What is the result when Holdoff is increased to say 0.5s ?
10 Hz is the BW limit where an Edge trigger will reliably catch a pulse regardless if either Normal or Peak Detect acquisition has been selected.
For LF pulses a Pulse trigger from the Trigger menu is required.
10 Hz is the BW limit where an Edge trigger will reliably catch a pulse regardless if either Normal or Peak Detect acquisition has been selected.
For LF pulses a Pulse trigger from the Trigger menu is required.
I have tried pulse trigger, but the result is exactly the same. It shouldn't be BW related, because the pulse risetime was ~100 ns, only the repetition rate was slow.
But did you manage to replicate the behavior?
But did you manage to replicate the behavior?Yes exactly the same with SDS1104X-E......same UI and features. Signal source SDG1032X.
...
Normal triggering was the most reliable and it keeps the pulse on the display ...
Interesting news...
I have tried several options and discovered that the problem disappears, even at fast horizontal settings, if I set "Acq Mode" to "slow" (Acquire button, second page). This function is not even mentioned in the user manual... the context help in the instrument says you can disable display hardware acceleration with "slow", in contrast with the standard "fast".
Obviously I'm not happy to leave the instrument in slow mode, but at least I have a solution.
Yesterday, while I was trying to use the scope to tune an Efratom LPRO-100 rubidium oscillator using a GPS, I have found a weird behavior. I had the scope in normal sync mode, rising front on the channel getting the slow PPS signal (1 Hz, 5 us duration); the other channel was connected to the 10 MHz output of the oscillator. I saw false syncs, signals with discontinuous phases and other weird things that let me think the scope was broken. Then I realized that it happened only above a certain horizontal speed setting: at low speeds like 500 us/div or slower (using then zoom to see the detail I needed) all was working correctly, letting me suppose it may be a firmware or FPGA problem. No problem at all in "single" mode, only in "normal".
I have tried applying trigger holdoff, even close to 1 s, without any improvement.
I have the latest downloadable firmware 1.1.19R5 with FPGA 2019-08-12 (however, I can't find any FPGA update), hardware version 00-01.
I could replicate the problem today using my Siglent SDG2042X, and I discovered that it happens even in single channel mode. I have posted a video here:
https://youtu.be/-re2NW7mH-k
where you can see the result of applying short pulses (1,2 or 5 us duration, 100 ns rise time) with 1 s PRF and normal sync mode, rising front, at half level. In this case all is ok for 200 us/div or less, while for 100 us/div or more it starts acquiring out of sync or showing a flat line. If I change the pulse duration, it even mixes the old and new duration, so the impression is that there is a bad pointer in the acquisition memory. If I increase the PRF above 10 Hz, the problem seems to disappear.
I'd like to know if someone else can see the same behavior, in which case it is almost surely an FPGA bug, or if my unit is faulty and should be returned (still under warranty period).
I have tried to turn it off and on, to start from default, to invoke autocal, but nothing...
The unit isn't hacked. I have purchased the MSO option, but the digital probe wasn't connected to the scope.
Maybe I'm wrong in some way... I'd appreciate your hints.
Thank you very much,
Roberto
RoV IW3IPD
@rf-loop: that still doesn't explain why pulses with the wrong width are appearing randomly.
Edit: I watched the video from the OP again and it definitely is a bug. After he changes the pulse width to 2us the 1us pulses are still popping up randomly. This is not acceptable behaviour for an oscilloscope.
In this point we know it do not work as can expect.
(later in video exist other mystery but lets first jump over it)
@rf-loop: that still doesn't explain why pulses with the wrong width are appearing randomly.
Edit: I watched the video from the OP again and it definitely is a bug. After he changes the pulse width to 2us the 1us pulses are still popping up randomly. This is not acceptable behaviour for an oscilloscope.
Better practice is divide messy problem to different parts. I was handling this first part, as I told. First need be sure how settings are and test with perfectly known setup and other things and find confirmation it trig in Normal trig mode without visible pulse, as can see it happen.
