Some DC measurements comparing with DM40C
Voltage reference chip is AD584LH
The DS4T252 has an option called "amplitude fine tuning" to manually calibrate DC values, but I didn't do it because I don't feel the need to. For me the displayed values are accurate enough.
Impressive. I was not expecting that level of precision on these low cost scopes.
So far, the only thing I don't like in the DSTouch DS4T252, it is the case construction.
It seems that it is caseless - They just sandwiched the assemblies and then applied vulcanized rubber around it to form a case.
It seems to be virtually impossible to open this case to peek around inside without deforming or even destroying the rubber case.
After looking to some videos where it has been dismantled, I am not keen to do the same thing on my unit 
Yes that's what I don't like either, to disassemble it we need to unglue the protective rubber, but let's hope we don't need to replace the battery for a long time.
This scope also runs directly from the charger but I've noticed that from time to time it is topping up the battery,
I did some consumption tests when the scope is running from the charger (although I don't recommend because of the increased noise it adds to signals, better to run from battery
The following values may vary depending on what the scope is doing, I tested with a simple sinewave:
Backlight 1/3
1CH, RUN, 540mA, 2.703W
1CH, STOP, 395.4mA, 1.977W
2CH, RUN, 592.6mA, 2.963W
2CH, STOP, 385.2mA, 1.926W
When both channels are enabled the increased consumption is around 260mW (52mA!)
Battery life: 2CH ON, RUN, triggering sweep 1KHz - 1MHz sine wave, after 5 hours the battery indicator was still green, then I had to turn off everything. So with some care it's possible to reach 6 hours or more.
It is able to trigger CAN signals although it cannot decode (yet) for obvious reasons. "Yet" means, maybe they add it.
Post your bugs or dream features you would like to see added to the next firmware release.