I recently got a trade for a 121GW meter. The previous owner did a great job calibrating DCV (thanks Scott!), but didn't have the tools to do ACV.
I tried using my SDG6052X to calibrate 5Vrms and below, but it simply didn't have enough resolution in the voltage control level to get within 7mV of the target.
Then I remembed the Magnova's AWG has more resolution. And it worked! My only complaint is that in the mVrms range, I could only adjust 2 decimal places, as the 3rd decimal was locked at 0.
Feature request: If possible, it would be awesome if that 3rd decimal place could be adjusted in the mV range.
Even with that limitation, I got some great results. Being able to adjust to 5 decimal places in the Vrms range made a huge difference!
For the test setup, the DAQ6510 is the calibrated reference, the AWG is connected directly to the 34465A, with 1 meter banana plug cables connecting the other meters.
Thanks,
Josh