Watch the video again and you'll see. In the first part (where the OP changes the pulse to 2us) there is no reason for the oscilloscope to start mixing 1us and 2us pulses. The oscilloscope is probably in auto trigger mode so it shows a flat line and a pulse in an alternating fashion but again: where does the 1us pulse comes from if there are only 2us pulses at the input?

@rf-loop: that still doesn't explain why pulses with the wrong width are appearing randomly.
Edit: I watched the video from the OP again and it definitely is a bug. After he changes the pulse width to 2us the 1us pulses are still popping up randomly. This is not acceptable behaviour for an oscilloscope.
Better practice is divide messy problem to different parts. I was handling this first part, as I told. First need be sure how settings are and test with perfectly known setup and other things and find confirmation it trig in Normal trig mode without visible pulse, as can see it happen.Watch the video again and you'll see. In the first part (where the OP changes the pulse to 2us) there is no reason for the oscilloscope to start mixing 1us and 2us pulses. The oscilloscope is probably in auto trigger mode so it shows a flat line and a pulse in an alternating fashion but again: where does the 1us pulse comes from if there are only 2us pulses at the input?
I assume you have a Siglent scope and a pulse generator so try to re-create this setup yourself with the test I described earlier (using incrementally wider pulses so each measurement can be identified).
Watch the video again and you'll see. In the first part (where the OP changes the pulse to 2us) there is no reason for the oscilloscope to start mixing 1us and 2us pulses. The oscilloscope is probably in auto trigger mode so it shows a flat line and a pulse in an alternating fashion but again: where does the 1us pulse comes from if there are only 2us pulses at the input?
No no, it was in normal trigger mode! Trigger arrived at 1 s pace, as the input pulses: when a flat line is displayed, it's because the scope is showing a wrong memory portion, I think. In fact, with an x-axis setting a bit slower, the pulse tends to appear in the wrong place, instead of giving a flat line.
As I told, all of this disappears above 200 us/div, or if I put the scope in slow mode.
I have also discovered that if the scope is in dual trace mode, with chA with the pulse triggered as usual and chB e.g. with a 10 MHz sinusoid, if I turn off chB, chA shows for a few pulses a mess of mix of previous chA and chB contents, similarly to what happens for the same channel when I change the pulse length.
I suppose it's a bug that will be fixed with a future update and not a defect of my scope
Watch the video again and you'll see. In the first part (where the OP changes the pulse to 2us) there is no reason for the oscilloscope to start mixing 1us and 2us pulses. The oscilloscope is probably in auto trigger mode so it shows a flat line and a pulse in an alternating fashion but again: where does the 1us pulse comes from if there are only 2us pulses at the input?
No no, it was in normal trigger mode! Trigger arrived at 1 s pace, as the input pulses: when a flat line is displayed, it's because the scope is showing a wrong memory portion, I think. In fact, with an x-axis setting a bit slower, the pulse tends to appear in the wrong place, instead of giving a flat line.
As I told, all of this disappears above 200 us/div, or if I put the scope in slow mode.
I have also discovered that if the scope is in dual trace mode, with chA with the pulse triggered as usual and chB e.g. with a 10 MHz sinusoid, if I turn off chB, chA shows for a few pulses a mess of mix of previous chA and chB contents, similarly to what happens for the same channel when I change the pulse length.
I suppose it's a bug that will be fixed with a future update and not a defect of my scope
So, for make sure all is as default, do full factory default.
After then set trigger rising edge. Level 2.5V and trigger mode "Normal"
scope for Set 2V/Div. Set time scale for 1us/div.
Take 1s period 1us wide pulses, level 0 - 5V using coaxial cable from your SDG to scope input.
, while I really acquired ten!!! The intermediate frames show a time which in my opinion is nonsense, because it is not an intermediate time.
12. Fixed the bug: Normal trigger can show more-than-one trigger event on the display at one time.

If you like you can run in slow mode long capture... so that time scale is enough short for get lot of these second interval pulses. (do it exist with more fast intervals it is not clear for me). After it run some hundreds of seconds it can look if there in history is any captures without pulse. Because slow mode can trig/capture only less than 40ms interval. But fast mode with your setup it can capture, very rough guessing, least faster than 500us interval.