Author Topic: Simple Oscilloscope Question  (Read 7879 times)
Offline frank2644Topic starter
« on: March 09, 2021, 02:22:17 am »
After 60 years of analog scopes, I just purchased a Siglent 1104x-E and I love it.
But I can't seem to do what should be a simple thing. I just want to stop the roughly 200hz signal on the display and scroll back in time looking for stuff, but I only see maybe 5ms in either direction using the play buttons (between history and Navigate button). When I use the History menu underneath the waveform display that menu shows only 3 frames available, also a very short period.
I've read the manual to no avail. Seems to me this should be easy to do (even for an old guy).
I've attached a JPG showing the limit using the fwd arrow button (bk arrow shows similar message)
Please tell me there is a simple way to do this.
Frank
Re: Simple Oscilloscope Question
« Reply #1 on: March 09, 2021, 03:15:08 am »
I've attached a JPG showing the limit using the fwd arrow button (bk arrow shows similar message) (Attachment Link)
Frank
Capture long and then zoom in.
If you need to see the small portion of waveform use Zoom mode (press timebase control) to enable the zoom window so you can work within the zoomed window.
By using this method and setting the primary timebase for the highest memory depth you'll have a large record to inspect in the zoomed window with any of the normal controls and be able to pan completely through the captured record.
Avid Rabid Hobbyist
Offline ataradov
    • Personal site
« Reply #2 on: March 09, 2021, 03:58:35 am »
There should be a manual control for the memory depth and/or sampling rate independent of the display setting. By default it would store enough to display on the screen. So yes, one possible workaround is to capture a lot and zoom in, but make sure that it also does not reduce the sample rate at long time base settings.
Alex
Offline SilverSolder
« Reply #3 on: March 09, 2021, 05:05:53 am »
 :-DD
Offline ataradov
    • Personal site
« Reply #4 on: March 09, 2021, 05:06:52 am »
What is so funny?
Alex
Offline Fungus
« Reply #5 on: March 09, 2021, 05:07:30 am »
This is a well known limitation of Siglents:
Not on a Siglent, sorry. It's an obvious thing to do but on a Siglent you have to mess around to capture more data.
« Last Edit: March 09, 2021, 05:09:51 am by Fungus »
The following users thanked this post: frank2644
Offline Someone
    • send complaints here
« Reply #6 on: March 09, 2021, 07:03:14 am »
Ignore the trolls above who are trying to cause alarm. The history mode only has a certain amount of memory available to it (practical limit of the instrument) and you've set a long memory depth for each capture so they fill up the history buffer very quickly.
To capture more history frames, go into the Memory Depth menu and reduce the number of points being captured in each frame.
Quote from the manual:
Press the [Acquire] button on the front panel; press the Mem Depth softkey and then turn the Universal Knob to select the desired value and push down the knob to confirm.
There is detail on the specific memory depths and number of history buffers compiled here:
https://www.eevblog.com/forum/testgear/siglent-sds1204x-e-released-for-domestic-markets-in-china/msg2098714/#msg2098714
« Last Edit: March 09, 2021, 07:16:14 am by Someone »
The following users thanked this post: tautech, frank2644
« Reply #7 on: March 09, 2021, 01:21:29 pm »
What is so funny?
Not anything specific that you said @ataradov, it was just the thought of the whole "Zoom" thread making a reappearance that made me chuckle!  :D
Offline Fungus
« Reply #8 on: March 09, 2021, 01:28:01 pm »
Why wouldn't it keep reappearing?
Stopping the signal and zooming in/out is one of the most basic functions of a DSO. Siglents don't do it, so...  :-//
« Reply #9 on: March 09, 2021, 03:26:39 pm »
Thank you Someone, following your instructions I made some progress. That is, I can now see a lot more frames are being captured, but the play buttons still only seem to shift the current window without showing any previous history. I suppose I could fiddle with the History soft menu and shuffle through the frames but that seems pretty awkward.
I would think I could simply play the history using the replay hard buttons, but I cannot.
