« on: June 14, 2023, 10:05:31 pm »
Before I ask my question....does anyone else here experience how hard it is to express to management underlying root cause electrical issues??   |O |O
I need a good process calibrator for some heat controllers that use J and K type thermocouples.  I like the fluke multipurpose (726) but its 5 grand US.   Can anyone make a suggestion on something thats as good without the price tag?  cant spend that much on fluke for the aforementioned management problems.... |O
Online TimFox
Re: Process Calibrator Reccomendations?
« Reply #1 on: June 14, 2023, 10:39:40 pm »
Do you own a sufficiently well-calibrated and sensitive voltmeter?
The thermal EMFs for J and K thermocouples are tabulated.
You need to buy a simple ice-point reference, and a cheap one can be made with a styrofoam cup, distilled water, and small ice cubes.
My old Keithley thermocouple meter had details for that junction construction.
This manual is for a similar instrument, and has ice junction construction in section 4.2.
https://assets.omega.com/manuals/M0618.pdf
The reason for the ice junction is to connect your two metals to two copper wires to avoid thermal errors at the voltmeter.
« Reply #2 on: June 14, 2023, 11:06:39 pm »
I use a fluke 289  for all my voltage metering issues.  im not trying to test the thermocouples accuracy, i'm trying to test the controllers calibration.   I dont think I follow how your advice helps me.
Offline Tomorokoshi
« Reply #3 on: June 15, 2023, 12:56:09 am »
1. By "test the controllers calibration" do you mean what the controller reads from the thermocouple? Its 4-20mA output? Something else? Those terms have both general definitions and very specific definitions depending on the industry they are being used in.
2. Google "process calibrator" and a few things show up. I took a quick look at the Uni-T UT725. Perhaps something like that is acceptable. Note: you get what you pay for with this kind of equipment.
However, I suspect the real problem, as you state, is how to talk to Management about this kind of thing. A few questions:
3. Is the organization financially stressed?
4. What kind of organization is it? Sales? Service? Manufacturing? Engineering? Other?
5. Is mechanical engineering the educational background of Management?
6. Is accounting the educational background of Management?
Offline J-R
« Reply #4 on: June 15, 2023, 12:57:37 am »
I picked up the Ronan X88 some months back and it's an interesting older device that may be an option for you.  You may have to search and wait a bit to find one for a reasonable price.
https://ronan.com/wp-content/uploads/2019/09/X88_brochure.pdf
https://www.ronansystems.com/assets/docs/X88_IOM.pdf
Online BillyO
« Reply #5 on: June 15, 2023, 03:22:35 am »
This is the best response I've seen in some time.  A bit brutal, but direct and to the point.
If your company really needs the accuracy, precision and traceability of the 5-10 grand instrument there is, unfortunately, no $200 Chinese drop-in replacement.
How precise do you need to be? +/- 0.0001 C?  More accurate?  +/- 10C?
What are your requirements as to target and tolerance?  Also, what's at stake if you are wrong?  Lives or refunds? :-//
 :popcorn:
Bill
Online BillyO
« Reply #6 on: June 15, 2023, 03:24:01 am »
Par for the course.
Bill
Offline J-R
« Reply #7 on: June 15, 2023, 05:06:25 am »
Fluke 726 is $4500 from Amazon, 725 is a bit less at $3600.
726 is about $2k used on eBay.
Online TimFox
« Reply #8 on: June 15, 2023, 01:46:13 pm »
Secton 4.2 of my citation shows how to calibrate a meter that measures a thermocouple voltage, using a voltage source and ice-point junction.
It is NOT a method to calibrate the thermocouple itself.
I used that procedure to calibrate a thermocouple meter without needing a variable temperature, using a measured voltage into the circuit (copper wires) and inserting the thermocouple wires into the meter.
I started with a commercial thermocouple (with appropriate connector), cutting the junction and using the two dissimilar wires in the circuit shown in section 4.2.
It's a nuisance to develop a known temperature of 1000o C on the bench, but a type J thermocouple in the circuit I cited needs 57.953 mV from the external voltage supply (measured on a normal voltmeter) to give the same results in the thermocouple meter.
I assume the expensive calibrators do essentially the same thing, although they do not use frozen distilled water for the ice-point junction.
Tables of thermocouple emfs for different metal combinations are readily available, e.g.  https://www.thermocoupleinfo.com/thermocouple-reference-tables.htm  and "Rubber Bibles" (Handbook of Chemistry and Physics, formerly published by the Chemical Rubber Company, now CRC)   https://hbcp.chemnetbase.com/faces/contents/ContentsSearch.xhtml;jsessionid=8E86BDA9072F1211BDD5A2DAF5F3F562
« Last Edit: June 15, 2023, 02:35:53 pm by TimFox »
The following users thanked this post: Tech_Monkey
Offline porter
« Reply #10 on: June 15, 2023, 03:32:18 pm »
1. I just need to simulate the thermoelectic voltage of a thermocouple.  The charts are online with the correlation of voltage to temperature. (roughly -10 to 30 millivolts) So I'm looking for a certified device to simulate those voltages and ensure that the temperature shown on the display matches what the thermocouple input is.  We have some situations where our machines are having to run hotter than expected on the controls. The problem is either ground looping, or lack of noise immunity in the controllers.   I believe the issue is with the grid, management believes its the controllers so frankly I'm about to spend 5k to prove them wrong.  (there is also harmonics issues and low power factor issues involved here. this whole grid is just a mess frankly)
2 I would like the equipment we buy to be good for the long term, calibrated, certified, and able to grow with a rapidly growing company.
