After repairing my Keithley 2010, I'm not yet impressed with its stability. Though it isn't bad, I don't believe it matches its datasheet specs, so I will be adjusting it (TC=0 Zener diode stability) further before I decide whether or not to experiment with the internal reference.
This made me curious to see what other people are getting for stability with their bench meters. All are welcome to play, but I'm especially interested to see other K2010s, DMM7510s, 34470A, and similar meters. I'd also love to see how other 6.5 digit meters perform, but don't be shy if you have a lower digit meter and want to compare it also.
Let's see how our bench meters perform when comparing stability at different test lengths. Some datasheets claim a specific stability at 2 minutes, so I'd like to compare that along with longer periods. Let's say 2 minutes, 10 minutes, 30 minutes, or any number of hours. It's fine to screen capture the same test run over a period of time. However, if you wish to restart the chart and Grid panel readouts for each test, that shouldn't make a big difference if enough warmup time has already passed.
If you don't already have
@HKJ's awesome free program TestController,
get it!
To perform these tests, there should be substantial warmup time. Let's say a minimum of 4 or 5 hours (including the meters and voltage reference), but more is better. Because I want to rule out any amount of warmup drift, my meters have been warmed up for over 12 hours.
A quality voltage source is required to rule out other sources of noise/instability. For my tests, I'm using a battery-powered VREF10-001 r9 (it appears to be slightly more stable with batteries compared to the external supply). If you have something similar or better, fantastic. Absolute accuracy isn't important; it's the stability that matters in this case.
To chart these comparisons, we should use TestController, and make sure to use similar settings. I will add a post showing some preferred setup options. Ideally, this will include a
chart, along with a "
Grid panel" and "
advanced readouts" for each meter (this is most important). That will show peak-to-peak (
PP) and standard deviation (
SD). If you're running a very old version of TestController I recommend
upgrading before performing these tests. Feel free to include data exports (be sure to pick simple CSV or TSV so it can imported into Excel), but I'm most interested in seeing screenshots with grid panels for quick info comparison.
To get started, I'm sharing a couple 2-minute tests, and a 30-minute test. I'll update this post with more as time elapses.
Meter settings:Range: Manual 10V (or 20V of your meter doesn't have a 10V range).
10NPLC.
1 sample.
No filtering or averaging.
Auto Zero on.
Line sync on if available.
TestController settings:Grid panel with advanced readouts (see below for setup)
Chart with logging set to 1 second
Thanks,
Josh
PS. Noting environmental variables might be helpful too, especially if they're unstable. My office is currently stable at 24C (air is off).