The "insulating spacer" now makes sense as shown in the manual images. Inserting it causes the 'blades' in the slots of the Hcur and Hpot which are normally shorted together as are Lcur and Lpot 'blades' to separate and "Open". Without this effectively makes the external Kelvin connections ineffective as the Kelvin connections are completed inside the meter slots. Inserting the insulator spacer separates the H and L blades allows proper external Kelvin Sensing.
Also noted in the manual images, that a Calibration can be performed with a Short (0 Ω) and Open separately and/or alone without the need for the other resistance calibration values. This is like most other LCR Meters we have experience with, and the ability to use various precision resistors as Calibration Standards in a nice additional feature IMO.
From what we've seen in the manual images, one could construct a 'Shorting Bar' like we've done for our DE-5000 and do a Short Calibration, then an Open Calibration which should produce the most repeatable results for high and low Z leaded components.
Wonder if Zoyi will offer a Shorting Bar like Tonghui has for their Bench LCR Meters?
This looks very interesting and curious how the ZT-DQ01 compares with the DE-5000 or other known LCR meters wrt accuracy and repeatability. If we had one would be more than happy to compare with our TH2830 and IM3536 Lab Bench LCR Meters

Anyway, thanks Atlan for posting the manual images

Best