Interesting that I couldn't replicate the noise test the same way you did it, can't see any noise on both scopes. Only if I set vertical to 1mV/div on the scope with the probe I'm able to see the "noise" at 1.5mV level
Then i decided to try a different approach instead of a shorted loop, i used a ferrite inductance as an antenna to "amplify" the noises and I'm able to detect more information from the lower middle to the left, on the back you can get a stronger noise detection.
With the little trick of using a "ferrite inductance" you can even "see" smps control chip working signals without having to touch the chip
My probe is just a 100MHz oscilloscope probe set on X1 and shorted by the ground cable to form a small loop, and the oscilloscope was set to X1 as well, with 10mV/div.
Try moving the probe *slowly* around the border of the left lower corner of the DUT as seen in my pictures, where the antenna is the ground cable.
A little deviation and the signal indeed reduces dramatically and fade away on a 10mV/div scope.
My DUT unit is generating in excess of 20mVpp on a single point that I guess is closer to the Touch Screen IC from the front side. This TS chip is installed on the flat cable. This noise amplitude and frequency readings changes with the selected time base on the probing oscilloscope.
If you can't detect it, probably you have a different version of DS4T252 hardware assembly.
My DS4T252 unit measures in excess of 700uVpp of internal noise on 1mV/div, while yours shows quite lower levels of noise.
I wonder if the manufacturer didn't install the shield covers at the front side of the PCB assembly facing the display on some production batches from where my unit came from.
That would partially explain why your units are quieter than mine.
Anyway, I am not planing to open my unit at this time to check the shielding.
when did you bought your scope? I bought both directly from the manufacturer, the first one 15 days ago, and the second 7 days ago. Overall I'm satisfied with them, the 2M memory are really handy. My understanding is that they are still working on them, improving the firmware and correcting bugs.
Are you sure that noise doesn't come from any other device you have connected? like a led bulb?
Are you sure that noise doesn't come from any other device you have connected? like a led bulb?
That is not possible in the test I did because, as seen in the posted pictures, I have no noise picked up until the DS4T252 is powered up.
And the noise is clearly coming from that lower left corner at a very specific place were the probe GND cable was placed.
Clearly your oscilloscopes were built differently from my unit. I find no other explanation.
Are you sure that noise doesn't come from any other device you have connected? like a led bulb?
That is not possible in the test I did because, as seen in the posted pictures, I have no noise picked up until the DS4T252 is powered up.
And the noise is clearly coming from that lower left corner at a very specific place were the probe GND cable was placed.
Clearly your oscilloscopes were built differently from my unit. I find no other explanation.
Did you see the Kerry review on YT?
Please see at 9:22 where he checks the 1mV/div sensitivity.
He got basically the same noise I have on my unit.
Here:
Probably you got one of the first batches, the same as he did, because the new box is smaller, that's my guess. And as I told you I bought directly from Dreamsourcelab.
Probably you got one of the first batches, the same as he did, because the new box is smaller, that's my guess. And as I told you I bought directly from Dreamsourcelab.
Yes, most probably there are at least two different hardware builds of this DS4T252 model.
And that is a good thing, it means they are addressing the issues.
Meanwhile I found some YT videos advertising the DreamSourceLab DSTouch DS4T252, and in one of them someone claims 90uVrms of noise at 1mV/div.
The small letters clearly state that "The noise level data is based on empirical measurements and may vary slightly due to environment factors".
Also they omit in what conditions they got that reading.
Now, those 90uVrms would translate to 90*1.414*2 = 254uVpp. on pure sinusoidal waveforms only. Here we have a complex waveform.
Checking Google AI and it enthusiastically assures me that for noise signals (assuming Gaussian noise distribution) we may multiply the RMS value by 6.6 to get Pk-Pk.
If we try that, we get 90 * 6.6 = 594uVpp, and this is close enough to my readings on the oscilloscope.
OK, to check that, I went to my balcony where there are no external EMF sources close by, switched the DS4T252 on and set it 1mV/div.
Only this time I have chose RMS in addition to Freq and Pk-Pk measurements.
As I was guessing by now, the above claim from someone seems to be valid: My unit never went above 90uVrms in any time base.
And the FPGA/MCU must be doing a good work estimating the RMS value from the Pk-Pk readings (or vice-versa).
For my testing, please see the attached pictures.
- DS4T252_noise_floor_1mVdiv_10mS: Time base at 10mS/div, reading 1.551 MHz, 800.000 uVpp, and 88.549 Vrms.
- DS4T252_noise_floor_1mVdiv_10nS: Time base at 10nS/div, reading 18.578 MHz, 366.667 uVpp, and 77.544 Vrms.
See the advert here at 0:30 playtime:
I'll repeat the test again tonight and see if I can catch anything like you did.
when did you bought your scope? I bought both directly from the manufacturer, the first one 15 days ago, and the second 7 days ago. Overall I'm satisfied with them, the 2M memory are really handy. My understanding is that they are still working on them, improving the firmware and correcting bugs.
Because they told me to report back bugs, improvements, etc so I'm assuming that they're working on it. can't tell you 100% that a new firmware is going to be released. We just have to wait and see.
FYI, DS Scopes cannot save and transfer captured data (only screen shots).
How do you know they are still working on the software? Do you think we can expect an update soon?
The following is a reply I received today from DreamSourcelab about one of my reports:
"Regarding the issue mentioned at the start of the email about the “Amplitude tuning” feature not responding, this feature is not implemented in the current DS4T252 firmware. We will add this function in a future firmware update."
So, my understating is that they're working on a new firmware update.
The following is a reply I received today from DreamSourcelab about one of my reports:
"Regarding the issue mentioned at the start of the email about the “Amplitude tuning” feature not responding, this feature is not implemented in the current DS4T252 firmware. We will add this function in a future firmware update."
So, my understating is that they're working on a new firmware update.So there is a pretty high probability that I will update, so today I ordered my DS4T252 through their website

Hello,
I think also to order one DS4T252, seem very valuable device for in the field measurements.
Anyway I don't know if I must order it directly from Dreamsourcelab web site, or via Aliexpress store.
I'm located to France and I don't know what would be the amount of duty if ordered to them directly.
Any feedback of buyers ?
Regards.
FRex
Thanks for your feedback dcm.
I did as like you, I ordered it from their web site.
Now, time to wait..
Frex
My arrived today.
It's a great tool!
1mV maximum sensitivity!!! It's unique!
FFT and protocol decoding are very usefull!
Trigger is reliable, 2MB memory.
I'm very happy with it!
It would be very interesting to compare DS4T252 with the Zeeway DSO3D12.