I scored a Fluke 80K-40 HV probe. According to the manual, it's rated for "1 kV to 40 kV dc or peak ac, 28 kV rms ac." Oddly, it has a sticker on it that says, "Warning: Not authorized for use over 5000 volts." Since this probe is used, I suppose someone other than Fluke may have put that on there.
Anyway, to make sure it isn't borked, I tested it to 1000V DC with my insulation tester. It measured 1GΩ from probe tip to ground and my DMM showed the expected 1000:1 ratio on the probe output. Also, the resistance on the output banana plugs measured 1.1MΩ. Seems OK.
I don't have any plans to go around measuring 40kVDC, but it will be useful to discharge CRTs (rather than shorting them with a screw driver) and verify that HT circuits are de-energized before repairing.
One thing that's not clear to me is the probe's rating starting from 1kV. I was able to correctly measure 250VDC and 500VDC being output by my insulation tester. So, what do you suppose the issue is with voltages below 1kV?