Author Topic: EEVblog 121GW Multimeter Issues  (Read 677388 times)
Offline dr.diesel
Re: EEVblog 121GW Multimeter Issues
« Reply #750 on: November 27, 2018, 06:32:59 pm »
I suspect this has already been noted, and perhaps a hardware limitation, but manual range is lost after APO and back on again.
As of v1.57
Offline n3mmr
« Reply #751 on: November 30, 2018, 02:11:43 pm »
A minor issue with the manual: it doesn't talk you thru the idiot startup phase you have with  any new piece of gear.
Make sure meter is off.
Open it
Batteries in
Reattach the back
Turn the rotary switch to V
Display will show "U- FWversion", like "U- 1.54" for FW v 1.54.
A minor tech issue with the meter:
I would have liked an AC DC mode where main display shows V DC and secondary display shows V AC.
Kudos for the meter being useful, albeit not too accurate, up to at least 40kHz for sub-1V voltage.
Offline nonentity
  • symbol formerly known as software engineer
« Reply #752 on: December 03, 2018, 08:01:48 pm »
Nanofarad values of capacitance yield readings of zero with nanofarad manual ranging.  Occasionally an incorrect low value for a capacitance is displayed which then very slowly increases over a ~30 minute period up to the correct value.
Also the Logging/Bluetooth value for the nonofarad capacitance is 10 times what it should be.
Using 121GW meter recently purchased from Welectron with V1.57 firmware.
iPhone bluetooth app version: build 1.78
Problem is reproduced by:
1.) Attach 0.14 nf cap                            [edited 4 Dec 2018]
2.) Turn on meter, select capacitance
3.) Set mode to capacitance - the correct value usually displays.
4.) Cycle through mode selection and ranges once or twice.
5.) The capacitance shows zero in both nonfarad ranges or an incorrect slowly increasing value toward the correct value.
Slowcap1 shows the initial Log file.
Slowcap2 shows Log file after about 20 minutes.
« Last Edit: December 05, 2018, 04:54:56 am by nonentity »
Online lowimpedance
  • Watts in an ohm?
« Reply #753 on: December 03, 2018, 10:26:15 pm »
There was some discussion that had started in relation to this over in the firmware detail thread, that discussion morphed somewhat but to preserve that detail it stayed in that thread.
If there is a way for the 'moderators' to move the relevant posts here that would be better than diluting the other threads intent.
Offline nonentity
« Reply #754 on: December 04, 2018, 12:07:47 am »
I agree.  This is the issues thread.  The other thread is for something else.
Offline Seppy
  • Curious
« Reply #755 on: December 04, 2018, 03:04:28 am »
1.) Attach .014NF cap
3.) Set mode to capacitance the correct value usually displays.
The resolution of the capacitance is 0.01 nF, unless I've misunderstood you might be using it for values out of the meters range.
Regarding the app, an update hasn't yet been released to add support for the additional capacitance ranges (there is a note on the 121GW firmware page)
"NOTE: This version has a change that will not function correctly with the app in capacitance range. An update to the app will be released soon."
« Last Edit: December 04, 2018, 03:12:32 am by Seppy »
The following users thanked this post: nonentity
Offline Seppy
  • Curious
« Reply #756 on: December 04, 2018, 03:38:16 am »
Hello EEVBlog!
I feel honored to now be a part of this awesome community. What makes me very sad is that my first post has to be a negative one.
I purchased my 121GW meter on July 26th 2018, during a small batch at official EEVBlog store, after being informed about its availability by the newsletter. Soon I discovered how big mistake I have made.
After receiving the meter and unboxing it, few things immediately stood out. The meter was scratched in many places (especially the knob and in the area around lead sockets) - see attachment. Scratches on two flat sides of the knob clearly came from someone's else fingernails. Part of the knob you grab by your fingers was noticably shinier than other areas. Batteries have also already been put inside the meter - From what I've seen on unboxing videos, these should be included seperately in the pouch.  Conclusion is simple - I recieved a used meter. I thought - "Ok, fine. This is not that bad. I can use this meter nevertheless." And I could - for a month. After that strange things started to happen.
Before making the purchase I did some research online (mainly this forum) and discovered that at some point a common problem with 121GW meters was the knob itself. Apparently there was a recall done at some point regarding the contact issues (knob replacement and addition of a shim). This took place months earlier. Common sense suggested that my meter should have already been free of this issue. This couldn't be farther from the truth.
