« Reply #1700 on: October 21, 2024, 06:49:50 pm »
I have bought a Siglent SDS804x HD oscilloscope some days ago. I feel a little lost in the using of digital oscilloscopes and I have a doubt whose answer is easy for sure.
In the attached image we can see that the signal frequency is 999.77 Hz (1 KHz theoretically) but on the upper side of the screen, with yellow characters, is shown a frequency of 2.08 KHz.
Sometimes the trigger fails to detect the signal and the signal shifts horizontally.
Why this difference in the measure and how is the yellow value obtained?

« Last Edit: October 21, 2024, 06:53:17 pm by IC_Toaster »
 

Offline electr_peter

  • Supporter
  • ****
  • Posts: 1616
  • Country: lt
Re: Siglent SDS800X HD 12 bit DSO's
« Reply #1701 on: October 21, 2024, 06:58:07 pm »
Yellow freq measurement is from HW counter. Signal looks noisy, counter likely picks up more some of that and gives higher reading. Depends on situation if HW or SW freq is more useful, but in general HW counter is nice to have.

From 800X HD user manual:
Quote
18.12 Hardware Frequency Counter
The hardware frequency counter is a stand alone, hardware based frequency measurement function. Differing from the “Frequency” parameter (see section “ Horizontal Measurement ”) which is measured by software, the hardware frequency counter directly use s the trigger signal as the input of a hardware counter, which yields 7 digit frequency measurement accuracy. The accuracy of the “Frequency” parameter measured by software is relative to the sample rate. The lower the sample rate, the worse the accuracy. In situations where the sample rate is less than 2 times the input frequency, the results will be incorrect . The hardware frequency counter does not have this limitation. Its result is always
 
The following users thanked this post: IC_Toaster

Offline IC_Toaster

  • Regular Contributor
  • *
  • Posts: 97
  • Country: es
Re: Siglent SDS800X HD 12 bit DSO's
« Reply #1702 on: October 21, 2024, 08:39:39 pm »
Yellow freq measurement is from HW counter. Signal looks noisy, counter likely picks up more some of that and gives higher reading. Depends on situation if HW or SW freq is more useful, but in general HW counter is nice to have.

From 800X HD user manual:
Quote
18.12 Hardware Frequency Counter
The hardware frequency counter is a stand alone, hardware based frequency measurement function. Differing from the “Frequency” parameter (see section “ Horizontal Measurement ”) which is measured by software, the hardware frequency counter directly use s the trigger signal as the input of a hardware counter, which yields 7 digit frequency measurement accuracy. The accuracy of the “Frequency” parameter measured by software is relative to the sample rate. The lower the sample rate, the worse the accuracy. In situations where the sample rate is less than 2 times the input frequency, the results will be incorrect . The hardware frequency counter does not have this limitation. Its result is always

The signal is noisy and probably this is the reason because the frequency measurement has desviations. Just I try to understand in which way the noise affect both measurement of the same parameter.
As I understand from the user manual, the counter depends on the trigger signal as you quoted, and the frequency parameter depends on the edges detected:
Quote
Period: Time between the middle threshold points of two consecutive like-polarity edges
Freq: Reciprocal of the period


If the counter obtains a value which is the double from the true frequency, does it means that the trigger is triggered twice on a period?
« Last Edit: October 21, 2024, 08:45:32 pm by IC_Toaster »
 

Offline 2N3055

  • Super Contributor
  • ***
  • Posts: 8738
  • Country: hr
Re: Siglent SDS800X HD 12 bit DSO's
« Reply #1703 on: October 21, 2024, 11:16:42 pm »
Page 88 of User manual explains how to set trigger Noise reject and HF reject.
Play with these two options to get more stable trigger.
If you get it hardware counter will show more accurate reading.
"Just hard work is not enough - it must be applied sensibly."
Dr. Richard W. Hamming
 
The following users thanked this post: IC_Toaster

Offline markus_jlrb

  • Regular Contributor
  • *
  • Posts: 179
  • Country: de
Re: Siglent SDS800X HD 12 bit DSO's
« Reply #1704 on: October 22, 2024, 06:22:54 am »
BTW,

why we had the sig dissymmetry in the screen image?

First thought would be, because we had a DC offset and the DC coupling enabled,
but perhaps after doing AutoSetup the sig become symmetrical.

Could you verify this on your new scope please.

I made same observation on mine device, see attached screen shots.

