks for the list  :-+
Where did you get these prices?
For instance the 1829978 costs me € 1,96 ex VAT a piece at Farnell.
 


Share me