Thanks, again
Frank
Offline masterx81
« Reply #10 on: March 09, 2021, 04:42:26 pm »
What's the problem using a smaller timebase and then zooming in?
Offline Fungus
« Reply #11 on: March 09, 2021, 04:59:16 pm »
Try it and see!
Normally you work zoomed in so you can see what's going on. Constantly thinking "Wait.... I have to zoom out before I can press stop then zoom in again" isn't a natural way to work. Natural is simply to press stop when you see something interesting then zoom out and see what was on either side.
There's also the problem that if the event was a one-off then you can't zoom out and recapture it.
« Reply #12 on: March 09, 2021, 05:12:05 pm »
I'm not convinced it is a "problem" per se, it seems more a matter of flexibility and convenience, "how does the car feel to drive" kind of thing.
[edit:] although @Fungus' point about one-offs is a real issue if you needed to zoom out afterwards.
Offline masterx81
« Reply #13 on: March 09, 2021, 05:30:35 pm »
I have from few month an sds1104x-e (before had an owon 7102), and as i've noticed this limitation on the new device, i've taken the habit to keep the timebase a bit less than necessary, then zooming in. I not find it changing too much my workflow, i'm still happy with all the other advantages of the device. Every device have it's own "quirks". Maybe a direct "rival" (for example a Rigol) have the zoom out, but not have other features of the siglent.
« Last Edit: March 09, 2021, 05:56:01 pm by masterx81 »
The following users thanked this post: Fungus
Online 2N3055
  • Country: hr
« Reply #14 on: March 09, 2021, 06:48:09 pm »
Try it and see!
Okay let me ask you one thing: you see something, and then stop and zoom in/out?
On what signal ?
On some fast repetitive signal, by the time you see something and press stop, there was thousands of triggers passed by.
One you saw is lost forever. Zoom in or zoom out. Only way to actually catch  a transient is to use Single mode and clever trigger...
If you use persistence, you might be able to observe something happened, but not catch it, Flash....
Try it and see!
On anything that repeats faster than once per second, your statement makes no sense.
Also, that Dave's video is not exactly thorough on the topic. Not his best work. In a video talking about zoom, he never, not once actually used Zoom mode on any of those scopes. He uses name "zoom" for changing timebase on stopped capture. While it changes displayed magnification, it is not Zoom mode, and not called Zoom in documentation for any of the scopes he uses. He also never, ever, showed use of proper Zoom mode on any of those scopes as described in manuals of those scopes. So people can see how that works, and weigh in pros and cons for themselves.
Oh, i just realized, I wasn't right, if you do see something flash on the screen, you might be able to go back if you 're fast enough...
If you have one of those scopes that have History, you can go back hundreds of previous triggers, and if you are fast enough, you might go back in time.... And that is why those scopes have history mode. Because it is usefull.
Regards,
« Last Edit: March 09, 2021, 06:51:10 pm by 2N3055 »
Dr. Richard W. Hamming
Offline Fungus
« Reply #15 on: March 09, 2021, 07:17:15 pm »
You might not even want to zoom out. You might simply want to go left or right a bit to see what comes before/next. Should you have to zoom out, recapture the signal, then zoom in again just to see it? That's a whole lot of extra knob fiddling for something very common.
Here's a question: What happens on a Siglent if you zoom in then scroll the trigger point off the screen and press "stop". Can you scroll the trigger point back onto the screen and see what was there?
« Last Edit: March 09, 2021, 07:22:44 pm by Fungus »
Online nctnico
« Reply #16 on: March 09, 2021, 07:25:03 pm »
I guess you never work on single events... Not every signal is repetitive or repeating itself quickly!