Markus

 

 

Offline 2N3055

  • Super Contributor
  • ***
  • Posts: 8738
  • Country: hr
Re: Siglent SDS800X HD 12 bit DSO's
« Reply #1705 on: October 22, 2024, 10:00:02 am »
People, this scopes have axis labels.
Go into Display Menu-> Axis label Setting.
I use Moving labels, so numbers follow position on screen
It cuts down on confusion.

@ markus_jlrb, IC_Toaster has 1 V of offset set on channel 1. And on that image trigger is set for 1.66V.
"Just hard work is not enough - it must be applied sensibly."
Dr. Richard W. Hamming
 
The following users thanked this post: IC_Toaster

Offline IC_Toaster

  • Regular Contributor
  • *
  • Posts: 97
  • Country: es
Re: Siglent SDS800X HD 12 bit DSO's
« Reply #1706 on: October 22, 2024, 03:46:17 pm »
Page 88 of User manual explains how to set trigger Noise reject and HF reject.
Play with these two options to get more stable trigger.
If you get it hardware counter will show more accurate reading.

Both options gave me a stable adquisition and a reliable frequency measurement.
Thank you for your advise.

The steep from analog scilloscopes to digital one is big.
 
The following users thanked this post: 2N3055

Offline 2N3055

  • Super Contributor
  • ***
  • Posts: 8738
  • Country: hr
Re: Siglent SDS800X HD 12 bit DSO's
« Reply #1707 on: October 22, 2024, 03:58:27 pm »
Page 88 of User manual explains how to set trigger Noise reject and HF reject.
Play with these two options to get more stable trigger.
If you get it hardware counter will show more accurate reading.

Both options gave me a stable adquisition and a reliable frequency measurement.
Thank you for your advise.

The steep from analog scilloscopes to digital one is big.

Glad to be able to help.
Yes, you are right, it needs some learning and getting used to, but it is so much more capable...
Take care!
"Just hard work is not enough - it must be applied sensibly."
Dr. Richard W. Hamming
 

Offline IC_Toaster

  • Regular Contributor
  • *
  • Posts: 97
  • Country: es
Re: Siglent SDS800X HD 12 bit DSO's
« Reply #1708 on: October 22, 2024, 04:01:28 pm »
BTW,

why we had the sig dissymmetry in the screen image?

First thought would be, because we had a DC offset and the DC coupling enabled,
but perhaps after doing AutoSetup the sig become symmetrical.

Could you verify this on your new scope please.

I made same observation on mine device, see attached screen shots.

Markus

Have you measured the mean value? For sine signals it is the DC value.
What source are you using?
My only signal generator is the Zoyi ZT-703s and this generates a 2.5Vpp sine with 1.25 Vdc..
 

Offline IC_Toaster

  • Regular Contributor
  • *
  • Posts: 97
  • Country: es
Re: Siglent SDS800X HD 12 bit DSO's
« Reply #1709 on: October 22, 2024, 04:03:46 pm »
People, this scopes have axis labels.
Go into Display Menu-> Axis label Setting.
I use Moving labels, so numbers follow position on screen
It cuts down on confusion.

@ markus_jlrb, IC_Toaster has 1 V of offset set on channel 1. And on that image trigger is set for 1.66V.

You are right, it is possible to activate the axis labels.
« Last Edit: October 22, 2024, 04:11:23 pm by IC_Toaster »
 

Offline IC_Toaster

  • Regular Contributor
  • *
  • Posts: 97
  • Country: es
Re: Siglent SDS800X HD 12 bit DSO's
« Reply #1710 on: October 22, 2024, 04:05:33 pm »
Page 88 of User manual explains how to set trigger Noise reject and HF reject.
Play with these two options to get more stable trigger.
If you get it hardware counter will show more accurate reading.

Both options gave me a stable adquisition and a reliable frequency measurement.
Thank you for your advise.

The steep from analog scilloscopes to digital one is big.

Glad to be able to help.
Yes, you are right, it needs some learning and getting used to, but it is so much more capable...
Take care!

Yes, digital oscilloscopes are much more capable.
Take care you too.
 
