50/75 Ohm Measurement Offset
«
on:
May 22, 2024, 11:26:10 pm »
Hi All,
If my measuring instrument is designed for a 50 ohm system and I measure a 75 ohm device, what is the correct offset to apply? For example, return loss?
Logged
Wallace Gasiewicz
Super Contributor
Posts: 1234
Country:
Re: 50/75 Ohm Measurement Offset
«
Reply #1 on:
May 23, 2024, 01:59:15 pm »
VSWR into a 75 Ohm load from a 50 Ohm device will yield a 1.5 VSWR
This is used as a quick check by some folks to see if everything is working OK.
1.5 VSWR is 13.979 dB Return Loss which means 96% of power goes into the load
Logged
metrologist
Super Contributor
Posts: 2219
Country:
Re: 50/75 Ohm Measurement Offset
«
Reply #2 on:
May 23, 2024, 04:25:15 pm »
OK, thanks. I was wondering why my instrument manual says a 7.5 dB offset is applied when selecting 75 Ohm impedance.
And an online converter says 1.5 VSWR is 0.18 mismatch loss.
«
Last Edit: May 23, 2024, 04:27:08 pm by metrologist
»
Logged
ConKbot
Super Contributor
Posts: 1394
Re: 50/75 Ohm Measurement Offset
«
Reply #3 on:
May 29, 2024, 06:32:19 pm »
Was it a 7.5 db offset or a 5.7 db offset? If the latter, then it's expecting Min loss pads to be installed.
https://www.keysight.com/us/en/product/11852B/adapter-50-ohm-type-n-f-75-ohm-type-n-m.html
With a good pad, it presents a good match to both ports (see the 27db return loss on that page)
For stuff like an amp, as noted by 1.5:1 mismatch you'll probably be fine without, as if your amp oscillates with 1.5:1 on the input or output, heaven help you in real life. But if evaluating filters, having the right load on the filter can make all the difference.
«
Last Edit: May 29, 2024, 06:42:56 pm by ConKbot
»
Logged
metrologist
Super Contributor
Posts: 2219
Country:
Re: 50/75 Ohm Measurement Offset
«
Reply #4 on:
May 30, 2024, 01:44:25 am »
Thank you for the continued interest.
I learned that the 7.5 dB offset is applied to spectrum analyzer applications for amplitude measurements when the mfg's impedance transformer is used. I think that means that the reference level is offset 7.5 dB due to the mfg specific loss specifications for their impedance transformer device.
There is an "Other" or a simple Reference Level Offset option to achieve the same result.
It just seemed odd to me that the Impedance menu would use some specific offset value without stating what it was.
«
Last Edit: May 30, 2024, 01:47:28 am by metrologist
»
Logged
cncjerry
Supporter
Posts: 1302
Re: 50/75 Ohm Measurement Offset
«
Reply #5 on:
June 03, 2024, 05:39:45 am »
On a 3586 selective level meter when you switch between 50 and 75 ohm, there is an offset of 1.76db applied. They mention it in the 3586 manual. I think the 3586 can add or subtract it automatically. Power is 10 log 75 divided by 50. S0 1.76dB would be .176 and 10^.176 is ~ 1.5. I might have the units upside down.
Jerry
Logged
Print
Search
Pages: [
1
]
Go Up
« previous
next »
Share me
Smf
EEVblog Electronics Community Forum
»
Electronics
»
RF, Microwave, Ham Radio
»
50/75 Ohm Measurement Offset
There was an error while thanking
Thanking...
EEVblog Main Site
EEVblog on Youtube
EEVblog on Twitter
EEVblog on Facebook
EEVblog on Odysee