After a month of very, very light use the meter is basically unusable. The knob wipers stopped making a good contact with the PCB (i think) and the meter reboots constantly while switching modes. Resistance mode cannot be turned on at all. The meter just turns off when I select it. The issue is present for every single knob position.
I made a short video where the issue is clearly visible:
As you can see, this is clearly not a 121GW revision with a fixed knob. I must have received an older, flawed version. This is absolutely unforgivable for a meter this expensive (and every other for that matter). What I find the most frustrating is that the recall was done some time before I bought my meter from the EEVBlog store. At that point I should expect this issue to be already resolved but in the end I recieved a meter with this exact defect still present.
I've sent Dave an email describing these issues on 5th October. He replied a week later and promised to send me a new meter. I was really happy about that. So why am I even writing this post, you may ask? Well, because the new meter has never been shipped and Dave has not replied to any of my emails since then :palm:. I've been trying many times to contact him through at least 3 EEVBlog email adresses and sending him PMs on the forum, asking why the new meter hasn't been shipped. Three weeks have passed and I have not recieved any reply. I have no idea why.
I can only hope that this post will grab Dave's attention and that my meter will get replaced eventually. Do you know if there is any other way to contact Dave or to fix the meter? I just want to be able to use the product i paid for.
Thank you for your help in advance,
A VERY frustrated customer.
Hello, not sure if this has been dealt with. Are you able to check whether a shim is installed internally in your meter.
It should be there. The shim resolved that issue earlier on, but it is possible your unit is missing this part...
Offline Antonov225
« Reply #757 on: December 04, 2018, 05:47:57 pm »
Hi, This issue has been resolved. I have already received a replacement meter from Dave. Thank you.
Offline nonentity
« Reply #758 on: December 04, 2018, 10:46:40 pm »
My apologies, my message and post had the wrong capacitance value.  The capacitance used was 0.14 nano-farad.   Please see the corrected EEVblog 121GW Multimeter Issues Reply #869
« Last Edit: December 05, 2018, 04:59:10 am by nonentity »
Offline DavidDLC
« Reply #759 on: December 05, 2018, 06:38:15 pm »
I just installed firmware version 1.57, and it still shows a fixed current value of about 34 mA with nothing connected.
I had to go back to 1.26.
Does anybody know about this issue ?
David DLC
Offline chronos42
« Reply #760 on: December 05, 2018, 09:24:23 pm »
David DLC
look at #819 in this thread, maybe that helps.
The following users thanked this post: DavidDLC
Offline Seppy
  • Curious
« Reply #761 on: December 06, 2018, 12:28:13 am »
Right, the app just hasn't been updated thats a planned update (needed because multimeter added some capacitance ranges).
After a while I have been able to reproduce the issue, we will look into this.
Offline joeqsmith
« Reply #762 on: December 15, 2018, 05:12:18 pm »
Seems like YOU are the one polluting it.  Have anything technical to add?
I'll be very clear again.
This thread is for REPORTS of bugs and issues and CONFIRMATIONS of those bugs and issues.
DO NOT discuss them here, PLEASE!
I did not realize you were censoring posts in this area.   I resubmitted the post you removed twice because I literally thought I had forgotten to press the send key, not because I was trying to be a dick about it.
I've had a go at cleaning up this area as well to save you the time.  Looks like about three pages were my posts so it does add up.
« Last Edit: December 16, 2018, 05:44:07 pm by joeqsmith »
Offline chronos42
« Reply #763 on: December 16, 2018, 10:17:47 am »
I want point out the capatitance measure issue for low values again.
It was first discussed in the Firmware tread and also here.
This is a serious bug and makes the capacitance measurement unreliable for low values.
I still assume this is a firmware issue but I am unsure whether the developers are aware of this issue or not.
With a 100pF capacitor the meter has this behaviour:
First it shows the correct value.
After cycling through the modes it shows 21pF instead of 100pF and comes back very slowly to 100pF.
I had measured the waveform across the capacitor with a scope and found that it did not change in amplitude and time after cycling through the modes, so there seems no hardware related issue like wrong setups for the Hycon chip. Maybe a bug in the averange filter for the low capatitance mode? After cycling throug the modes it seems that the filter has a very long settling time for some reason. This issue should be investigated from UEI.
« Last Edit: December 16, 2018, 11:28:43 am by chronos42 »
Offline chronos42
« Reply #764 on: December 16, 2018, 11:11:32 am »
beside of the unacceptable slow autorange in ohms mode there is another issue in this mode.
