Bringing a question by PM from member
@hip here for all to benefit from.
hip asked:
Let me ask you a question about the Siglent SDS804X HD, please. When turned on, I observe a constant offset across all channels, but especially on channel one, reaching 1 mV. It takes about 25-30 min for stabilize this offset to 0. The other three channels stabilize a little faster, about 15 minutes. No calibration manipulation or resetting to factory default has any effect. If, after the offset has stabilized, I turn the device off for 5 min and then turn it back on, the offset reappears. Is this a malfunction? If not, why is channel one so different from the others and takes longer to stabilize? I also have a Siglent SDS2104X HD, and its offset when cold is much lower across all channels.
Please see the links for a screenshot of the SDS804 and a photo of the SDS2104. Thank you.
Instrument warm up to full operational temp is a real thing if best accuracy is required.
Most of us are aware that a Self Cal must be done only after some 20-30mins of operation to provide fully reproducible results after the instrument fully warms up.
However we can also activate the Quick Cal feature in the Utility/Maintenance menu to periodically compensate drifts in accuracy while the scope reaches full operating temperature.
Quick Cal momentarily stops measurements while the active channel zero and accuracy is adjusted and those that have used LeCroy DSO's will be familiar with its process.
It could be that hip has performed the Self Cal process with local interference nearby which could impact on the Self Cal result where in his screenshot below there is a 1mV 0V offset on Ch1.
It should be noted, hip mentions his SDS804X HD whereas his screenshot filename tells a different story.


My SDS814X HD with the same setup AFAICT and with the Quick Cal showing however at time of screenshot it had not run.