To the OP: for zooming out after the fact you have choosen the wrong scope. It doesn't support that (I sincerely hope Siglent changes that soon though; I might have an application where a Siglent scope could work out well). Anyway... given the relatively low frequency of your signal the memory might not be long enough to offer more than the regular width of the screen (depending on time/div setting). If that is the case then zoom mode is the only way to force an oscilloscope to record for a longer time (forcing the samplerate down so the memory gets filled slower) than fits on the screen so there is data to scroll left/right to. In the end the amount of time a DSO can record a signal is equal to memory length / samplerate. If you have 10Mpts and 1 Gs/s samplerate then it can record 10M  / 1G = 10 milliseconds of data. If you need a longer time period, then the samplerate must go down.
« Last Edit: March 09, 2021, 08:30:55 pm by nctnico »
Offline Fungus
« Reply #17 on: March 09, 2021, 07:38:56 pm »
Yep. A signal can easily be "interesting" for a second or two then go away again, eg. a sudden burst of noise or transients on the line. Plenty of time to react and press "stop", not really enough time to go through the sequence of zooming out and then pressing stop.
Simple version: Zooming out and recapturing takes time.
Online 2N3055
  • Country: hr
« Reply #18 on: March 09, 2021, 09:22:47 pm »
Read again. I was replying to Fungus, who is simply trolling without any examples.
I know you use it and you explained how you use it, and seem to have valid reasons to do so and know how to set it up to work for you. While I many times said I disagree it is easier and that I use some other method, you can at least demonstrate what and how you're doing it.
But he is touting some magical properties and keep repeating things he doesn't understand completely, and relying on references that were proven wrong and such...
Online 2N3055
  • Country: hr
« Reply #19 on: March 09, 2021, 09:27:23 pm »
No, troll, read a manual. You don't look at something without clue, then stop and zoom  out and do all kinds of things without using brain.
You're looking at something that is a long infrequent burst, so you set timebase to capture whole burst, enable zoom, go and magnify portion of signal you want to monitor, and keep it running. If you see something, you then stop and look around. And you keep an eye on a magnified portion and whole signal at the same time.
Offline Someone
« Reply #20 on: March 09, 2021, 09:31:37 pm »
Reread all that carefully. This is one of the example situations where "zoom out after stopping" wouldn't be achievable as the memory depth is already at maximum. So its all just adding noise to the OPs question.
Offline Someone
« Reply #21 on: March 09, 2021, 09:35:51 pm »
Thanks, again
Frank
There are many many options and it can be a bit hard to find them all, the navigate controls can be set to do different things (user manual section pictured below).
Offline Fungus
« Reply #22 on: March 09, 2021, 10:25:18 pm »
I notice you're all ignoring this situation, which doesn't seem at all artificial or "trolling" to me.
Bottom line: It's not expected or natural behavior. Dave saw it and was surprised, the person who started this thread was surprised, there are quite a few threads on this now. Other 'scopes just do it without thinking, Siglent needs to pull their finger out and fix this problem.
Online Martin72
« Reply #23 on: March 09, 2021, 10:34:19 pm »
« Reply #24 on: March 09, 2021, 10:44:11 pm »
To all,
I'm managing to get what I want mostly using Someone's suggestions, but sometimes I'm not sure how I got there. I'm pretty sure I'll eventually get fluent, but it will take me some time.
I'm working on the suggestions about zooming, but so-far it's eluding me. I think a lot of things will begin to make sense as I get more familiar with the scope.
Yes, so-far with my low knowledge the replay is not as convenient as I would like, but some of the other capabilities still make me think I bought the right scope for my budget. The ability to access the scope via a PC browser on my 27 inch screen is really nice for old eyes. Not to mention the ease of capturing screen prints and annotating them for my electronic project notebook. As Fungus mentioned, hopefully Siglent will update their software some day to improve the replay capability, or maybe I get savvy enough to minimize the apparent inconvenience.
Thanks to all for the comments and suggestions. Keep them coming, I read them all, but sometimes if I don't respond quickly it's because I'm overwhelmed and I'm not sure I can comment coherently.
Frank
P.S. I read hundreds of reviews and viewed many videos before I purchased this scope, but I don't think anyone mentioned this particular "well known" replay drawback. Although I might not have purchased this scope had I known, and at this moment I think that might have been a mistake (for me).