After shorting the leads and pushing the rel button the meter zeros out the low ohm range and compensates the lead resistor as expected. But it also disables simultaneous the autorange function.
Why? That makes no sense, no other of my meters does this.
Re-enabling the auto mode deletes the resistor compensation  :palm:.
That should be changed.
Online IanB
« Reply #765 on: December 16, 2018, 04:32:09 pm »
As far as I know this is typical behaviour in DMMs. The Rel function is specific to the range you are in. If a meter autoranges to a different range the Rel function is cancelled.
It appears the 121GW is locking in the Rel'd range rather than autoranging away and undoing the Rel setting. This seems to me like a reasonable and valid design choice.
« Last Edit: December 16, 2018, 04:34:07 pm by IanB »
Offline GeoffreyF
« Reply #766 on: December 16, 2018, 05:00:35 pm »
When people characterize something as "unacceptable" - in this case auto range, why not state what you are comparing it too and how you have used that other "acceptable" device.  Price, other functions? All really acceptable?  Why didn't you buy that one?
US Amateur Extra W1GCF.
Offline CDaniel
« Reply #767 on: December 16, 2018, 05:23:37 pm »
GeoffreyF , 121GW is very slow in comparison with cheap or expensive multimeters , so stop being a fanboy troll . This firmware program should be scrapped completely and start from scratch with speed efficiency in mind . With a PIC family microcontroller a long time ago many , many users could have done a better firmware .
To add something , autorange hysteresis is 55000 counts up and 40000 counts down .Unreasonable huge . So , when you measure a 470ohm resistor from infinite resistance ( autoranging down ) , it will autorange in the 5K range and you will see 0,4700K .
But if you short the leads first and wait for 0 ohm reading ,  then measure fast ( now autorange is going up in ranges )  that 470ohm resistor , it will autorange correctly in 500ohm range as 470.00ohm .
« Last Edit: December 16, 2018, 05:38:08 pm by CDaniel »
The following users thanked this post: genghisnico13
Offline Marco1971
« Reply #768 on: December 16, 2018, 05:43:52 pm »
Hi, my Fluke 89IV/189 and 287 (with firmware 1.16) all do it in the same way with automatically switching to MANUAL ranging.
Regards from Italy.
Marco1971
Offline GeoffreyF
« Reply #769 on: December 16, 2018, 06:14:35 pm »
So you can't make any real point and you call me a fan boy troll?  You discredited yourself completely.   Did I say I was a "fan"?  No.  I own one. I own others.   Troll?  I am not one of these people who never gets tired of repeating the same point again and again, completely unproductively.   As for "Scrapped completely" (as opposed to correcting bugs).  It is equally obvious you never really worked on production code.
This is an Engineering blog - that means numbers, facts, comparisons - not rude school boy remarks.
Offline HKJ
    • Tests
« Reply #770 on: December 16, 2018, 06:28:07 pm »
A few meters do not disable auto range when doing min/max or REL, one of them is Hioki DT4282
The following users thanked this post: newbrain, Marco1971
Offline CDaniel
« Reply #771 on: December 16, 2018, 07:12:25 pm »
Yes , sure I'm discredited not you for implying that 121GW slowness has something to do with price ... Fanboys allways find excuses for any issues .
I don't think this firmware could be "corrected" that much . This is a core feature when designing a software to work efficiently . Not 3 lines of code that they couldn't change in 1 year or more ...
But I don't care how is done
« Last Edit: December 16, 2018, 07:14:39 pm by CDaniel »
Offline Simon
  • Global Moderator
    • Simon's Electronics
« Reply #772 on: December 16, 2018, 07:23:11 pm »
The 121GW is cheaper than the venerable Fluke 179, far more functional and nearly twice as accurate with one more digit. stop bleating.
Offline chronos42
« Reply #773 on: December 16, 2018, 08:16:28 pm »
Regards from Italy.
Marco1971
yes, some Fluke meters do it the same way and i also don't like that. But my Agilent U1252B, Brymen 869S and the Gossen Metrahit Energy does not. These meters ignore the low ohm offset correction in higher ranges, but they don't switch off the auto mode. This is the better way in my opinion.
« Last Edit: December 16, 2018, 08:28:30 pm by chronos42 »
Offline GeoffreyF
« Reply #774 on: December 16, 2018, 09:05:54 pm »
It's about Electronics - NOT YOU.   You keep repeating yourself.  Adding nothing.  Who found excuses?  This is an engineering forum.   Act like you are in one.