Understood. I left my runners off, and I dislike conformal coatings passionately. Time will tell if I'm wrong.
I'm fairly accident prone by nature and just pulling the cards out over the last couple of days I had to be very careful as they *WILL* clash against each other (with the runners removed).
My logic is that (if you read the webpage on replacing the runners) it was speculation by the 3M app engineer that this tape would work best. I figure conformal coatings (I used silicone) are well proven wrt traces so I doubt very much they will cause an issue and maybe if there is a long term issue with the replacement 3M tape it will help prevent damage.
Of course it took 20+ years for the issue to occur with the original OEM tape so I suspect I'll be well into senility before anything happens but ......
On the 16534A card again...
Besides the previous question on the self-test, are the cal errors associated with one channel or both?
Both. I just removed the runners from the failing board.
It still works (both channels) connected to my Siglent function generator. I tried a variety of waveforms and pulsing. All worked. Maybe slightly noisier than the fully working one.
I just re-ran calibration and didn't get the previously mentioned 'whoa this board has issues are you sure you want to continue' message and this time Trigger Delay passed but ADC, Gain and Offset still fail on both channels.
Same calibration T cable works fine on the other card.
You can probe two cards at the same time by using a 16701B expansion chassis, or a home-made card extender. User DogP has some gerbers available for an extender that works well:
I'd seen this before. I assume by "two cards at same time" one is via removing the unit cover.
I'd like a 16701B. One of my units came with the cable to connect. They're stupid expensive. Max I'd pay is $50 and it would need to be local.
I'm parting out a couple of 16700Bs. I think one has opt-003. I've had zero luck selling them for $100. Just mentioning it in case anyone wants any parts (for spares). I'll probably keep the boards from one for spares for my 16702B. I also have a 16702B that will be for sale (two logic cards (low end) with original runners replaced, all cables/pods, optional pattern gen card) if anyone is interested.