Hi all,
I had posted in Metrology but I have not received any help. I was hoping maybe @TiN can provide some help.
Anyway I hope somebody here has some information.
I have the following error when the device does the self-test:
ERRSTR 202, “HARDWARE FAILURE – SLAVE TEST: DC BOARD”
No other errors reported.
No measurement is being made. Even TRIG didn't trigger a measurement.
We sent the device for service at Keysight, they asked for "only" $7.5k to replace the DC Board and A/D board plus a calibration. I didn't foresee such a bill. So I didn't do the repair. It's such a valuable device to sit as a paper-weight though. Please help me get it running again.
Here are the faults reported by Keysight:
Time out unable to read A/D. Cal value Out of Range "95"
Can anyone who knows this device explain what this means?
I found U2 on the DC board damaged, they relays were not running, after replacing it they are working again but the failure didn't clear.
Please note that this is a device I bought brand new around 2014 and it was never tampered with before this incident.
After injecting 1V through the terminals, I see 1V on both TP10 and TP11 but Q12 is off and it is not passing this voltage to P4. Q12 is controlled by a signal coming from the 8051 uC which is indeed disabling it. Does the system do this because it failed the self test? maybe as a safety feature?
Do you have suggestions?