On mine, FW version 1.0.7, there is another bug there.
The tests are done using 1% resistance, with the BAT connector on Zoyi I get 99.45 mOhm, with YR1035+ 99.9 mOhm, and with the YR2050 99.88 mOhm.
Inserting the resistor in the rectangular connectors of the Zoyi, the results are in milliohms at levels 0.1V / 0.3V / 0.6V, the last digits sometimes are not stable and can fluctuate up to about +/- 1 mOhm:
100 Hz: 101 / 101 / 101
120 Hz: 101 / 101 / 101
1000 Hz: 101 / 46 / 46
10 KHz: 102 / 102 / 102
100 KHz: 110 / 124 / 105
There are also bugs in the manual ranges, sometimes I get measurements of 150 or even 170 mOhm in the lower 100 Ohm range, and good ones in the 100 KOhm range, sometimes they are solved by changing the range and selecting it again, and sometimes not.
It is true that for more precision in such low resistances the higher connector should be used, but these errors are unacceptable, I don't have time to test higher resistances, but I wouldn't be surprised if they also happen.