I saw one of your other posts about drift, and the meter does do weird open drift behavior when it's first powered on, and then continues less dramatically.
For example, when first powered on in DCV, it may show -5V or 5V, and then drop down closer to 0. Right now, once it reaches close to 0 (maybe 0.03V), it takes about 4 seconds to drop to -1.17V, and then back to 0.
400.2 and 305.2 (which is back after a warmup), are both performed open.
305.2 is bias current. My guess is that the state of the drift caused by 305.2 is causing the intermittent 400.2.
Thanks,
Josh