The following users thanked this post: 2N3055

Offline eTobey

  • Super Contributor
  • ***
  • Posts: 1573
  • Country: de
  • Virtual Features for the SDS800XHD -> My website
    • Virtual feature script
Re: Siglent SDS800X HD 12 bit DSO's
« Reply #1711 on: October 24, 2024, 09:03:37 am »
I found this in the manual:
Quote
... the hardware frequency counter directly uses the trigger signal as the input of a hardware
counter ...
I set up a delay trigger, which i checked on the trigger output, which is as expected, at about 4.5Hz. But when looking at the frequency value of the hardware counter it shows 100Hz.
Interestingly, i have a 100Hz signal between the trigger. So either the manual is wrong, the scope has another issue or its an user error.
2412203-0
"Sometimes, after talking with a person, you want to pet a dog, wave at a monkey, and take off your hat to an elephant."(Maxim Gorki)

Siglent SDS800X HD bugs/issues/workarounds (Partially updated 3. Mar 2026)
 

Offline 2N3055

  • Super Contributor
  • ***
  • Posts: 8738
  • Country: hr
Re: Siglent SDS800X HD 12 bit DSO's
« Reply #1712 on: October 24, 2024, 10:46:12 am »
I found this in the manual:
Quote
... the hardware frequency counter directly uses the trigger signal as the input of a hardware
counter ...
I set up a delay trigger, which i checked on the trigger output, which is as expected, at about 4.5Hz. But when looking at the frequency value of the hardware counter it shows 100Hz.
Interestingly, i have a 100Hz signal between the trigger. So either the manual is wrong, the scope has another issue or its an user error.
(Attachment Link)

Error in presumptions.
Fact that counter uses data from parts of trigger path is not same as trigger using counter circuits.

On beginning of trigger path there is a comparator block that will detect every transition through threshold. Those transitions are fed to hardware counter block and rest of trigger block. Rest of trigger functions then start acquisition if other prerequisites are fulfilled.. That happens asynchronously with comparator block that keeps detecting edges and feeding it to counter block, and rest of trigger functions ignore it until they are ready again.
Trigger out outputs pulses that are synchronous with actual acquisitions.

Note in documentation that counter is connected through parts of trigger circuit is there to point to the fact that counter won't work properly unless you have stable trigger.
"Just hard work is not enough - it must be applied sensibly."
Dr. Richard W. Hamming
 
The following users thanked this post: newbrain, Antonio90, tmeub

Offline temperance

  • Super Contributor
  • ***
  • Posts: 1589
  • Country: 00
Re: Siglent SDS800X HD 12 bit DSO's
« Reply #1713 on: October 24, 2024, 11:54:22 am »
I found this in the manual:
Quote
... the hardware frequency counter directly uses the trigger signal as the input of a hardware
counter ...
I set up a delay trigger, which i checked on the trigger output, which is as expected, at about 4.5Hz. But when looking at the frequency value of the hardware counter it shows 100Hz.
Interestingly, i have a 100Hz signal between the trigger. So either the manual is wrong, the scope has another issue or its an user error.
(Attachment Link)

One can clearly see a 100 Hz signal on CH2. Unless the scope is making up some signal it must be you feeding that signal into CH2. You are probably seeing a glitch generated a counter from a micro controller or something. Some controllers like an AVR will output a glitch even with the compare value set to 0.
 

Offline SHF

  • Regular Contributor
  • *
  • Posts: 83
  • Country: de
Re: Siglent SDS800X HD 12 bit DSO's
« Reply #1714 on: October 25, 2024, 06:20:50 pm »
One question, I got my new SDS804-HD today, connected a GPSDO with 10 MHz as a test signal and the counter at the top right shows 10.00015 MHz! Is that within the normal range?
Firmware V1.1.3.6
Precision is no coincidence
 

Online Martin72

  • Super Contributor
  • ***
  • Posts: 8877
  • Country: de
Re: Siglent SDS800X HD 12 bit DSO's
« Reply #1715 on: October 25, 2024, 06:34:30 pm »
Hi,

The accuracy of the time base is +/-25 ppm, so no worries.
How accurate should your GPS be, to put it another way... ;)




Offline SHF

  • Regular Contributor
  • *
  • Posts: 83
  • Country: de
Re: Siglent SDS800X HD 12 bit DSO's
« Reply #1716 on: October 27, 2024, 06:34:33 pm »
Does anyone know what the time base frequency is?
The SDS1104X-E runs at 25MHz
The SDS800X-HD runs at .....?
Precision is no coincidence
 

Online ebastler

  • Super Contributor
  • ***
  • Posts: 7735
  • Country: de
Re: Siglent SDS800X HD 12 bit DSO's
« Reply #1717 on: October 27, 2024, 07:07:06 pm »
Does anyone know what the time base frequency is?
The SDS1104X-E runs at 25MHz
The SDS800X-HD runs at .....?