3 no and thats what annoys me more than anything.
4 heavily engineering reliant manufacturing.
5 lol
6.  lol x2.
Online TimFox
« Reply #11 on: June 15, 2023, 03:38:41 pm »
The method I suggested is calibrated from first principles.
It may be cumbersome and labor-intensive, but apparently you can't get enough budget to purchase the faster, higher-cost, digitally controlled calibration equipment.
Perhaps your boss can hire, as a summer intern, a nephew who studied high-school physics and can read instructions to do it at an acceptable cost?
« Reply #12 on: June 15, 2023, 03:49:57 pm »
idk why i didnt think about fleabay. Thanks.
Offline RAPo
« Reply #13 on: June 15, 2023, 03:52:55 pm »
I use The service of .
About 250 euro on Ali.
« Reply #14 on: June 15, 2023, 03:54:15 pm »
I understand what youre saying. But thats not really going to work for the type of tests  i need to run across a range of loads and power quality situations.   Process calibrator makes it much simpler.  I found one for 1800 i think I can get. so no worries. thanks for your input
Online TimFox
« Reply #15 on: June 15, 2023, 04:22:36 pm »
You can invert the measurement I suggested to check the calibration on your reasonably-priced calibrator against a physical ice-point and a calibrated voltmeter.
Offline J-R
« Reply #16 on: June 15, 2023, 05:03:09 pm »
Extech PRC30 seems to be an interesting budget option, $850 with NIST certificate.  They are in stock.
Online TimFox
« Reply #17 on: June 15, 2023, 05:53:08 pm »
Note that for my J example at 1000o C, the thermal emf has a slope of 0.06 mV/K, or 0.1% of value per degree.
This is useful to determine what voltmeter accuracy is required either for calibrating the thermocouple meter, or the inverse circuit for checking the calibration of the thermocouple-simulating calibrator.
Offline rvalente
  • Country: br
« Reply #18 on: June 15, 2023, 09:42:44 pm »
Offline jchw4
« Reply #19 on: June 16, 2023, 09:12:49 am »
There are also Omega calibrators: https://www.omega.com/en-us/calibration/temperature-calibrators/c/handheld-calibrators
I own the simplest CL3515R and it's pretty good. They still sell it and offer support and recalibration.
But there are certainly better / newer models.
Upd: they have multifunction calibrators as well https://www.omega.com/en-us/calibration/c/multifunction-calibration but I have no experience with them.
« Last Edit: June 16, 2023, 09:15:10 am by jchw4 »
Online tszaboo
« Reply #20 on: June 16, 2023, 09:23:33 am »
When I had issues about equipment calibration, I turned to the quality management of my company. We raised it as a Nonconformity issue for ISO9001, then everything got calibrated. But yeah, those Fluke units are expensive for what they do.
Offline armandine2
« Reply #21 on: June 16, 2023, 10:45:55 am »
RS Pro 133 - sources and measures thermocouples
Gold Capacitor Prize - 2025
Offline Ground_Loop
« Reply #22 on: June 16, 2023, 03:15:07 pm »
How about a Fluke 787 process meter with a precision 10 ohm, or less, resistor across the outputs. You can simulate the TC voltage to a reasonable accuracy and the 787 is also a great all around multimeter otherwise. It’s still a bit pricey though at around $1000.   BTW I’ve done just this with my 787.
There's no point getting old if you don't have stories.
Offline Aussie Instro Tech
« Reply #23 on: January 24, 2026, 01:10:12 am »
Firstly, Fluke 787 is a good meter but has been superseded by the 789 which allows you to loop power a 4 - 20 mA loop.
In your post you mention the following,
I would like to know if you hand write calibration reports. If so, what happens to them ? Does anybody review them ? Where and how are they stored ? In this day and age if a business is still relying on hand written reports for calibration of their equipment, the management is living in the dark ages.
You mention your test gear needs to be certified, calibrated and good for the long term and suitable for a rapidly growing company.
Are management interested in optimising  the workplace and making profits ? If so there is only one choice, a Beamex MC6, the best multi function calibrator IN THE WORLD, bar none. It is a documenting calibrator which stores the result, tells you if your calibration result is a pass or fail and you store the results on a PC using the Beamex CMX software A report can be generated and printed out, listing all the instruments parameters, the calibration result, as found and as left calibration data, day date, year etc, the list goes on. It also gives a graphical representation of the result.
For T/C simulations the Beamex has reference or cold junction compensation, no need for an ice water slurry. Noah used that on his Ark !
Expensive yes, but worth every cent. If you are serious about getting the best results there is no other choice, cheers !