Working on the 465B “parts” unit.
Deoxit'ed the dirty switches and pots.
Cleaned up the front. Case is in good shape but needs a good cleaning.
Both Volts/Div knobs are cracked but still functional. If they get worse epoxy will fix.
Checked +55VDC reference. Dead on.
Checked all supply voltages for value and ripple. All in spec and ripple on all 2.0mv or less.
Checked HV. 2453V vs 2450V spec. Again, dead on.
Here's the best part. Did a DC continuity check of both channel attenuator switch decks and they are perfect. Don't even need cleaning.

The bad: The trace has noise on it. It's not the power supply so could be only one thing. TANTS! If this guy winds up being a keeper I'm going to have to do a massive tant change at least on the vertical board. But I noticed as the scope warms up the noise clears slightly. All the more indication of bad caps somewhere. Going to leave the scope cook for a few hours.
The 4X attenuator for the other 465B is somewhere over the Atlantic between Athens and JFK.