Here's a post from rf-loop, a Siglent "insider", stating that it's a 25 MHz reference clock as well:
https://www.eevblog.com/forum/testgear/siglent-sds800x-hd-12-bit-dsos-coming/msg5453810/#msg5453810

I seem to recall teardown photos showing the 25 MHz quartz (larger and presumably more stable than a couple of others on the PCB), and also a discussion whether it might be possible to "force" the clock to sync with an external reference. But can't find that thread, and may be confusing it with some other scope model.
 
The following users thanked this post: SHF

Offline SHF

  • Regular Contributor
  • *
  • Posts: 83
  • Country: de
Re: Siglent SDS800X HD 12 bit DSO's
« Reply #1718 on: October 27, 2024, 09:05:52 pm »
Thank you very much, I have to do some research
Precision is no coincidence
 

Online Martin72

  • Super Contributor
  • ***
  • Posts: 8877
  • Country: de
Re: Siglent SDS800X HD 12 bit DSO's
« Reply #1719 on: October 27, 2024, 09:09:03 pm »
What are you up to, if I may ask?

Offline SHF

  • Regular Contributor
  • *
  • Posts: 83
  • Country: de
Re: Siglent SDS800X HD 12 bit DSO's
« Reply #1720 on: October 27, 2024, 09:19:12 pm »
to synchronize the time base with an external GPSDO
Precision is no coincidence
 
The following users thanked this post: Martin72, EvgenyG

Offline maxmatteo

  • Contributor
  • Posts: 37
  • Country: de
Re: Siglent SDS800X HD 12 bit DSO's
« Reply #1721 on: October 28, 2024, 09:19:51 pm »
finally pulled the trigger for an sds804x.

unfortunately the fan noise is quite loud :/

has anyone already tested similiar fan noise mods like with the SDS1204X ?

replacing with noctua fan? lowering voltage with diodes or 7809 regulator?

thanks for any insights
Audio / Electronics / Web
 

Online tautechTopic starter

  • Super Contributor
  • ***
  • Posts: 31930
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Siglent SDS800X HD 12 bit DSO's
« Reply #1722 on: October 28, 2024, 09:39:56 pm »
finally pulled the trigger for an sds804x.

unfortunately the fan noise is quite loud :/

has anyone already tested similiar fan noise mods like with the SDS1204X ?

replacing with noctua fan? lowering voltage with diodes or 7809 regulator?

thanks for any insights
Fan noise is similar to SDS11/1204X-E as expected with a small fan spinning fast.
The OEM fans are roughly the same noise as Noctua therefore reducing noise is only possible by slowing them down < not advised.
Several threads contain info on fan mods to the 4ch X-E models if you take the time to hunt them out.

SDS1000X HD in the much larger enclosure uses a ~80mm fan so is somewhat quieter.
Avid Rabid Hobbyist
 
The following users thanked this post: Martin72

Online Martin72

  • Super Contributor
  • ***
  • Posts: 8877
  • Country: de
Re: Siglent SDS800X HD 12 bit DSO's
« Reply #1723 on: October 28, 2024, 10:14:50 pm »
What you could try is to find a fan that is quieter at the same air flow rate.
To do this, you would need to know the data of the “Siglent” fan and check other fans accordingly.
Whether this effort is worthwhile remains to be seen, because as Tautech has already noted, the fans will not differ much from each other – size pays tribute to physics.
Another thing you can do, but it will certainly void the warranty, is to remove the fins in the inner metal casing to reduce drag for the flow and thus minimize air noise.
Overall, it's an effort I wouldn't go to.
« Last Edit: October 28, 2024, 10:34:13 pm by Martin72 »
 

Offline 2N3055

  • Super Contributor
  • ***
  • Posts: 8738
  • Country: hr
Re: Siglent SDS800X HD 12 bit DSO's
« Reply #1724 on: October 28, 2024, 10:25:59 pm »
Removing fins on fan opening might increase EMC emissions... I said might.. I didn't try.
"Just hard work is not enough - it must be applied sensibly."
Dr. Richard W. Hamming
 
The following users thanked this post: Martin72


Share me

Digg  Facebook  SlashDot  Delicious  Technorati  Twitter  Google  Yahoo
Smf